Our product is a sterile (irradiation) polypropylene syringe used for administering pharmaceutical products to patients by injection (we don't supply needles). We have carried out cytotoxicity testing but not sensitisation, irritation, systemic toxicity and haemocompatibility. These are VERY expensive, we are prepared to do if necessary but before we commit is there a plausible justification for not doing these for our type of device?
Also, are the tests normally carried out at the start and end of the claimed shelf-life? Or will an up-front test suffice?
Thanks,
Philip
ps my first post - nice to be here!
Elsmar Forum Sponsor
Marcelo
Inactive Registered Visitor
#2
Re: ISO 10993 Question
You really need to perform a biological safety evaluation, the need for testing is an output of the evaluation process. See, for example, this thread.
Philip B
#3
Thanks Marcelo; our evaluation asks us to consider sensitisation, irritation, systemic toxicity and haemocompatibility (in addition to cytotoxicity which we already do). I am looking for any advice on arguments / justifications which are accepted by NBs for not carrying out these tests.
Marcelo
#4
Well, the only way to have justifications for not testing is if existing data shows that the related risk is already acceptable. Didn't you find any data that would deem the risk acceptable in the literature search or on your own data?
Philip B
#5
Thanks again; we have two products, one for which we are not the OM and the product has a long history of safe use; the other is a brand new product, I think we will have to do the testing for this
Marcelo
#6
Please note that "a long history of safe use" is not an accepted justification, you need data that shows that the risk is acceptable.
In the case of the brand new product, you mean that it's using new materials (after performing the material characterization per 10993-18)?
chris1price
Trusted Information Resource
#7
I agree with Marcelo about the need for a biological safety evaluation or obtaining appropriate data. Something else to consider is the use of the syringe. If you are selling to a pharma company to be used as a pre-filled , they will have to perform their own testing. You may be able to piggy-back of their results.
If you are selling direct to hospitals or nurses as a general purpose syringe, consider the different types of medicine and the effects these could have on the polypropylene. Just performing one set of tests may not be sufficient.
Regarding when to do the testing, it should be at the end of the shelf-life. The effects of gamma radiation on polymers continues long after the exposure. Also, the leaching of chemicals from the polypropylene will increase in concentration of any toxic substances in the drug over time.
Thread starter Similar threads Forum Replies Date
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
R ISO 10993-18 Medical Device and FDA Regulations and Standards News 1
M ISO 10993-17 - Ancillary Medicinal substance leachables Medical Device and FDA Regulations and Standards News 1
M Medical Devices - Ancillary - ISO 10993-17 Medical Device and FDA Regulations and Standards News 6
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
L Biological Assessment (ISO 10993-1) Other Medical Device Related Standards 1
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
A ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations Other Medical Device Related Standards 3
P European versions of standards - ISO 10993-1 and ISO 11607-1 EU Medical Device Regulations 6
S ISO 10993 Biocompatibility Evaluation - Electronic thermometer Other Medical Device Related Standards 3
P ISO 10993-1:2018 - When will the European equivalent become a Harmonized Standard? Other Medical Device Related Standards 13
I ISO 10993 Biocompatibility Requirements - manufacturing process chemicals Other Medical Device Related Standards 8
N Technical File Reviewer has requested more testing to ISO 10993 Other Medical Device Related Standards 10
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
planB ISO 10993-1:2018 has just been published Other Medical Device Related Standards 2
M Does ISO 10993 Apply to Packaging? (e.g. Travel Case) Other Medical Device Related Standards 4
K ISO 10993 Requirements For 510k - Substantial Equivalence Other US Medical Device Regulations 11
T Using Risk Management in ISO 10993 - Medical Device Accessory 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Actual Definition of Non-Contact for ISO 10993 Other Medical Device Related Standards 1
R Biocompatibility - ISO 10993 and SVHC Requirements Other Medical Device Related Standards 4
S ISO 10993-1: External Communicating Device biocompatibility test requirements Other ISO and International Standards and European Regulations 4
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
T Plastic of an infusion pump enclosure and ISO 10993 ISO 13485:2016 - Medical Device Quality Management Systems 2
E Upcoming changes for ISO 10993 and IEC 60601 in South Korea Other Medical Device Regulations World-Wide 2
M FDA issues draft guidance to replace G95-1 (Application of ISO 10993) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
B Verify GB/T 16886 is the same as ISO 10993 US Food and Drug Administration (FDA) 5
B New ISO 10993-12 - Consistency of Extraction Conditions through different Assays Other ISO and International Standards and European Regulations 1
S Suppliers of ISO 10993 compliant Material for Skin Contact Other ISO and International Standards and European Regulations 3
M ISO 10993 - Repeat tests necessary if colour changes slightly? Other Medical Device Related Standards 11
M Pre-Selecting Materials for ISO 10993-1 Biocompatibility Compliance Other ISO and International Standards and European Regulations 5
F ISO 10993-10 - Removed from List of Harmonized Standards Other Medical Device Related Standards 4
S ISO 10993 Cytotox Testing - Direct Contact rather than Extractables Other Medical Device Related Standards 4
E Meeting ISO 10993-1 2009 Material Risk Assessment Requirements Other Medical Device Related Standards 13
M ISO 10993 Labs which can test Cytotoxicity, Sensitizaton and Irritation Other Medical Device Related Standards 10
M ISO 10993 for Fabric based Electrode - Finding already FDA Approved Materials Other Medical Device Related Standards 11
J Biological Evaluation requirements per ISO 10993 - Medical Device Testing ISO 13485:2016 - Medical Device Quality Management Systems 2
A Claiming Biocompatibility without ISO 10993-1 Other Medical Device Related Standards 19
J ISO 10993 (Biological Compatibility) Approved Materials US Food and Drug Administration (FDA) 23
T ISO 10993-10: 2002 vs 2010 version - What are the Major Changes? ISO 13485:2016 - Medical Device Quality Management Systems 2
W ISO 10993 (Biological Evaluation of Medical Devices) for Suppliers Other ISO and International Standards and European Regulations 8
I EO (Ethylene Oxide) Residuals - EN ISO 10993-7:2008 EU Medical Device Regulations 5
D Does anyone know of a biocompatible ABS tested & certified to ISO 10993-1 Other Medical Device Related Standards 5
W ISO 10993 - Replacing Sensitization Test with Systemic Toxicity Test? Other Medical Device Related Standards 4
I ISO 10993-10:2010 Guidance Document for Irritation and Sensitization Test - Changes ISO 13485:2016 - Medical Device Quality Management Systems 3
M EO (Ethylene Oxide) and ECH (Ethylene Chlorohydrin) Residual as per ISO 10993-7:2008 ISO 13485:2016 - Medical Device Quality Management Systems 15
I ISO 10993-1:2009 Sensitization (Initial Evaluation Test) vs. USP Other Medical Device Related Standards 3
B What ISO 10993 considers Non-Invasive vs. Invasive MDD 93/42/EEC EU Medical Device Regulations 3
C Differences between the biocompatibility standards ISO 7405 and ISO 10993 Other Medical Device Related Standards 2