ISO 11135:2014, B.1.4, BI resistance x product bioburden

akyt

Registered
#1
Hello guys, could you help me?

The standard ISO 11135: 2014, item B.1.4, says that the resistance of the product's bioburden must be less than the resistance of BI. I read the guide (item D.14.1.4), but I still have doubts about how I could run the tests to meet this item.

The standard talks about a fractional cycle and the sterility test after the cycle, but if there is growth in the BI and the product, how can I guarantee that the BI is still more resistant? Should the exposure time of this fractional cycle be defined to allow BI growth? Should I use product samples as if it were a routine sterilization? And how do I define the sampling?

If someone has another test suggestion to ensure that the BI is more resistant than the product's bioburden, please share. I will be very grateful for your support.

Thank you in advance.
 
Elsmar Forum Sponsor

levatorsuperioris

Involved In Discussions
#2
Characterize your bioburden (population) with fractional cycles, if your first fractional cycle kills everything in both your BI and your population, you have to restart. Its not uncommon to find your bioburden dead after your first fractional cycle, but your BI is not. This situation is normally ok, but you need to ensure the population of bioburden is => BI when you start.

If you are using Annex A you need one Fractional Cycle, if you are using Annex B you need multiple Fractional cycles to build a curve. You should likely hire a professional certified lab to design these as accuracy is critical.

You can use multiple fractional cycles per annex B in this situation you describe above.
 

adir88

Involved In Discussions
#3
If there is growth in both the product and BI, you need to increase the exposure until you get sterility in product. You could run some preliminary fractional cycles to work out the exposure time that gives you sterility for product but growth in your BI. Once you're happy, you can formally run it under a validation protocol to demonstrate that the product's bioburden is less than the resistance of the BI. Basically, you cannot pass your fractional cycle until you get all sterile (no growth) product.

Hope this helps.

Cheers.
 

akyt

Registered
#4
Thank you so much for your contributions.
Now I could understand the fractional cycle strategy, but I have a question:
Can I use the product's natural bioburden (in its normally found population), or should I inoculate a known population (greater than or equal to the population of BI, 10^6) in a sterile product in order to compare resistance?
And having this resistance defined, how often should I redo this validation?
 

chris1price

Trusted Information Resource
#5
No, use the natural bioburden. You are trying to show that the product is easier to sterilise than the BI. That way, when the BI is fully sterilised, you know the product will be as well.
 

levatorsuperioris

Involved In Discussions
#6
The point is to find out if the natural bioburden is more resistant to sterilization than your BI, innoculating would ruin the point, ie strains of species can have different resistances to Sterilization Methods -> famously troublesome was Gluteraldehyde and Mycobacterium Spores .... There are devices with extremely low CFU counts naturally, you work with what you have.
 

adir88

Involved In Discussions
#7
Agree with chris1price and levatorsuperioris. The point is to compare the natural resistance of the product to the BI.
 
Thread starter Similar threads Forum Replies Date
planB ISO 11135:2014 cancels and replaces ISO 11135-1:2007 and ISO/TS 11135-2:2008 Other Medical Device Related Standards 15
I Certification in Sterilization for ISO 11135 and ISO 11137 Professional Certifications and Degrees 3
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 10
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
G EtO Sterilization - DIN/EN 550 - ISO 11135 Other US Medical Device Regulations 20
I The Status of EN 550 and Whether it will be Replaced Soon by ISO 11135 ISO 13485:2016 - Medical Device Quality Management Systems 5
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
J Opportunity in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
P ISO 8 classified medical manufacturing room Qualification and Validation (including 21 CFR Part 11) 1
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 13
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
Sidney Vianna Release of ISO 10013:2021, Quality management systems – Guidance for documented information Other ISO and International Standards and European Regulations 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
R Remote Audits for ISO 9001 (or any other standard) General Auditing Discussions 29
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
D ISO 17021 Certificate Registrars and Notified Bodies 1
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
S Sequence of ISO 9001:2015 Implementation Steps ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
qualprod Business Continuity Planning in ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
T Help with BS EN ISO - IEC 80079-34 2020 (Explosive atmospheres QMS) Other ISO and International Standards and European Regulations 0
C Sampling - ISO 16269-6 vs ISO 2859 Other ISO and International Standards and European Regulations 0
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 9001:2015 Internal Auditing Internal Auditing 8
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
N ISO 27001 for Jumb Burger - Risk Assessment sheet IEC 27001 - Information Security Management Systems (ISMS) 10
H What ISO certification is for an IT department? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
P ISO 9001 certification with zero customers? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4

Similar threads

Top Bottom