ISO 11135:2014, B.1.4, BI resistance x product bioburden

akyt

Registered
#1
Hello guys, could you help me?

The standard ISO 11135: 2014, item B.1.4, says that the resistance of the product's bioburden must be less than the resistance of BI. I read the guide (item D.14.1.4), but I still have doubts about how I could run the tests to meet this item.

The standard talks about a fractional cycle and the sterility test after the cycle, but if there is growth in the BI and the product, how can I guarantee that the BI is still more resistant? Should the exposure time of this fractional cycle be defined to allow BI growth? Should I use product samples as if it were a routine sterilization? And how do I define the sampling?

If someone has another test suggestion to ensure that the BI is more resistant than the product's bioburden, please share. I will be very grateful for your support.

Thank you in advance.
 
Elsmar Forum Sponsor

levatorsuperioris

Involved In Discussions
#2
Characterize your bioburden (population) with fractional cycles, if your first fractional cycle kills everything in both your BI and your population, you have to restart. Its not uncommon to find your bioburden dead after your first fractional cycle, but your BI is not. This situation is normally ok, but you need to ensure the population of bioburden is => BI when you start.

If you are using Annex A you need one Fractional Cycle, if you are using Annex B you need multiple Fractional cycles to build a curve. You should likely hire a professional certified lab to design these as accuracy is critical.

You can use multiple fractional cycles per annex B in this situation you describe above.
 

adir88

Involved In Discussions
#3
If there is growth in both the product and BI, you need to increase the exposure until you get sterility in product. You could run some preliminary fractional cycles to work out the exposure time that gives you sterility for product but growth in your BI. Once you're happy, you can formally run it under a validation protocol to demonstrate that the product's bioburden is less than the resistance of the BI. Basically, you cannot pass your fractional cycle until you get all sterile (no growth) product.

Hope this helps.

Cheers.
 

akyt

Registered
#4
Thank you so much for your contributions.
Now I could understand the fractional cycle strategy, but I have a question:
Can I use the product's natural bioburden (in its normally found population), or should I inoculate a known population (greater than or equal to the population of BI, 10^6) in a sterile product in order to compare resistance?
And having this resistance defined, how often should I redo this validation?
 

chris1price

Trusted Information Resource
#5
No, use the natural bioburden. You are trying to show that the product is easier to sterilise than the BI. That way, when the BI is fully sterilised, you know the product will be as well.
 

levatorsuperioris

Involved In Discussions
#6
The point is to find out if the natural bioburden is more resistant to sterilization than your BI, innoculating would ruin the point, ie strains of species can have different resistances to Sterilization Methods -> famously troublesome was Gluteraldehyde and Mycobacterium Spores .... There are devices with extremely low CFU counts naturally, you work with what you have.
 

adir88

Involved In Discussions
#7
Agree with chris1price and levatorsuperioris. The point is to compare the natural resistance of the product to the BI.
 
Thread starter Similar threads Forum Replies Date
planB ISO 11135:2014 cancels and replaces ISO 11135-1:2007 and ISO/TS 11135-2:2008 Other Medical Device Related Standards 15
I Certification in Sterilization for ISO 11135 and ISO 11137 Professional Certifications and Degrees 3
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
S ETO ISO 11135-1 & ISO 17665-2 - Purity and Quality of Steam (Humidity) Qualification and Validation (including 21 CFR Part 11) 3
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 12
F Implementing ISO 11135 - Facilities Conditions - Adding ETO Sterilization Other Medical Device Related Standards 4
I ETO Sterilizer Validation (ISO 11135:2007) Other Medical Device Related Standards 23
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
K EN550 (ISO 11135-1:2007) - ETO (Ethylene Oxide) Validation Qualification and Validation (including 21 CFR Part 11) 5
B ISO 11135:2007 (EtO Sterilization) ISO 13485:2016 - Medical Device Quality Management Systems 5
G EtO Sterilization - DIN/EN 550 - ISO 11135 Other US Medical Device Regulations 20
I The Status of EN 550 and Whether it will be Replaced Soon by ISO 11135 ISO 13485:2016 - Medical Device Quality Management Systems 5
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 0
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 6
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
D "certified" in ISO 19011, as well as IATF required? IATF 16949 - Automotive Quality Systems Standard 4
S ISO/IEC 15408 - Is this is Certifiable Standard? Other ISO and International Standards and European Regulations 2
D Lead time to schedule an ISO 13485 audit Auditing Quality and Environmental Management Systems 2
T Do we need an SOP for ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
K ISO 9001 Auditing in a Healthcare setting Auditing Quality and Environmental Management Systems 15
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 1
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 2
Y How can i integrate ISO 13845 into ISO 27001? ISO 13485:2016 - Medical Device Quality Management Systems 4
vickyva ISO 14155:2020 CIP CIR templates Other Medical Device Related Standards 0
C ISO 9001:2015 8.3.2. h) Design and Development Planning - What is required? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Employee Handbook in ISO 9001:2015 Section 7 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 10
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
G Trying to get a financially reasonable ISO Certification Body Registrars and Notified Bodies 8
L ISO 45001:2018 - Clause 5.4: Consultation and Participation of Workers Process Maps, Process Mapping and Turtle Diagrams 1
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
Richard Regalado Informational ISO/IEC DIS 27001:2021, to be published soon. IEC 27001 - Information Security Management Systems (ISMS) 0
Q Audit report template ISO 9001/14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11

Similar threads

Top Bottom