Hi everyone-

What is the typical lead time for a quarterly dose audit of a gamma sterilized device?

My packaging/sterilization vendor is claiming that it takes 8 weeks, and they need to hold the entire lot of product for those 8 weeks post sterilization. There has been no significant change in product or manufacturing since the last dose audit.

However, our in-house sterilization "expert" says the test only takes 2 days.

This doesn't make sense to me - why would anyone use gamma if 4 times a year their shipments are delayed by 8 weeks?

Thanks very much for any input!
 
Elsmar Forum Sponsor
M

MIREGMGR

#3
I wouldn't call myself a subject matter expert, but:

What test are they doing over that eight weeks? That's a long time for lethality testing.

Are they also running testing for ISO 10993 biocompatibility, i.e. pyrogenicity of your materials at the designated dose? I assume that was tested and cleared during your original process validation. If your specific materials haven't changed and the dose hasn't changed, I wouldn't be clear on why new pyrogenicity testing would be required.

You must have asked for more information. What explanation have they offered you?
 

chris1price

Trusted Information Resource
#4
Hi

Our quartly dose audits (for VDMax25) only take a few weeks. Probably a week for bioburden and 2 for sterility testing. Pyrogen testing normally only takes 2 weeks. As this is all part of a routine, historically well controlled process, we don't hold up production waiting for the results.

You may want to check, but I believe if you have 12 months of good information and reliable bioburden data, you can reduce the sterility testing to once a year, but keep quarterly bioburden. Have a look at 11137-1 Section 12.1.3.2 and see if it applies to your process.

Chris
 
J

Jimmy the Brit

#5
I agree with Chris, the dose audit testing can be done on samples taken from the batch and certainly does not require the retention of the whole batch while the testing is being done. Its purpose is to demonstrate that the orginal dose validation versus your finished product bioburden remains valid, it is intended to pick up changes in the resistance and composition of the bioburden, and, as a result, is an ongoing commitment to verify the original dose choice was appropriate. Tell your irradiator to stop messing about, take the samples they need and then release the batch as normal.

Your in house expert is right in that the testing only requires two days, however it requires a further 14 days incubation time, and may require some dose sensitivity analysis which can take a further few days. Still no reason to sit on the batch though.

Jimmy
 
A

arrabellaangel

#6
Your in house expert is right in that the testing only requires two days, however it requires a further 14 days incubation time, and may require some dose sensitivity analysis which can take a further few days. Still no reason to sit on the batch though.

Jimmy
i agree with jimmy..our sterilizer only takes 5 days for bioburden and sterility test need 14 days...total lead only 3 - 4 weeks only..
 
K

Kees Fremery

#7
Indeed a Periodic Dose Audit can be performed within 3-4 weeks. First of all irradiate the samples using the initial varification dose and test for sterility. During those two weeks of incubation the Bioburden test can be performed. No need to do the Bioburden test before the verification experiment because the verification dose already is determined.
 
A

Aqeel Abbas

#8
We are planning to use gamma radiation for the sterilization of our single use medical device. The materials used in the device are all compatible with gamma radiation. In our country there is only one goverbment organization that offer gamm radaitioon servcies. They have clients like Johnsosn and Johnsons and other big local pharamceutical companies.

This organization however is not certified for ISO 13485.

They do help in process validation to develop a specific process for our devices.

My question is that can we subcontract their services? If the bioburden, sterility and pyrogencity tests performed as part of process validation are clear, can these alone be considered as performance qualification? Do we or the local notified body has to audit this organization? Are there any specific qualification for the auditor who can audit a gamma sterilization plant?

Thanks in advance for help.
 
Thread starter Similar threads Forum Replies Date
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
S ISO 11137- Simulated product vs SIP Other Medical Device Related Standards 2
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
I Certification in Sterilization for ISO 11135 and ISO 11137 Professional Certifications and Degrees 3
M ISO 11137 - Verification dose audit for VD max method Other Medical Device Regulations World-Wide 7
S If this EN ISO 11137 certificate acceptable for Contract Sterilization? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Does anyone have a summary of changes made to ISO 11137-2? Other Medical Device Related Standards 3
F ISO 11137 - Flow Chart | Decision Tree sterilization Other Medical Device Related Standards 4
K Very small batch validation ISO 11137 (VDmax25) ISO 13485:2016 - Medical Device Quality Management Systems 6
S Controlled Environment for Manufacturing Sterile Product - ISO 11137 Other Medical Device Related Standards 2
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
C ISO 11137:1995 Sterilisation of healthcare products - Amd1:2001 ISO 13485:2016 - Medical Device Quality Management Systems 11
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet ISO 26262 ISO 26262 - Road vehicles – Functional safety 3
Marc ISO 26262- Road vehicles – Functional safety ISO 26262 - Road vehicles – Functional safety 0
John C. Abnet ISO 26262 IATF 16949 - Automotive Quality Systems Standard 0
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
K PDCA cycle and ISO processes alternative model Quality Management System (QMS) Manuals 14
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
D Which ISO Standard to purchase? ISO 13485:2016 - Medical Device Quality Management Systems 7
V ISO 10360-5: 2020 Gap analysis and Action plan Excel .xls Spreadsheet Templates and Tools 1
Q ISO 9001 - Reseller Exclusions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO 45001 6.1.2.1 Hazard Identification Occupational Health & Safety Management Standards 1
T The difference between ISO 14644-3:2005 and ISO 14644:2019 Other Medical Device Related Standards 2
S Any ISO standards around Artificial Intelligence and Machine Learning? Medical Information Technology, Medical Software and Health Informatics 4
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
N ISO 9001 - Training business with fewer than 5 employees ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4

Similar threads

Top Bottom