ISO 11137 - Verification dose audit for VD max method

M

muzammil21ad

#1
Hi All,

I need one help with respect to verification dose audit for VD max method, we are validating our gamma sterilization process, i face problem with respect to verification dose audit, my gamma subcontractor have dose range starting at 9kGy only but my bio burden is less so the verification dose according to table 9 of 11137 is 7.5kGy,
so please let me know how i should complete my verification dose audit with this type of challenge

:thanx:
 
Elsmar Forum Sponsor

Stijloor

Staff member
Super Moderator
#2
A Quick Bump!

This looks like a fairly complex issue, but can someone help with this question?

Thank you very much!
 
B

brianthermophilus

#3
If that's your only dose option and your bioburden is lower, I would run your samples through 9kgy and perform material & packaging assessments to confirm that your product may withstand a higher dose. Nothing wrong with a higher dose as long as your product can handle it.
 
M

muzammil21ad

#4
If that's your only dose option and your bioburden is lower, I would run your samples through 9kgy and perform material & packaging assessments to confirm that your product may withstand a higher dose. Nothing wrong with a higher dose as long as your product can handle it.


Thanks for your comments, this is verification dose to have confidence on my sterility not material compatibility excersise


Sent from my iPhone using Tapatalk
 

pkost

Trusted Information Resource
#5
I would suspect using a higher dose will make the validation ineffective - from recollection the dose is low enough to give a an SAL of 1x10^-2....The subsequent sterilisation tests are sufficiently powered establish the SAL is validated....if you use a higher verification dose then the sterility tests will not be sufficiently powered

I think it is permitted to carry out the VDmax study at any facility, it does not have to be the main site of sterilisation - if you can find somewhere that operates at 7.5kGy, that may be a solution
 
Last edited:
M

muzammil21ad

#6
I think doing dose mapping with 2 different other sources will give you different uncertainty and sterilizing agent characteristics , i am not sure if standard allow me to do ths because i am not as much expert than you may be pls correct me if i am wrong, if it is ok thn my prblm is solved


Sent from my iPhone using Tapatalk
 

pkost

Trusted Information Resource
#7
From my understanding, the dose setting is to demonstrate that the the 25kGy is sufficient to give an SAL of 10-6...25kGy is the same regardless of equipment used; the sterilising agent is gamma radiation...it's consistent

Dose mapping is different as it demonstrates that the entire load receives a dose of 25kGy...I suspect that this may need to be carried out in the intended sterilising facility, but I'm not sure.

The lab that is conducting the validation for you should be able to answer these questions and provide more qualified guidance!
 
Thread starter Similar threads Forum Replies Date
S ISO 11137- Simulated product vs SIP Other Medical Device Related Standards 2
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
I Certification in Sterilization for ISO 11135 and ISO 11137 Professional Certifications and Degrees 3
S If this EN ISO 11137 certificate acceptable for Contract Sterilization? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Does anyone have a summary of changes made to ISO 11137-2? Other Medical Device Related Standards 3
F ISO 11137 - Flow Chart | Decision Tree sterilization Other Medical Device Related Standards 4
K Very small batch validation ISO 11137 (VDmax25) ISO 13485:2016 - Medical Device Quality Management Systems 6
S Controlled Environment for Manufacturing Sterile Product - ISO 11137 Other Medical Device Related Standards 2
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
E ISO 11137 gamma sterilized device dose audit - Lead time? Other ISO and International Standards and European Regulations 7
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
C ISO 11137:1995 Sterilisation of healthcare products - Amd1:2001 ISO 13485:2016 - Medical Device Quality Management Systems 11
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 10
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
G Trying to get a financially reasonable ISO Certification Body Registrars and Notified Bodies 7
L ISO 45001:2018 - Clause 5.4: Consultation and Participation of Workers Process Maps, Process Mapping and Turtle Diagrams 0
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
Richard Regalado Informational ISO/IEC DIS 27001:2021, to be published soon. IEC 27001 - Information Security Management Systems (ISMS) 0
Q Audit report template ISO 9001/14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet ISO 26262 ISO 26262 - Road vehicles – Functional safety 3
Marc ISO 26262- Road vehicles – Functional safety ISO 26262 - Road vehicles – Functional safety 0
John C. Abnet ISO 26262 IATF 16949 - Automotive Quality Systems Standard 0
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
K PDCA cycle and ISO processes alternative model Quality Management System (QMS) Manuals 14
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2

Similar threads

Top Bottom