ISO 11737 - Bioburden testing for non-sterile products

C

Cec28007

#1
What type of microbial testing is required for non-sterile product containers before filling? I have not found any regs. that require testing for incoming containers. Does anyone do any microbial testing on their non-sterile product containers? We are planning on following the ISO 11737 for bioburden and then simply disregarding the information on sterility, since our products are not claiming sterility. This will give us a good indication of what microbial load our containers are coming in with, then we just monitor as we would any environmental monitoring. Any input is appreciated.


Clint Christensen
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: Bioburden testing for non-sterile products

Out of curiousity, what product type are you making?

I have not found any regs. that require testing for incoming containers.
I don't know of any such regs either, but my focus is primarily medical devices for which the sterile/non-sterile regulatory division is rather absolute.

Does anyone do any microbial testing on their non-sterile product containers?
We don't. Non-sterile for us means that the product's intended use involves contact with normal unbroken skin at most, and that there is no expectation on the part of the clinical user or the patient of absence of typical environmental microorganisms on surfaces.
 
Last edited by a moderator:
C

Cec28007

#3
Re: Bioburden testing for non-sterile products

Out of curiousity, what product type are you making?
We manufacture a variety of dental products. Some of which include hemostats, bleaching gels, toothpastes and fluoride treatments. All intended for use in the oral cavity of course.
 
J

Jimmy the Brit

#4
Re: Bioburden testing for non-sterile products

We manufacture a variety of dental products. Some of which include hemostats, bleaching gels, toothpastes and fluoride treatments. All intended for use in the oral cavity of course.
Whoa - I had no idea hemostats were non-sterile - doesn't this fly in the face of the requirements of ANSI/AAMI ST67 Annex B? Even tongue depressors are expected to be supplied sterile, albeit with an SAL of just 10 x -3.

I speak from a position of no-knowledge of dental supply requirements, I am just surprised that a "device" that is required to have blood and mucosal contact as part of its function is not even required to have a bioburden limit.

Anyway, back to your original question, no I don't believe that there are limits for bioburden for non-sterile devices, other than those limits that you may choose to impose yourself to demonstrate control of process. Bioburden is really only useful to ascertain the magnitude and resistance of the challenge to the sterilization process.

Since toothpastes, gels and treatments are quite close to pharmaceutical oral dosage forms you could always use the USP2007 limits defined in section <1111> Microbiological Examination of Non-Sterile Products: acceptance criteria for pharmaceutical preparations and substances for pharmaceutical use. Table 1 states a limit of 100 cfu/ml total aerobic microbial count and 10 cfu/ml for yeasts & moulds (combined). It also includes a test for two objectionable microorganisms in that dosage form Staph. aureus and Pseudomonas aeruginosa - pathogens which have the potential for human infection - easy enough to exclude from the process, but by monitoring you can protect yourself from legal claims from patients/dentists trying to pass the responsibility.

Not sure if this helps at all, or if I have just made your world much more complicated than it needs to be,

Jimmy
 
Last edited by a moderator:

Ajit Basrur

Staff member
Admin
#5
Re: Bioburden testing for non-sterile products - ISO 11737

What type of microbial testing is required for non-sterile product containers before filling? I have not found any regs. that require testing for incoming containers. Does anyone do any microbial testing on their non-sterile product containers? We are planning on following the ISO 11737 for bioburden and then simply disregarding the information on sterility, since our products are not claiming sterility. This will give us a good indication of what microbial load our containers are coming in with, then we just monitor as we would any environmental monitoring. Any input is appreciated.

Clint Christensen
Welcome to the Cove, Clint :bigwave:

Jimmy has given great advise. To add on, some of the guidleines that have mentioned about the Bioburden in dental devices are -

Guidelines for Infection Control in Dental Health-Care Settings - 2003

Additonal info also available at Infection Control in Dental Settings

Hope this is useful to you - should you have more queries, pl revert back :)
 

6thsense

Involved In Discussions
#7
well I work for a beverage making concern and The owners of the company have a requirement for Bioburden analysis of Bottles and Closures. the procedure is really good and i strongly recommend that u do it.
 
C

Cec28007

#8
well I work for a beverage making concern and The owners of the company have a requirement for Bioburden analysis of Bottles and Closures. the procedure is really good and i strongly recommend that u do it.
I would be interested to know what procedures you follow. Are you rinsing the interior of the bottle with an extraction solvet? How do you validate the testing? Any more information on the testing procedure you use would be helpful. Thanks!!
 

6thsense

Involved In Discussions
#9
The Bioburden Method

Preparing the extraction fluid

Requirements are:
1 ml Tween 80 (polysorbate 80)
1 ltre deionized water
2.5g peptone

Place all the above in an Autoclavable Flask (can use aluminium foil/cap to close the the top of the flask) and Autoclave the flask at 121 degrees celcius for 15mins.

When you have the Extraction fluid you can proceed. for the bottles we approximate 50ml of extraction fluid/bottle that you pour inside the bottle and swirl. then membrane filteration to plate the sample:
;)
 
E

eeckler

#10
Re: Bioburden testing for non-sterile products

Whoa - I had no idea hemostats were non-sterile - doesn't this fly in the face of the requirements of ANSI/AAMI ST67 Annex B? Even tongue depressors are expected to be supplied sterile, albeit with an SAL of just 10 x -3.

I speak from a position of no-knowledge of dental supply requirements, I am just surprised that a "device" that is required to have blood and mucosal contact as part of its function is not even required to have a bioburden limit.

Anyway, back to your original question, no I don't believe that there are limits for bioburden for non-sterile devices, other than those limits that you may choose to impose yourself to demonstrate control of process. Bioburden is really only useful to ascertain the magnitude and resistance of the challenge to the sterilization process.

Since toothpastes, gels and treatments are quite close to pharmaceutical oral dosage forms you could always use the USP2007 limits defined in section <1111> Microbiological Examination of Non-Sterile Products: acceptance criteria for pharmaceutical preparations and substances for pharmaceutical use. Table 1 states a limit of 100 cfu/ml total aerobic microbial count and 10 cfu/ml for yeasts & moulds (combined). It also includes a test for two objectionable microorganisms in that dosage form Staph. aureus and Pseudomonas aeruginosa - pathogens which have the potential for human infection - easy enough to exclude from the process, but by monitoring you can protect yourself from legal claims from patients/dentists trying to pass the responsibility.

Not sure if this helps at all, or if I have just made your world much more complicated than it needs to be,

Jimmy
I've been very interested in this area myself. We'll be manufacturing a single-use disposable with vaginal mucosal contact. It has been our read of AAMI ST67 that it applied only to devices claiming sterility, not as a supply mandate. Most vaginal specula are sold clean, non-sterile. I have been looking for cleanliness standards/allowable microbial count similar to the USP standards for drugs,but for nonsterile medical devices.
 
Thread starter Similar threads Forum Replies Date
O Bioburden Testing - Non Sterile Product / Tapes - ISO 11737-1 Other ISO and International Standards and European Regulations 9
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
L EN ISO vs. ISO standards - Why to use ISO 11737-1 instead of EN ISO 11737-1 Other Medical Device and Orthopedic Related Topics 24
A ISO 11737 - Sterilization of Medical Devices - Microbiological Methods EU Medical Device Regulations 1
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Q ISO 9001/IATF 16949 Audit Finding Question - Document Retention IATF 16949 - Automotive Quality Systems Standard 10
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
G Trying to get a financially reasonable ISO Certification Body Registrars and Notified Bodies 7
L ISO 45001:2018 - Clause 5.4: Consultation and Participation of Workers Process Maps, Process Mapping and Turtle Diagrams 0
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
Richard Regalado Informational ISO/IEC DIS 27001:2021, to be published soon. IEC 27001 - Information Security Management Systems (ISMS) 0
Q Audit report template ISO 9001/14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet ISO 26262 ISO 26262 - Road vehicles – Functional safety 3
Marc ISO 26262- Road vehicles – Functional safety ISO 26262 - Road vehicles – Functional safety 0
John C. Abnet ISO 26262 IATF 16949 - Automotive Quality Systems Standard 0
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
K PDCA cycle and ISO processes alternative model Quality Management System (QMS) Manuals 14
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S ISO 11137- Simulated product vs SIP Other Medical Device Related Standards 2
D Which ISO Standard to purchase? ISO 13485:2016 - Medical Device Quality Management Systems 7
V ISO 10360-5: 2020 Gap analysis and Action plan Excel .xls Spreadsheet Templates and Tools 1
Q ISO 9001 - Reseller Exclusions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO 45001 6.1.2.1 Hazard Identification Occupational Health & Safety Management Standards 1
T The difference between ISO 14644-3:2005 and ISO 14644:2019 Other Medical Device Related Standards 2
S Any ISO standards around Artificial Intelligence and Machine Learning? Medical Information Technology, Medical Software and Health Informatics 4
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2

Similar threads

Top Bottom