We are a contract manufacture who are currently iso 13485:2003. Our transition to 2016 is in September. Although 13485 we do not manufacture any medical devices. I have listed all regulatory elements within the standard as N/A within our manual as we do not work to any reg requirements. Is this acceptable??
Marcelo
Hello, and welcome to the Cove!
I'm pretty sure some certification body will accept what you did as ok (I've seen crazier stuff accepted :p), but ISO 13485 does not have an option for that. You can have NA clauses in clause ˆ, 7 or 8 depending wither in the nature of your device (sterile, for example), or activities you do not perform (such as several if you are not a manufacturer).
Even as a contract manufacturer, you probably have several regulatory requirements applicable to you. For a general example, a contract manufacturer performs manufacturing in the name of the "legal"manufacturer, and thus the contract manufacturer complies with regulatory QMS requirements for the activity they perform (although the full responsibility is still with with the "legal"). That's one of the reasons we added a "quality agreement" to the third party activities requirements in 4.1.5, the most important thing this quality agreement has is a definition of which activities and responsibilities of each party in relation to compliance with regulatory requirements.
So, per 4.1.1 you have to define your role in the applicable regulatory requirements (as I mentioned elsewhere, this is probably the most important clause of the 2016 version of the standard) and contract manufacturer is a clear role in several regulatory systems. Then you have to determine and comply with (per 5.2) the regulatory requirements for each of those roles (probably together with you contracting clients as they will define in quality agreements what regulatory requirements you need to comply with).
So, no, in my opinion, it's not acceptable.
G
gmaltby
I understand what you are saying but we manufacture essentially tubing. We have no idea in most cases what this tubing is even used for so cannot comply to regulatory requirements as we do not know what they would be. In addition no customer has ever requested we meet any specific regulatory requirements?
Marcelo
Ah ok, you do not manufacture a device. But then are you really a contract manufacturer (third party) for tubing or a supplier?
Marcelo
Sorry, which one? They are two different things.
- Contract manufacturer for tubing (which manufacturers a tubing following "legal"manufacturer design and requirements)
Or
- Tubing (your own design and requirement) supplier (that complies with customer requirements)?
If a contract manufacturer, you do have regulatory requirements to fulfill, even if your "legal"manufacturer has never asked for that (this is a usual problem because most of the time they don't know).
If you are a supplier, you won't have in principle regulatory requirements to comply.
Also, for a contract manufacturer, there's no "customer", you are them (you act on their behalf).
Marcelo
So you do have regulatory requirements to comply with, as you are doing the work they should be doing (which is the idea of the third party mentioned in 4.1.5 of ISO 13485).
G
gmaltby
In what sense? For example they send us a request to manufacture 4.0 x 7.0mm tubing with a defined tolerance in a specific material nothing more? How would we know what regulatory requirements are required as we have other information?
Marcelo
They should have a quality agreement with you that tells you which regulatory requirement you would fulfill in their name. As I mentioned, it's more their problem than your, but this does not mean that you do not have any applicable regulatory requirements to fulfill.
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
D Audit Report details when ISO 13485:2016 and cGMP 21 CFR 820 are applicable ISO 13485:2016 - Medical Device Quality Management Systems 6
E Theoretical project: Implementing ISO 13485:2016 into a start up acrylic bone cement manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
M ISO 13485: 2016 Lot numbering question ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Documented Evidence of Training ISO 13485: 2016 ISO 13485:2016 - Medical Device Quality Management Systems 33
D Customer Survey Example - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
M Requirements ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 12
I EN ISO 13485:2016/A11:2021 ISO 13485:2016 - Medical Device Quality Management Systems 0
A ISO 13485:2016 Clause 8.2.1 ISO 13485:2016 - Medical Device Quality Management Systems 3
FuzzyD ISO 13485:2016 Clause 8.2.6 ISO 13485:2016 - Medical Device Quality Management Systems 4
K Can I make an exclusion of Design and Development in ISO 13485:2016 if my product is not regulated ISO 13485:2016 - Medical Device Quality Management Systems 12
B ISO 9001:2015 vs ISO 13485:2016 Gap analysis ISO 13485:2016 - Medical Device Quality Management Systems 9
K 3rd party auditor for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 4
S Electronic Signatures - Non-Conformance - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 36
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 4
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
H ISO 13485:2016 Gap Analysis by NB ISO 13485:2016 - Medical Device Quality Management Systems 7
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 16
JoCam Difference between Approval and Registration - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
T Document control ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
L ISO 13485:2016 Clause 8.4 - Analysis of Audit Observations ISO 13485:2016 - Medical Device Quality Management Systems 8
S When is ISO 13485:2016 6.4.2 Contamination Control appropriate? ISO 13485:2016 - Medical Device Quality Management Systems 11
L Templates for three ISO 13485:2016 SOPs ISO 13485:2016 - Medical Device Quality Management Systems 8
C What falls under the 'Customer Property' according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 29
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1