ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations?
ISO 8.5.3 d and 8.5.4 d state:
verifying that the action does not adversely affect the ability to meet applicable regulatory
requirements or the safety and performance of the
medical device;
One NB auditor insisted that every CAPA has to be checked or approved by regulatory department. We think that is only related to the device, otherwise the standard would say so. For instance, a CAPA triggered by QA related training , which is not related to the 'medical device' does not need RA" blessing"- your opinion?
blackholequasar
I would think as long as the organization verifies that the action does not adversely affect the requirements, then there is really no need to have it approved any further. In the event that changes need to be made that may affect regulatory requirements or safety or performance of the medical device, then that could possibly justify checking with regulatory... though, do you have an internal regulatory department that may be required to sign off on a CAPA?
Do you have in your internal procedures that regulatory must check / approve each CAPA? If not, then I would think the auditor is overstepping a bit.
Our CAPA procedure states that when the CAPA is related to the product (the medical device) the related action items and activities are going through the ECO (change control) process, which always includes a mandatory regulatory assessment.
Otherwise (non-product related CAPAs) do not need regulatory supervision. We think we cover the ISO requirement, but the auditor insists that every CAPA needs regulatory assessment. Don't want to change the procedure just because of the auditor's interpretation. Your advice?
Hi- Tnx! - please see my previous reply explaining the case.
yodon
Well, you basically have 2 choices: fight or flight.
You can press the auditor for clarification on what, exactly, the nonconformance is and then escalate the matter if you still feel it's unwarranted.
Or you can just loop Regulatory in on each CAPA. (Do you have so many CAPAs that it would be a burden for regulatory to approve each?)
I don't necessarily like the capitulation option as it only promotes bad auditor practices. But sometimes you pick your fights.
Tagin
medical device;
It all seems to hinge on the interpretation of 'verifying'. To me, 'verifying' leaves it up to the company as to how they satisfy that requirement. Indeed, it could be argued that you are using a two-stage verification, where stage one is:
"CAPA group classifies this CAPA as non-product related, and therefore is verification that applicable regulatory requirement are not adversely affected"
This could be strengthened if your regulatory department make some kind of policy statement, along the lines of:
"We have determined that non-product CAPAs do not require any additional verification review by our department to ensure that applicable regulatory requirement are not adversely affected."
Edit: Another way to state this is that the auditor is claiming that verification can only be performed by the regulatory department, whereas I am speculating that the regulatory department can delegate some of that review process to the CAPA group.
levatorsuperioris
I have a through on the possible issues: They are probably coming from a certain perspective. "Accept as is" NCR/CAPA result is not ok if it violates regulatory requirements ... this call is somewhat regulatory complex.
François
In the handbook of ISO 13485, the following is written in the intent :
verifying that a corrective action does not adversely affect the ability (of product, process or QMS) to meet applicable regulatory requirements, or
a medical device’s safety and performance.
So it seems that all CAPA must have a regulatory assessment (but not necesseraly by your RA team).
TWA - not the airline
I'm with Francois here. There may be types of CAPAs where you can generally determine beforehand that they never impact regulatory requirements, but these will be very few and special and the distinction product-related vs. not product-related will not be a satisfactory definition for this. If you add a signature field for RA on your CAPA form, then you're done, no further description/SOP needed. If you or your RA department wants to delegate then you need to put more into the CAPA SOP to show how you fulfil the requirement.
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 3
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
H ISO 13485:2016 Gap Analysis by NB ISO 13485:2016 - Medical Device Quality Management Systems 7
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
JoCam Difference between Approval and Registration - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
T Document control ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
L ISO 13485:2016 Clause 8.4 - Analysis of Audit Observations ISO 13485:2016 - Medical Device Quality Management Systems 8
S When is ISO 13485:2016 6.4.2 Contamination Control appropriate? ISO 13485:2016 - Medical Device Quality Management Systems 11
L Templates for three ISO 13485:2016 SOPs ISO 13485:2016 - Medical Device Quality Management Systems 8
C What falls under the 'Customer Property' according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 29
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1
F ISO 13485:2016 Quality Policy Requirements Other ISO and International Standards and European Regulations 18
M Contract Manufacturers and MDF Responsibilities, ISO 13485:2016, Clause 4.2.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
J ISO 13485:2016 sample exam/test ISO 13485:2016 - Medical Device Quality Management Systems 3
M Informational Questionário – Análise crítica sistemática – ISO 13485:2016 (Portuguese-only) Medical Device and FDA Regulations and Standards News 0
M Informational ISO 13485:2016 under systematic review Medical Device and FDA Regulations and Standards News 5
C Updates on Documentation for outsourced OEM from ISO 13485:2003 to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 4
GStough Informational SN EN ISO 9001:2015 and SN EN ISO 13485:2016 on Same Certificate? Registrars and Notified Bodies 7
E MDSAP Audit - Our QMS conforms to ISO 13485:2016 and FDA GMP Canada Medical Device Regulations 9
Ronen E Informational ISO 13485:2016 Transition Period End - 1 March 2019 ISO 13485:2016 - Medical Device Quality Management Systems 0
B ISO 9001:2015 vs ISO 13485:2016 for MDR Compliance EU Medical Device Regulations 4
T ISO 13485:2016 - Processes exempt from process validation ISO 13485:2016 - Medical Device Quality Management Systems 12
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
J EU ISO 13485:2016 Recertification Audit - Effect of 10 Minor Nonconformances EU Medical Device Regulations 2
E Template of a Management Review Agenda or Report in compliance with ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 6
J ISO 13485:2016 Section 6.2 - Documenting the process for establishing competence ISO 13485:2016 - Medical Device Quality Management Systems 6