ISO 13485:2016 - Processes exempt from process validation
ISO 13485:2016 defines criteria for processes which need to be validated. I want to ask that the processes which do not need to be validated, is there any requirement for equipment qualification for those processes, or IQ, OQ, PQ requirements? If yes, how to go about it as complete validation is not required.
Thanks in advance
Ronen E
Problem Solver
Moderator
In principle, processes don't require validation where the output is verified 100%. In such cases there is no concern about the process because it's effectiveness is checked directly on the output (the product). IMO, from a 13485 compliance perspective you should be fine in such cases with any extent and manner of qualification you choose to implement. Make sure, however, that you are still in compliance your own SOPs requirement-set.
Marcelo
I always think that there's some misconceptions when we talk about process validation.
First, we need to separate between before a process is put into service/operating - A (which related to the process planning) and after the process is put into service/operating - B.
Process validation requirements are application to B, not A - special processes are processes where I cannot verify that the output of the process, when the process is operating that, and thus I cannot guarantee conformity (please note that ISO 13485:2016 also includes the situation where you do not WANT to verify the output, thus transforming the process into a special processes.
In relation to A, any process needs verification. So, when you plan the process, you define process requirements, and, before putting the process into service/operating, you need to verify that it can proceed the expected results. In this way, you are verifying that the process can be put into service/operating.
I have a confusion about equipment qualification, which is a part of process validation. For process which does not need validation, is there a need for equipment qualification of equipment used in that process? If yes, please give some detail or reference document as to what extent equipment qualification would be sufficient?
somashekar
Equipment qualification aka IQ.
I understand that your process does not need validation, however there is an equipment.
The check for having provided all of the proper connections (inputs) to the equipment as suggested, including its proper fixing / suspending etc etc per recommendation.
The check of the satisfactory operation of the equipment. All indicators, meters operating satisfactorily, no abnormal vibrations / noise... (No error code / error signal visible or alarm heard)
The check of the desired output when operational, when you feed the correct inputs (materials)., including operation of any emergency stops etc.
All these done by qualified personnel and made as a form of a report .....
becomes your equipment qualification.
I hope such an exercise will give you enough confidence that the equipment is qualified for your intended use.
The check for having provided all of the proper inputs to the equipment as suggested, including its proper fixing / suspending etc etc per recommendation.
The check of the satisfactory operation of the equipment. (No error code / error signal visible or alarm heard)
The check of the desired output when operational, including operation of any emergency stops etc.
Yes that's what I was looking for.
Another question. I could not find a requirement for equipment qualification in ISO 13485, so it seems to be optional......?
somashekar
You please see how the equipment qualification fits into the ISO 13485:2016 clause 6.3 and in that the 6.3 b)
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
H ISO 13485:2016 Gap Analysis by NB ISO 13485:2016 - Medical Device Quality Management Systems 7
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
JoCam Difference between Approval and Registration - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
T Document control ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
L ISO 13485:2016 Clause 8.4 - Analysis of Audit Observations ISO 13485:2016 - Medical Device Quality Management Systems 8
S When is ISO 13485:2016 6.4.2 Contamination Control appropriate? ISO 13485:2016 - Medical Device Quality Management Systems 11
L Templates for three ISO 13485:2016 SOPs ISO 13485:2016 - Medical Device Quality Management Systems 8
C What falls under the 'Customer Property' according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 10
M Is it possible to get iso 13485:2016 certified as a one man band ISO 13485:2016 - Medical Device Quality Management Systems 1
F ISO 13485:2016 Quality Policy Requirements Other ISO and International Standards and European Regulations 18
M Contract Manufacturers and MDF Responsibilities, ISO 13485:2016, Clause 4.2.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
J ISO 13485:2016 sample exam/test ISO 13485:2016 - Medical Device Quality Management Systems 3
M Informational Questionário – Análise crítica sistemática – ISO 13485:2016 (Portuguese-only) Medical Device and FDA Regulations and Standards News 0
M Informational ISO 13485:2016 under systematic review Medical Device and FDA Regulations and Standards News 5
C Updates on Documentation for outsourced OEM from ISO 13485:2003 to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 4
GStough Informational SN EN ISO 9001:2015 and SN EN ISO 13485:2016 on Same Certificate? Registrars and Notified Bodies 7
E MDSAP Audit - Our QMS conforms to ISO 13485:2016 and FDA GMP Canada Medical Device Regulations 9
Ronen E Informational ISO 13485:2016 Transition Period End - 1 March 2019 ISO 13485:2016 - Medical Device Quality Management Systems 0
B ISO 9001:2015 vs ISO 13485:2016 for MDR Compliance EU Medical Device Regulations 4
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
J EU ISO 13485:2016 Recertification Audit - Effect of 10 Minor Nonconformances EU Medical Device Regulations 2
E Template of a Management Review Agenda or Report in compliance with ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 6
J ISO 13485:2016 Section 6.2 - Documenting the process for establishing competence ISO 13485:2016 - Medical Device Quality Management Systems 6
Q Any good Checklists for ensuring SOPs cover ISO 13485:2016 and 21CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
H Transition to ISO 13485:2016 together with ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 12
B Classes/ Online Training on ISO 13485:2016 and FDA QSR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 5
T Software Validation Certificate (ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 19
D ISO: 13485:2016 Sec. 7.5.2 (C) - Requirements for cleanliness of product or contamination control ISO 13485:2016 - Medical Device Quality Management Systems 2