ISO 13485 - 7.5.2 cleanliness , 6.4.2 validation of special processes, and 6.4.2 contamination ability to be exempt?

Jkc3usc12

Involved In Discussions
#1
Exempt from these 3 ISO 13485 clauses?

7.5.2 - My company is a distributer of a non-sterile product and have no required cleaning instructions besides a recommendation for the user.

6.4.2 - Returned product is received but any product that has been returned from the hospital is decontaminated there. The product distributed becoming contaminated has not happened.

7.5.6 - we do not design or manufacture any product.
 
Last edited by a moderator:
Elsmar Forum Sponsor

Rincewind

Involved In Discussions
#2
Hello,

7.5.2:
I do not believe you can exclude clause 7.5.2 from your QMS, it is stated:
"The organization shall document requirements for cleanliness of product or contamination control of
product if:
[...]
b) product is supplied non-sterile and is to be subjected to a cleaning process prior to sterilization or
its use; [...]", this reads to me like it fits your product pretty well from what you told us.

6.4.2
I just had a pretty similar discussion with an auditor the other day, the question is, how do you make sure the device is properly disinfected/sterilized when it gets back to you? Do you have a form the hospital has to fill out and send you back with the product? The clause clearly states "potentially contaminated product" as well so depending on your auditor or a risk analysis performed by your people you might need a procedure for this as well.

7.5.6
This does not only apply to manufacturing processes but to service provision as well, so if you can not claim that all your processes are verifiable you might need to validate some, some companies get around this by performing final tests on every product they manufacture, as you do not manufacture anything I do not know if this can be used by you.
 

Jkc3usc12

Involved In Discussions
#3
Agree with your points. Again being solely a distributor of the product I just have a hard time with some of the elements. We receive it in a box, check it, then rebox, and send. That is the main function of the facility.

6.4.2 could I possibly say we do not receive contaminated product. If we do it is scrapped or not accepted?

7.5.6 again we do not manufacture so all validation and testing of product is done at before we receive it.
 

Rincewind

Involved In Discussions
#4
6.4.2 could I possibly say we do not receive contaminated product. If we do it is scrapped or not accepted?
If that is what happens then that would be your procedure. If the device is possibly contaminated it is scrapped. A simple procedure could maybe be: You only accept products which have some kind of certificate from the customer that it was decontaminated before it was send out to you (I have seen companies do that, for example companies selling surgical instruments), if this certificate is missing you scrap the product (or clean it yourself before further processing but that might be not possible for you if you are just a distributor.)

7.5.6 again we do not manufacture so all validation and testing of product is done at before we receive it.
Again, this clause does not only apply to the manufacturing process but service provision as well. It depends on what your part in the cycle of this product is, I am thinking about 7.5.4 servicing activities for example (as you mentioned you receive products back). But if you have concluded non of your processes need to be validated you could exempt it.
 

Lokus200

Starting to get Involved
#5
Agree with your points. Again being solely a distributor of the product I just have a hard time with some of the elements. We receive it in a box, check it, then rebox, and send. That is the main function of the facility.

6.4.2 could I possibly say we do not receive contaminated product. If we do it is scrapped or not accepted?

7.5.6 again we do not manufacture so all validation and testing of product is done at before we receive it.
6.4.2 - you can add receiving inspection requirements to verify that it was decontaminated.
7.5.6 - if your scope is distribution only, you should document the justification for it in your Quality Manual. You are not a manufacturer so you don't have to validate manufacturing processes, however, you may have to validate your distribution process.
 

Jkc3usc12

Involved In Discussions
#6
Hello,

7.5.2:
I do not believe you can exclude clause 7.5.2 from your QMS, it is stated:
"The organization shall document requirements for cleanliness of product or contamination control of
product if:
[...]
b) product is supplied non-sterile and is to be subjected to a cleaning process prior to sterilization or
its use; [...]", this reads to me like it fits your product pretty well from what you told us.

6.4.2
I just had a pretty similar discussion with an auditor the other day, the question is, how do you make sure the device is properly disinfected/sterilized when it gets back to you? Do you have a form the hospital has to fill out and send you back with the product? The clause clearly states "potentially contaminated product" as well so depending on your auditor or a risk analysis performed by your people you might need a procedure for this as well.

7.5.6
This does not only apply to manufacturing processes but to service provision as well, so if you can not claim that all your processes are verifiable you might need to validate some, some companies get around this by performing final tests on every product they manufacture, as you do not manufacture anything I do not know if this can be used by you.
both 7.5.2 and 6.4.2 we did declare in our QM but was written a NC for not declaring the non applicable. This is what was written -
The Scope of the Quality Management System is not fully effectively described because not all non-applicable clauses have been declared
clauses 7.5.2 cleanliness, 7.5.6, validation of special processes, and 6.4.2 Contamination control, have not been declared.
 
Thread starter Similar threads Forum Replies Date
D ISO: 13485:2016 Sec. 7.5.2 (C) - Requirements for cleanliness of product or contamination control ISO 13485:2016 - Medical Device Quality Management Systems 2
J ISO 13485:2003 Section 7.5.1.2.1 Cleanliness of Product ISO 13485:2016 - Medical Device Quality Management Systems 1
C ISO 13485 Clause 7.5.1.2.1 Cleanliness of Product and Contamination Control Question ISO 13485:2016 - Medical Device Quality Management Systems 7
B ISO 13485 7.5.1.2.1 Cleanliness and Contamination of Product Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 5
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7
M Getting started in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 31

Similar threads

Top Bottom