ISO 13485 - Advisory Notices - Regulations - Quality Manual

Q

QA Tech

#1
ISO 13485- Advisory Notices/Regulations/Quality manual

We are an electronics manufacturing company that is ISO 9001:2000 and adding ISO 13485:2003 by January of 2005. ISO 14969 explains advisory notices to an extent, however it would be nice to see some examples. Anyone have a model for advisory notice process/procedure?

ISO 13485 constantly referrs to national or regional regulations and typically this is FDA QSR, but are there other regulating bodies for medical devices?

Any tips/suggestions on how to incorperate two quality systems (ISO 9001:2000 and ISO 13485:2003) in one quality manual, quality policy, and one set of procedures?

Thanks!
 
Elsmar Forum Sponsor
J

Jimmy Olson

#2
Hello and welcome to the cove :bigwave:

I am in the same situation as you. We are an electronic manufacturing company that is looking at adding 13485:2003 (so we can compare notes :) ) I have a copy of a manual based on the 1996 version of 13485 that is helpful, but I have to clean it up and remove names and references first. I also have some other material that I've gathered that compares the two that I can share, I am on my way out the door, so I will sort through all my stuff tomorrow (just thought I would say hi before taking off).

There are also other people here that have a very good understanding of the medical device field and are very helpful.
 
V

Vincnet

#3
QA Tech said:
ISO 13485 constantly referrs to national or regional regulations and typically this is FDA QSR, but are there other regulating bodies for medical devices?
Hi QA Tech,

As ISO is an Internationnal standard, regional regulations refers to the EU Directives in Eurpore 93/42/EC and 98/79/EC, regulated by the so called Competent Authorities, and to the Japanese regulation or any other regulations in the world.

Vincnet
 

Al Rosen

Staff member
Super Moderator
#4
QA Tech said:
Any tips/suggestions on how to incorperate two quality systems (ISO 9001:2000 and ISO 13485:2003) in one quality manual, quality policy, and one set of procedures?

Thanks!
QATech:

I wrote my manual using 9k2k, then I made a comparison clause by clause against 13485:2003(FDIS). Where they differed I modified it approriately to incorporate the additional requirements of 13485 and still meet the customer satisfaction requirement of 9001. I haven't been assessed yet to 13485:2003, but I have been assesed and registered to 13485:1996 and CMDCAS. I haven't had any problems.

BTW, you will need more than 6 documented procedures.
 
Last edited:
C

Cathy

#5
Hi QA Tech - don't know if you have found an anwser to your advisory notice question. Here's my input

Advisory notices can be issued from the country's competent authorities or by the manufacturer to the competent authority.

When you (as the manufacturer) issue an advisory notices you need to include detailed information on the following:
1. the type of incident (or potential incident)
2. whether any medical device has been involved that the manufacturer is responsible for
3. whether the incident was caused, wholly or partly, by the medical device
4. if there is any short comings in the information supplier with the product.

To see exmaples of advisory notices check out FDA website or try this link though it is for UK competent authority.
http://www.medical-devices.gov.uk/mda/mdawebsitev2.nsf/webvwMDASafetyWarnings?Open&count=9999

One thing I will say about the transition from manufacturing to manufacturing medical devices is you need to be very careful in understanding the medical device vigilance. This is built into ISO13485:2003 to some extent but consult the FDA regulations on this area.
 
K
#6
Advisory Notices

I work for a contract manufacturer that is persuing ISO13485, and builds several different finished devices. I am still confused about Advisory Notices.
Would our customers (i.e., the corporations marketing the devices) be responsible for dealing with them?

(We have never had to deal with adverse event reporting; it is handled by our customer).
 
#7
Kari said:
I work for a contract manufacturer that is persuing ISO13485, and builds several different finished devices. I am still confused about Advisory Notices.
Would our customers (i.e., the corporations marketing the devices) be responsible for dealing with them?

(We have never had to deal with adverse event reporting; it is handled by our customer).
I would think that advisory notices would be those issued by your company, not your customer. (See Cathy's post). Advisory notices are related to the work your company does.
 

Al Rosen

Staff member
Super Moderator
#8
Kari,

Advisory notices are not adverse event reports to the competent authority, they are notices to your customers.

I believe it is the responsibility of the entity that places the device on the market to issue the advisory notices to their customers. But, if you are a contract manufacturer and are to be registered to ISO 13485, it is necesary to address the issue of advisory notices to your customer in accordance with 7.3(d) and 8.5.1.
 

amjadrana

Involved - Posts
#9
Advisory Notices

advisory notices are issued by the manufacturer to the customers in case a device is not functioning as intended. Advisories can be issued by the regualtory bodies as well. For example Health canada posts advisories on its website for medical professionals as well as for general public.

I company should have documented procedures for issuing and implementing advisory notices to customers, regulatory bodies or to general public.

Amjad Rana
 
B

Bridget

#10
Advisory Notice

I have both a procedure and a form that has been through several auditors without any findings. I kept it simple and my form is not sent to our customers. They would be contacted via phone, email, fax, etc more or a personal touch. This form identifies the steps needed to record the problem as well as who to contact if regulatory reporting is needed. Therefore if I am not available anyone can start the recall. The procedure is very simple and echos the form, in fact I made the form first.

If further reporting is needed I have forms for the Competent Authorities which can be used for Canada as well as the FDA.

If it helps at all I am glad I posted it. Of course if anyone sees any gaps in it I welcome the feedback.
Bridget
 

Attachments

Thread starter Similar threads Forum Replies Date
S Difference between an Advisory Notice (ISO 13485) and a Field Safety Notice? ISO 13485:2016 - Medical Device Quality Management Systems 3
R ISO 13485 - Advisory Notice Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
A Advisory Notices - ISO 13485 - applicable to suppliers? ISO 13485:2016 - Medical Device Quality Management Systems 20
C Established procedure for Advisory Notices - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
G Advisory Notices - What is the Requirement in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
D Implementing ISO 13485:2003 - 8.5.1 - Advisory Notices - Recall Procedure ISO 13485:2016 - Medical Device Quality Management Systems 12
D Notified Bodies - ISO 13485 & MDR Technical Files ISO 13485:2016 - Medical Device Quality Management Systems 3
D Deviations - Where in ISO 13485 deviations are covered? ISO 13485:2016 - Medical Device Quality Management Systems 7
B ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 2
Y ISO 13485:2015 Software Validation IQ/OQ/PQ ISO 13485:2016 - Medical Device Quality Management Systems 12
J ISO 13485 for Metal Finishing Medical Device and FDA Regulations and Standards News 5
S How to calculate Effective Number of People for ISO 13485 Certification? General Auditing Discussions 2
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
E Any template/ form of Monitoring and Measurement of Processes and product to ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 1
H ISO 13485-paragraphs for a SaaS SAMD needed or not? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
M ISO 13485 and document control for programs ISO 13485:2016 - Medical Device Quality Management Systems 6
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
D Lead time to schedule an ISO 13485 audit General Auditing Discussions 2
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 1
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 3
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3

Similar threads

Top Bottom