Dear experts,
For ISO 13485 requirement, do you know if the suppliers of equipment calibration must be evaluated and listed in "approved vendor list" in accordance with clause 7.4.1? As I know, just material/component and subcontractor need to do this.
But the auditor of NB said "Yes" when he audited for ISO 13485.
Thank you for kindly comment.
 
Elsmar Forum Sponsor

mdurivage

Quite Involved in Discussions
#2

The short answer is yes.

7.4.1 Purchasing process
The type and extent of control applied to the supplier and the purchased product shall be dependent upon theeffect of the purchased product on subsequent product realization or the final product.

and

Sec. 820.50 Purchasing controls
Each manufacturer shall establish and maintain procedures to ensure that all purchased or otherwise received product and services conform to specified requirements.
Anything that is purchased that has a quality system impact (goods, services, consultants) must be on the ASL. The requirements to place the supplier on the list and the controls to maintain them should be risk based.



 

medwise

Involved In Discussions
#3
In simple terms, if any step or process which is outsourced and can impact the quality of the final product should be evaluated and monitored. For e.g. if you are sterilising the product and if your thermocouple is not calibrated then your sterilisation process may not be valid. Other e.g. if your pressure gauge (magnehelic) which is used to monitor the pressure differential of your clean room area is not calibrated then how will you know that your clean room is clean. Therefore, as a manufacturer you are required to identify your critical supplier and monitor them periodically.

P.S - Not an expert though an obedient student of ISO 13485:2003.:tg:
 
Thread starter Similar threads Forum Replies Date
D Sample Calibration SOP for ISO 13485 General Measurement Device and Calibration Topics 4
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
H Is there ISO 13485 Guidance for Calibration Tolerance? General Measurement Device and Calibration Topics 14
M Do basic Microscopes require Calibration per FDA and ISO 13485 General Measurement Device and Calibration Topics 10
M "As Found" (aka "As Received") Calibration Data - ISO 13485 Requirements ISO 13485:2016 - Medical Device Quality Management Systems 8
somashekar Calibration interval (Clause 7.6 ISO 13485:2003) - Biomedical test instruments ISO 13485:2016 - Medical Device Quality Management Systems 4
J Calibration service requirements for device manufacturers - ISO 13485 requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
G Limited Calibration requirements in ISO 13485 - Temperature controller ISO 13485:2016 - Medical Device Quality Management Systems 3
B ISO 13485 Calibration Requirements - Requiring a Documented Procedure for control? ISO 13485:2016 - Medical Device Quality Management Systems 8
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 5
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3

Similar threads

Top Bottom