ISO 13485 Class I Medical Device Design Exclusion

B

Bomiss

#1
Hello All,

Thanks to everyone for participating on this form. The participation allows the novice (like me) to gain a better understanding through the discussions. I have been reading the content for a few weeks and I am impressed, not only about the knowledge content, but also the considerate approach to finding resolution to the questions proposed on the forum.

Now for the issue I have at hand, I would appreciate a better understanding of the Design requirements for ISO 13485. I have recently started a new business and I need validation on my perception of Design Exclusions in regard to ISO 13485. The FDA’s GMP CFR 820 standard states that Class I and some Class II medical devices are excluded from Design Controls. The product I manufacture is a class I, non-invasive, non-sterile, plastic/rubber medical device used in Dentistry. The product category is not listed in the chart of Class I Medical devices that have to have Design Controls.

It is my interpretation, that even if I designed the product, I would not need to use design Controls in my business. At this time, we are a very small company and all Design, Engineering, Quality, and other business functions are managed by one person.

If I am misled please help me understand.
Thanks
 
Elsmar Forum Sponsor

RA Guy

Involved In Discussions
#2
I believe your assertions are basically correct, but will put them in my own words for clarity.

For FDA purposes, it is explicity prescribed which devices require design controls/gmp and which devices do not require. Having said that, even for simple, low risk devices there are valid reasons, beyond simple compliance, to have design documentation and a project management approach as implied by design controls.

The second paragraph of 13485 Section 1.2 defines the criteria for exclusions under sec. 7.3. Essentially it is allowable only if permitted by the applicable regulator in the country whee you will maket the device.

For instance in Canada regarding device licencing requirements:
Class I - ISO13485 not required
Class II - ISO13485 CMDCAS required for manufacturing
Class III & IV ISO13485 CMDCAS required for design and manufacturing

That means for a class II device in Canada, design is an allowable exclusion from the application of 13485 CMDCAS.
 
B

Bomiss

#3
Thanks for the quick reply. I appreaciate you shaing your knowledge and understanding of the requirement.
 
M

MIREGMGR

#4
Generally speaking, Europe requires that all medical devices conform to the Medical Device Directive, and in particular to the Essential Requirements, Annex I. The Essential Requirements are normally interpreted to require design controls, and do not exclude any device classes.
 
B

Bomiss

#5
MIREGMGR,

Thanks for the reply. For clarification, I should not have an issue with ISO 13485, however, when I plan on selling product in the EU, I will need to satisfy the requirements of Design Control. Is that your understanding?

Thanks
 
M

MIREGMGR

#6
The latter part is correct.

Regarding implementation of ISO 13485, standards are applied in a context, as RAGuy illustrates above. They aren't meaningful without knowing that context, since it determines their applicability and interpretation. Who will be determining your conformance to ISO 13485, and with regard to what jurisdiction's requirements?
 

Ronen E

Problem Solver
Staff member
Moderator
#7
The Essential Requirements are normally interpreted to require design controls, and do not exclude any device classes.
This is quite a broad statement that doesn't match my knowledge and/or experience; I'd be glad if you could kindly provide a pointer to a regulation, guidance or otherwise, that support this approach. I can imagine some of the stricter NBs applying such an approach, but definitely not all of them for all device classes. In the EC, plain class I (non-S non-M) devices do not require a qualified QMS, let alone design controls; a NB intervention would also not be required, so I wonder who would then require the implementation of design controls.
 
M

MIREGMGR

#8
That's our reading of the MDD, and my understanding of how our competitors see it as well, from my contacts with them. Some applicability of the MDD is legally self-controlled, of course, rather than being imposed via NB enforcement. My understanding however is that our NB and our Authorized Representative have the same interpretation of the MDD, as communicated during their various oversight of our tech files, processes and listings for the many Class I and I-s devices we make.

Perhaps as you suggest our NB and Authorized Representative are more rigorous than some others.

As a possibly relevant data point, we know that US FDA and the EU authorities among others have been discussing what evolutions of the various systems may occur as those systems are eventually moved toward harmonization. My understanding is that US FDA's thinking is that their current exclusion of design control for (mostly) Class I devices will be one of the things they'll sacrifice in that harmonization process, moving instead to (as I understand it) an MDD-like expectation of design control for all devices.
 

Ronen E

Problem Solver
Staff member
Moderator
#9
That's our reading of the MDD, and my understanding of how our competitors see it as well, from my contacts with them. Some applicability of the MDD is legally self-controlled, of course, rather than being imposed via NB enforcement. My understanding however is that our NB and our Authorized Representative have the same interpretation of the MDD, as communicated during their various oversight of our tech files, processes and listings for the many Class I and I-s devices we make.

Perhaps as you suggest our NB and Authorized Representative are more rigorous than some others.

As a possibly relevant data point, we know that US FDA and the EU authorities among others have been discussing what evolutions of the various systems may occur as those systems are eventually moved toward harmonization. My understanding is that US FDA's thinking is that their current exclusion of design control for (mostly) Class I devices will be one of the things they'll sacrifice in that harmonization process, moving instead to (as I understand it) an MDD-like expectation of design control for all devices.
Once again, so we don't confuse newcomers, there's no explicit MDD expectation of design controls for all devices. To me, yours seems to be an interpretation that exists in a closed set of NB(s?) and companies. The authorised representative may provide advice, but is certainly not in a legal position to require anything of that sort (unless such power is granted to them in a specific contract). I don't know what the FDA intends to do with regards to design controls in the future, my comment related to the state of things in the EC, as I know them today.

Perhaps your NB promotes implementation of design controls for your class I devices because:
1. It is a good practice and makes a lot of sense anyway.
2. it is an easy/simple way (not the only one) of complying with the ERs.
3. It makes a lot of sense to employ a single set of principles / system for all devices you make, regardless of class, and since some of your devices are higher class than I(?), you have to have design controls established in your organisation anyway.

Either way, I see it more as a strong recommendation than as a mandatory requirement.
 
M

MIREGMGR

#10
Once again, so we don't confuse newcomers, there's no explicit MDD expectation of design controls for all devices.
The Essential Requirements state in a number of locations, "The devices must be designed (...)", followed by a specification for particular required performance or other characteristics. It is at least a common interpretation of this set of stated requirements that the design process, just as the manufacturing and packing processes in certain cases, must be controlled so as to accomplish the stated requirement...and of course, as a general requirement in line with the universal dictum that "if it's not documented it didn't happen", that design process must be documented.

I gather that you're arguing for an alternate interpretation of the "must be designed, manufactured and packed" requirements, in which there isn't any requirement for control of processes as long as the device outcome is acceptable. I don't think the rather extensive historical record of drafter intent supports an interpretation that doesn't include control of processes, though. Certainly that's the understanding that's expected by the regulatory parties with which my employer interacts.
 
Thread starter Similar threads Forum Replies Date
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
C ISO 13485 Requirement in Australia/NZ for class 1? ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
F MDR ISO 13485 for Class I sw/hw company to be class II due dates ISO 13485:2016 - Medical Device Quality Management Systems 3
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
M ISO 13485 Required for CE Mark on Class I product CE Marking (Conformité Européene) / CB Scheme 1
B ISO 13485 certification advantages when I certify a Class 2a medical device CE Marking (Conformité Européene) / CB Scheme 12
K ISO 13485:2016 Self Certification for Class I (annex 9) Stand-Alone Software? ISO 13485:2016 - Medical Device Quality Management Systems 3
GoSpeedRacer Pre-Work and Study for ISO 13485:2016 Lead Auditor Class ISO 13485:2016 - Medical Device Quality Management Systems 1
K ISO 13485 requirements for Stand Alone Medical Device Software (Class II a) ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485 and Class 1 Medical Device - FDA thinking ISO 13485:2016 - Medical Device Quality Management Systems 14
Q Class I Technical File reviewed during ISO 13485 Certification Audit by NB CE Marking (Conformité Européene) / CB Scheme 10
N Extent of Product Traceability Required by ISO 13485 (Low Level Class I Devices) ISO 13485:2016 - Medical Device Quality Management Systems 3
Q ISO 13485 Contract Manufacturer & Class II Devices Other US Medical Device Regulations 1
C ISO 13485 Class 7 Cleanroom Requirements ISO 13485:2016 - Medical Device Quality Management Systems 7
T ISO 13485 Book and Training Class Recommendations wanted Training - Internal, External, Online and Distance Learning 4
S Medical Device Class IIa Nonsterile ISO 13485 Alternatives ISO 13485:2016 - Medical Device Quality Management Systems 3
B Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope? ISO 13485:2016 - Medical Device Quality Management Systems 2
A ISO 13485 Risk Analysis - Class II Medical Device Engineering Changes ISO 13485:2016 - Medical Device Quality Management Systems 9
I Distributors for class II medical devices in US - Do we need ISO 9001 or ISO 13485 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C CAP for Medical Device Class Is product - Slightly confused - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 3
O ISO 13485 sample document or template - Class II medical device ISO 13485:2016 - Medical Device Quality Management Systems 16
Q Training Matrix - Job Class vs. Process Instructions - ISO 13485 Training - Internal, External, Online and Distance Learning 4
D ISO 13485 Certification - Class II product - Fluoride-free toothpaste - Records ISO 13485:2016 - Medical Device Quality Management Systems 1
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1

Similar threads

Top Bottom