ISO 13485 Clause 7.5.1.2.1 Cleanliness of Product and Contamination Control Question
cartman1417
Background: Our device is a Class IIa diagnostic medical device. It is a cart based system for vaginal examination, and uses a transvaginal probe for examination (similar to an ultrasound probe). The system consists of a touch screen monitor on medical cart and other parts/components.
The probe is covered with a disposable sheath/cover and hence the probe is reusable but might need to be "disinfected" (not sterilized) at periodic intervals, by soaking in chemicals. The touch-screen monitor will either be covered with a disposable cover or be disinfected by disinfection wipes.
So how do I go about writing the content of the clause 7.5.1.2.1? Could anyone please suggest?
Reference:
7.5.1.2.1 Cleanliness of product and contamination control
The organization shall establish documented requirements for cleanliness of product if
a) product is cleaned by the organization prior to sterilization and/or its use, or
b) product is supplied non-sterile to be subjected to a cleaning process prior to sterilization and/or its use, or
c) product is supplied to be used non-sterile and its cleanliness is of significance in use, or
d) process agents are to be removed from product during manufacture.
If product is cleaned in accordance with a) or b) above, the requirements contained in 6.4 a) and 6.4 b) do not
apply prior to the cleaning process.
Stijloor
Re: ISO 13485 Clause 7.5.1.2.1 Cleanliness of Product and Contamination Control Quest
A Quick Bump!
Can someone help?
Thank you very much!!
Stijloor.
DannyK
The procedure should define the product cleanliness requirements, use of process agents, and decontamination processes if it applies.
somashekar
Reference:
Most likely your product or precisely the transvaginal probe for examination could fall under the highlighted in blue category as in above.
So have a documented procedure how the factory does this "The disinfecting (not sterilizing) before the final packing, by soaking in chemicals, and how these chemicals are further removed, making the transvaginal probe ready to use by the technician for the first time."
The rest of your product can be cleaned by whatever method that you deem fit and need not be covered within your procedure
cartman1417
How do I decide whether b) applies or c) ?
b) product is supplied non-sterile to be subjected to a cleaning process prior to its use, or
somashekar
b) As you say, the device is supplied non-sterile. So are you supplying it in a way ready to use, OR do you say within your device information for use that the device has to be cleaned before the first use in a specified way ?
c) Does your device risk management address the significance of cleanliness in its use ? towards this, do you supply the device with necessary cleaning ? OR
Do you recommend the cleaning method for the cleaning of the device before the first use to the technician ?
I am sure you have addressed this in the risk management, since you have mentioned above in post #1 this:
cartman1417
Our risk management does address the cleaning issue. From your questions, it seems to me that c) applies in our case.
We would be the suppliers of the probe covers as well. The probe needs to be disinfected if there is apparent rupture in the cover during/after examination or once every week otherwise. We will provide specific instructions on which chemicals to use and the procedure.
Ronen E
Problem Solver
Moderator
Those are not exclusive-or's (xor's). One could have any combination of a, b, c and/or d apply in a specific instance.
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
A ISO 13485 - Clause 8.3.3 appropriate actions in response to nonconforming product ISO 13485:2016 - Medical Device Quality Management Systems 2
K ISO 13485:2016, Clause 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 4
F ISO 13485 8.2.3 Reporting to regulatory authorities: Question regarding a procedure for this clause. ISO 13485:2016 - Medical Device Quality Management Systems 4
L ISO 13485:2016 Clause 8.4 - Analysis of Audit Observations ISO 13485:2016 - Medical Device Quality Management Systems 8
M Contract Manufacturers and MDF Responsibilities, ISO 13485:2016, Clause 4.2.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
JoshuaFroud Interpretation of Clause 5.5.2 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
S What is the clause in ISO 13485 for SAP Software Validation? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Damaged Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 7
GoSpeedRacer ISO 13485:2016 Clause 4.2.5 - Control of Records - HIPAA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 11
T Struggling with ISO 13485:2016 Clause 6.2 - Level of Risk Associated with Work ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 13485:2016 Clause 7.4.2 regarding Supplier Quality Agreements ISO 13485:2016 - Medical Device Quality Management Systems 11
R ISO 13485:2016 Clause 6.2 - Effectiveness Check ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 Clause 7.3.2 - Purchased Data Approval and Record Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S Interpreting clause 4.1.5 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 9
S Software Validation – Clause 4.1.6 of ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 12
P Which ISO 13485 clause(es) are applicable for Value Proposition Creation ISO 13485:2016 - Medical Device Quality Management Systems 1
J Definition Servicing - Definition - ISO 13485 Clause 7.5.1.2.3 - Servicing Activities Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
T ISO 13485 Clause 4.2.1 - Clarification of Intent ISO 13485:2016 - Medical Device Quality Management Systems 6
R ISO 13485 Clause 4.2.1 - Documentation Requirement question ISO 13485:2016 - Medical Device Quality Management Systems 4
D Taking an Exception to Clause 7.3 in ISO 9001 and ISO 13485 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
X Help with ISO 13485 Clause 8.3 "Rework" as applicable to Medical Device Servicing ISO 13485:2016 - Medical Device Quality Management Systems 39
X Help with ISO 13485 Clause 6.4 as it applies to Medical Device Servicing ISO 13485:2016 - Medical Device Quality Management Systems 25
Project Man The "Ifs" in ISO 13485 Clause 6.4 Work Environment - Machine Shop which makes parts ISO 13485:2016 - Medical Device Quality Management Systems 1
A Human Resources (ISO 13485 Clause 6.2) example of SOP ISO 13485:2016 - Medical Device Quality Management Systems 1
T Quality Staff - Independence (ISO 13485 Clause 5.5.1) ISO 13485:2016 - Medical Device Quality Management Systems 1
P Control of Monitoring and Measurement Devices - ISO 13485 Clause 7.6 says ... ISO 13485:2016 - Medical Device Quality Management Systems 3
A Purchasing Information Requirements - ISO 13485 Clause 7.4.2 and FDA QSR Supplier Quality Assurance and other Supplier Issues 2
D Quality System Manual - Exclusions from ISO 13485 Clause 7.5 ISO 13485:2016 - Medical Device Quality Management Systems 4
N Validation of Special Processes - N/A of ISO 13485 Clause 7.5.2 ISO 13485:2016 - Medical Device Quality Management Systems 12
I Not having Training Records for Documentation Changes - ISO 13485 Clause 4.2.3 Document Control Systems, Procedures, Forms and Templates 22
J ISO 13485 Clause 1.2 Application Clarification ISO 13485:2016 - Medical Device Quality Management Systems 1
K ISO 13485 Clause 7.4 Purchase Order Requirements Question ISO 13485:2016 - Medical Device Quality Management Systems 4
somashekar Calibration interval (Clause 7.6 ISO 13485:2003) - Biomedical test instruments ISO 13485:2016 - Medical Device Quality Management Systems 4
R ISO 13485 Clause 7.6 Control of Monitoring and measuring devices - Requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
Q Interpreting Responsibility and Authority Clause 5.5.1 ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 18
Q ISO 13485 Clause 4.2.2 Documentation Structure - Quality Manual clarification ISO 13485:2016 - Medical Device Quality Management Systems 6
R Interpretation of ISO 13485 Clause 8.4 - Analysis of Data ISO 13485:2016 - Medical Device Quality Management Systems 20
J ISO 13485:2003 Clause 1.3 - Applicable Standards and Regulations ISO 13485:2016 - Medical Device Quality Management Systems 8
N Poor control of testing chemicals ? which ISO 13485/9001 clause does it contravene? ISO 13485:2016 - Medical Device Quality Management Systems 8
J How to accomplish ISO 13485 Clause 8.2.3 Monitoring and Measuring? ISO 13485:2016 - Medical Device Quality Management Systems 5
F ISO 13485 Clause 6.3 - System to determine and provide infrastructure ISO 13485:2016 - Medical Device Quality Management Systems 5
V Planning of Product Realization - Risk Management - ISO 13485 Clause 7.1 ISO 13485:2016 - Medical Device Quality Management Systems 6
D Effective Control of Processes - ISO 13485 Clause 4.2.1 (d) ISO 13485:2016 - Medical Device Quality Management Systems 2
A ISO 13485 Clause 8.2.2 - Internal auditor from within the company to be audited Internal Auditing 2
L ISO 13485: 2003 - Clause 8.4 Analysis of Data - Seeking an example procedure ISO 13485:2016 - Medical Device Quality Management Systems 2
H New Clause Structure of ISO 13485:2003 ISO 13485:2016 - Medical Device Quality Management Systems 5