ISO 13485 Definitions - Medical Device Component Supplier
S
sbio1
#1
Need help. A chemical supplier for a medical diagnostics company (Class I medical devices) received ISO 13485 certification with the scope of development, manufacture and marketing a certain product group for diagnostics, i.e., critical biotech. reagent, e.g., enzyme. Is the reagent, e.g. enzyme, a "medical device" or just a "product" by definition within the framework of the ISO 13485 quality management system??? within the framework of FDA QSR? within the framework of EU IVDD?
Elsmar Forum Sponsor
Sidney Vianna
Post Responsibly
Staff member
Admin
#2
Based on the ISO 13485:2016 definition of medical device, it seems to me the product you described qualifies as one.
medical device
instrument, apparatus, implement, machine, appliance, implant, reagent for in vitro use, software, material or other similar or related article, intended by the manufacturer to be used, alone or in combination, for human beings, for one or more of the specific medical purpose(s) of:
— diagnosis, prevention, monitoring, treatment or alleviation of disease;
— diagnosis, monitoring, treatment, alleviation of or compensation for an injury;
— investigation, replacement, modification, or support of the anatomy or of a physiological process;
— supporting or sustaining life;
— control of conception;
— disinfection of medical devices;
— providing information by means of in vitro examination of specimens derived from the human body;
and does not achieve its primary intended action by pharmacological, immunological or metabolic means, in or on the human body, but which may be assisted in its intended function by such means
S
sbio1
#3
GHTF/SG1/N29:2005 (forward ref. from ISO 13485 clause 3.11)
5.2 In Vitro Diagnostic (IVD) Medical Device
‘In Vitro Diagnostic (IVD) medical device’ means a medical device, whether used
alone or in combination, intended by the manufacturer for the in-vitro examination of
specimens derived from the human body solely or principally to provide information for
diagnostic, monitoring or compatibility purposes.
Note 1: IVD medical devices include reagents, calibrators, control materials, specimen
receptacles, software, and related instruments or apparatus or other articles and are
used, for example, for the following test purposes: diagnosis, aid to diagnosis,
screening, monitoring, predisposition, prognosis, prediction, determination of
physiological status.
Note2: In some jurisdictions, certain IVD medical devices may be covered by other
regulations
Still in doubt... Let's say the finally assembled IVD test system is just one tube with a solution (i.e., clearly a "medical device" and "reagent). This solution is a simple mixture of three different solutions supplied by chem. company that is only ISO 13485 certified. Are those 2-tier reagents also "medical devices"?:
1. no intent to provide any diagnostic information? just an intent to supply "a part" for final assembly.
2. legal liability concerns, e.g., a supplier is in EU and does not have EC mark, just ISO 13485, and therefore cannot name the products a "medical devices"?
S
Sarah Stec
#4
Welcome!:bigwave:
Whether a product is a medical device or not depends on whether the product meets the regulatory definition, not on the status of the manufacturer/supplier of the product. So, the manufacturer/supplier may have no certifications at all and still have a medical device. Likewise, just because you have a medical device may not mean that you are a manufacturer within the regulatory definition.
Is your company putting this on the market? Or are you shipping it to someone else for them to put on the market?
The whole test system can be the IVD medical device if they are intended to be sold together. An IVD is a tube (the vitro in in vitro diagnostic) and a solution.
The question then becomes who is doing the intending: the chemical company doing the supplying or the manufacturer putting the IVD (tube + reagent) on the market? Is the chemical company "just intending to supply a part for final assembly"...for the purpose of diagnosis (or another IVD purpose)? Or is the chemical company selling a general purpose chemical that the manufacturer happens to use for a diagnostic purpose?
Tl;dr: You don't have to be a "manufacturer of IVD medical devices" to have an IVD medical device. :bigwave:
And you may want to stick to the legal documents (so, not the GHTF, but the FDA regs and the IVDD) when you're figuring out your legal liability. :2cents:
S
sbio1
#5
The chemical company: - is not putting the IVD (tube + reagent) on the market
- is manufacturing and supplying 3x different bulk chemicals to the customer in US
- the QMS is ISO 13485 certified; the chemicals are mentioned in the issued certificate
- is accepted supplier; passed on-site audits conducted by the customer
- declares intent that these bulk chemicals can be used for manufacturing the IVD
- does not formally label the bulk chemicals "for in vitro diagnostics" or with "EC Mark"
The customer: - is manufacturing the IVD + reagent from the 3x different bulk chemicals received from The chemical company
- meets all applicable FDA and IVDD regs
- is putting the IVD (tube + reagent; Class I device acc. to FDA regs classification) on the market in US and EU
All the questions concern The chemical company and the 3x bulk chemicals:
1. From perspective of FDA each of the 3x bulk reagents is medical device? Yes/No?
2. Within framework of the ISO 13485 (e.g, from ISO 13485 auditor perspective) each of the 3x bulk chemicals is medical device? Yes/No?
3. From perspective of EU regs each of the 3x bulk chemicals is medical device? Yes/No?
4. Is the company allowed to use in marketing materials the term "medical device" when referring to those bulk chemicals?
1. No; 2. Yes; 3. No; 4. Yes
Thread starter Similar threads Forum Replies Date
somashekar Definition Exclusion vs. Non-Application - Definitions in ISO 13485 Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
Q How to find definitions from ISO 9001 when implementing ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 8
R Definition MDR, MDVR, MPR - Definitions - ISO 13485 - Medical Devices Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 11
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 5
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 6
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7
M Getting started in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 31
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
Similar threads
Top Bottom