ISO 13485 for Class 1 Medical Device

#1
Are the certification requirements the same for Class 1 and Class 2 medical devices? In other words is the same level of system robustness, including records the same.
To achieve certification is there a requirement to complete a full generation of all records retrospectively for every product?
I want to map out resource requirements
 
Elsmar Forum Sponsor

DannyK

Trusted Information Resource
#2
Please specify the jurisdiction that you want to distribute products so that we could provide you with a specific answer.
 

yodon

Staff member
Super Moderator
#4
Short answer is no, the burden for Class I devices is (generally) lower. For example, in the US, Class I devices *may* be exempt from design controls. You'll have to consider specifics, though, for a precise answer.

To achieve certification is there a requirement to complete a full generation of all records retrospectively for every product?
Not completely sure I understand what you're asking here. When there is a regulatory requirement you need to meet, you do have to provide evidence of meeting it. Generally, this is done at the product level; e.g., Verification and Validation records for the product, not for every device built of that product. There are records required from production so in that respect, there would be records for every item built. Maybe clarify the question?
 
#5
Are the certification requirements the same for Class 1 and Class 2 medical devices? In other words is the same level of system robustness, including records the same.
To achieve certification is there a requirement to complete a full generation of all records retrospectively for every product?
I want to map out resource requirements
Are you referring to ISO 13485 QMS certificate? Or CE CERTICATE so on and so forth?
your question is not clear enough.
 
#6
Certification to the standard requires an organization’s quality management system to pass a third-party Medical Device Single Audit Program, or “MDSAP” Audit . Learn about the standard. Perform Gap analysis of your current QMS in line of the ISO13485. Then create your project plan and trained your staff about the standards and check regularly the results. Design and document your ISO 13485 Quality Manual and Procedures. The biggest portion of the project is looking at your current processes and redesigning them to address all of the requirements of the standard. Once you have modified or developed processes to meet the standard, you will need to control those processes.

Once your system is developed and documented, employees will follow the procedures, collect records, and make improvements to the system. For approximately three months or more, your organization will run the QMS, collecting records. To achieve registration, you will need to select a Registrar who will perform a Registration Audit.
 
#7
Thanks for the reply. Yes we are taking this approach. I’m struggling with the level of records/proof we need for certification. For example.. prior to the audit, would you expect to see a record of competency for all employees based on qualifications and experience along with a training plan linked to all relevant procedures and the associated records for training to every relevant document. This would apply to all employees involved in activities throughout the lifecycle with their competency assessment and training proportionate to the risk of their tasks? I am trying to get an idea of the work involved to fill the gaps. Another example... would you expect a retrospective review of all complaints with a documented assessment to determine if the risk file should be updated or a design update. Thank you
 

RAQA Support

Supporting your goals and objectives since 1989.
#8
Thanks for the reply. Yes we are taking this approach. I’m struggling with the level of records/proof we need for certification. For example.. prior to the audit, would you expect to see a record of competency for all employees based on qualifications and experience along with a training plan linked to all relevant procedures and the associated records for training to every relevant document. This would apply to all employees involved in activities throughout the lifecycle with their competency assessment and training proportionate to the risk of their tasks? I am trying to get an idea of the work involved to fill the gaps. Another example... would you expect a retrospective review of all complaints with a documented assessment to determine if the risk file should be updated or a design update. Thank you
You seem to have two main questions here:
1. Yes the training files need to be as thorough as you are stating. Training files are a common audit finding so do not be afraid to make them thorough when setting them up.
2. A retrospective review would not be expected. If you have any open items, those would need to be captured in the new system. And I would recommend if you have any high trending complaints to mention it in your first management review meeting. This will capture the past major problems without dragging all the previous documentation into the process.
 
Thread starter Similar threads Forum Replies Date
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
C ISO 13485 Requirement in Australia/NZ for class 1? ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
F MDR ISO 13485 for Class I sw/hw company to be class II due dates ISO 13485:2016 - Medical Device Quality Management Systems 3
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
M ISO 13485 Required for CE Mark on Class I product CE Marking (Conformité Européene) / CB Scheme 1
B ISO 13485 certification advantages when I certify a Class 2a medical device CE Marking (Conformité Européene) / CB Scheme 12
K ISO 13485:2016 Self Certification for Class I (annex 9) Stand-Alone Software? ISO 13485:2016 - Medical Device Quality Management Systems 3
GoSpeedRacer Pre-Work and Study for ISO 13485:2016 Lead Auditor Class ISO 13485:2016 - Medical Device Quality Management Systems 1
K ISO 13485 requirements for Stand Alone Medical Device Software (Class II a) ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485 and Class 1 Medical Device - FDA thinking ISO 13485:2016 - Medical Device Quality Management Systems 14
Q Class I Technical File reviewed during ISO 13485 Certification Audit by NB CE Marking (Conformité Européene) / CB Scheme 10
N Extent of Product Traceability Required by ISO 13485 (Low Level Class I Devices) ISO 13485:2016 - Medical Device Quality Management Systems 3
Q ISO 13485 Contract Manufacturer & Class II Devices Other US Medical Device Regulations 1
C ISO 13485 Class 7 Cleanroom Requirements ISO 13485:2016 - Medical Device Quality Management Systems 7
B ISO 13485 Class I Medical Device Design Exclusion ISO 13485:2016 - Medical Device Quality Management Systems 10
T ISO 13485 Book and Training Class Recommendations wanted Training - Internal, External, Online and Distance Learning 4
S Medical Device Class IIa Nonsterile ISO 13485 Alternatives ISO 13485:2016 - Medical Device Quality Management Systems 3
B Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope? ISO 13485:2016 - Medical Device Quality Management Systems 2
A ISO 13485 Risk Analysis - Class II Medical Device Engineering Changes ISO 13485:2016 - Medical Device Quality Management Systems 9
I Distributors for class II medical devices in US - Do we need ISO 9001 or ISO 13485 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C CAP for Medical Device Class Is product - Slightly confused - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 3
O ISO 13485 sample document or template - Class II medical device ISO 13485:2016 - Medical Device Quality Management Systems 16
Q Training Matrix - Job Class vs. Process Instructions - ISO 13485 Training - Internal, External, Online and Distance Learning 4
D ISO 13485 Certification - Class II product - Fluoride-free toothpaste - Records ISO 13485:2016 - Medical Device Quality Management Systems 1
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1

Similar threads

Top Bottom