ISO 13485 - Monitoring and Measurement of Processes (Sections 8.2.5 / 8.2.6)

H

HeatherC-S

#1
Hello

I am implementing 13485:2016 in my company and i am a bit confused with section 8.2.5 / 8.2.6 Monitoring and measurement of processes and product. To me this is just a way of flagging up when something has gone wrong with the product or the process, coming from customer complaint, audit, near-miss etc - which then leads on to CAPA. Am I missing something here?

Many thanks
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Hello

I am implementing 13485:2016 in my company and i am a bit confused with section 8.2.5 / 8.2.6 Monitoring and measurement of processes and product. To me this is just a way of flagging up when something has gone wrong with the product or the process, coming from customer complaint, audit, near-miss etc - which then leads on to CAPA. Am I missing something here?

Many thanks
Hello Heather and welcome to the Cove :bigwave:

8.2.5 speaks of the QMS as a whole. You need to state the results the processes are expected to produce, then actively monitor them (e.g. trend) to show that the processes are capable, them implement CAPA if they're not.

8.2.6 is more focused on demonstrating that the product meets its specifications and that the prescribed procedures and inspections have been followed as specified. It's quite straightforward so I don't think you're missing anything of essence.

Cheers,
Ronen.
 
#3
Hi Ronen!

I have the same issue. I see that thread is a little bit old, but I hope my question will get back to you. :)

I'm very confused with 8.2.5. How to demonstrate that processes are capable and effective?

Hello Heather and welcome to the Cove :bigwave:

8.2.5 speaks of the QMS as a whole. You need to state the results the processes are expected to produce, then actively monitor them (e.g. trend) to show that the processes are capable, them implement CAPA if they're not.

8.2.6 is more focused on demonstrating that the product meets its specifications and that the prescribed procedures and inspections have been followed as specified. It's quite straightforward so I don't think you're missing anything of essence.

Cheers,
Ronen.
 

Ronen E

Problem Solver
Staff member
Moderator
#4
Start by listing your QMS processes and the planned results for each.
Then collect and collate information about each of the above processes' performance (in relation to the planned results). This should be an ongoing process.
This should allow you to reach conclusions and take action as necessary.
 

indubioush

Quite Involved in Discussions
#5
Every process should have a metric or metrics. For example, nonconformity reports should be closed within 30 days. That is just an example. Then, metrics data is periodically reviewed, typically during management review. If the average number of days to close a nonconformity report is actually 90, you need to determine whether a corrective action is required. Let's say you determined that a corrective action is required. You would then initiate the corrective action process and determine the root cause of the issue, i.e., why your nonconformity reports take so long to be closed.
 
Thread starter Similar threads Forum Replies Date
B Measuring and monitoring equipment - Understanding which procedures to be compliant with ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
E ISO 13485 7.6 Control of monitoring and measuring equipment - Assess the Validity ISO 13485:2016 - Medical Device Quality Management Systems 4
R ISO 13485 Software validation procedure and Quality Objectives Monitoring wanted Document Control Systems, Procedures, Forms and Templates 1
T ISO 13485 8.2.5. Monitoring and measurement of quality management system processes ISO 13485:2016 - Medical Device Quality Management Systems 1
K ISO 13485:2016 Cl. 4.2.3 - Determine QMS Processes, Monitoring, Measuring, etc. ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Supplier Monitoring Minimum Requirements Supplier Quality Assurance and other Supplier Issues 3
P Control of Monitoring and Measurement Devices - ISO 13485 Clause 7.6 says ... ISO 13485:2016 - Medical Device Quality Management Systems 3
R ISO 13485 Clause 7.6 Control of Monitoring and measuring devices - Requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
J How to accomplish ISO 13485 Clause 8.2.3 Monitoring and Measuring? ISO 13485:2016 - Medical Device Quality Management Systems 5
C Monitoring and measurement - 7.6 of ISO 13485 - Documented procedure required? ISO 13485:2016 - Medical Device Quality Management Systems 5
N Quality Compliance Officer - ISO 13485, London Job Openings, Consulting and Employment Opportunities 0
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 4
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7
M Getting started in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 31
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
C SOP Template needed for ISO 13485 6.3 Infrastructure ISO 13485:2016 - Medical Device Quality Management Systems 9
T ISO 13485 8.3 - Non-Conforming Materials - on-line rework or part of process? ISO 13485:2016 - Medical Device Quality Management Systems 11
B Do IFU designs have to be document controlled under ISO 13485? Document Control Systems, Procedures, Forms and Templates 2
H ISO 13485 - Separate Microbiology Audits ISO 13485:2016 - Medical Device Quality Management Systems 8
C Production and Post Production feedback - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 4
T ISO 13485 - 5.5.1 Responsibility and authority - Small Company Independence ISO 13485:2016 - Medical Device Quality Management Systems 13
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
O ISO 13485 vs. GMP - Comparison matrix wanted EU Medical Device Regulations 4
O ISO 13485 - Is management review required before stage 1? ISO 13485:2016 - Medical Device Quality Management Systems 6
O In addition to the standard, what other ISO 13485 sources do people recommend? ISO 13485:2016 - Medical Device Quality Management Systems 5
Watchcat ISO 13485 for IVD (In-vitro Diagnostic Device) Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 9
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12

Similar threads

Top Bottom