Hi :bigwave:
I am quite new to the requirements of ISO13485:2003.
One of our medical devices supplier is ISO13485 Certified in UK.
One of their scope is calibration of test equipment. They are the OEM of the said medical devices and falls under test equipment.
My concern is NIST Traceability not required for calibration of medical devices since they are ISO13485 Certified? Is it acceptable that a medical device calibration not traceable to a national/international standard as long as you are ISO13485 certified? Please advice. :frust:
Thank you very much in advance.
Best regards,
Raffy :cool:
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#3
Hello Raffy!

If I understand what you're asking, you have a supplier that calibrates instruments for you. This supplier is certified within ISO13485, but the calibrations that are performed are not traceable to a national standard.

To my knowledge, if the calibration standard does not maintain traceability to a national standard/ SI unit, it's not a calibration.

That activity may be referred to as some kind of verification or performance check. But if there is no traceability, it's not a suitable standard for calibration.
 

howste

Thaumaturge
Super Moderator
#4
One of their scope is calibration of test equipment. They are the OEM of the said medical devices and falls under test equipment.
If their scope includes calibration of their product under ISO 13485, then they must have traceability to national/international standards. Here's what ISO 13485 says:
ISO 13485:2003 said:
Where necessary to ensure valid results, measuring equipment shall a) be calibrated or verified at specified intervals, or prior to use, against measurement standards traceable to international or national measurement standards.... Records of the results of calibration and verification shall be maintained.
My concern is NIST Traceability not required for calibration of medical devices since they are ISO13485 Certified? Is it acceptable that a medical device calibration not traceable to a national/international standard as long as you are ISO13485 certified? Please advice. :frust:
Per the above, traceable devices must be used, and they must maintain records. However, ISO 13485 does not require them to provide those records to YOU.

What does your purchasing information (PO or contract) specify to the supplier? These are the legal agreements between you and your suppliers. If your purchase order only specifies the product (without documentation) then they aren't required to provide the traceability information. If your PO specifies that they must provide traceability documentation, then they have to provide it.

I know many companies that will provide traceability information for an added cost. This needs to be negotiated BEFORE you buy. If you have the product without the traceability, then you may have to calibrate it yourself.
 
Thread starter Similar threads Forum Replies Date
H Contract Manufacturer as Design Owner ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 4
blackholequasar ISO 13485 certification prior to Medical Device Manufacturing... worth it? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Electronic Signatures - Non-Conformance - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 29
E ISO 13485 software validation ISO 13485:2016 - Medical Device Quality Management Systems 7
D Notified Bodies - ISO 13485 & MDR Technical Files ISO 13485:2016 - Medical Device Quality Management Systems 3
D Deviations - Where in ISO 13485 deviations are covered? ISO 13485:2016 - Medical Device Quality Management Systems 7
B ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 2
Y ISO 13485:2015 Software Validation IQ/OQ/PQ ISO 13485:2016 - Medical Device Quality Management Systems 13
J ISO 13485 for Metal Finishing Medical Device and FDA Regulations and Standards News 5
S How to calculate Effective Number of People for ISO 13485 Certification? General Auditing Discussions 2
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
E Any template/ form of Monitoring and Measurement of Processes and product to ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 1
H ISO 13485-paragraphs for a SaaS SAMD needed or not? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
M ISO 13485 and document control for programs ISO 13485:2016 - Medical Device Quality Management Systems 6
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
D Lead time to schedule an ISO 13485 audit General Auditing Discussions 2
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 3
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 4
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 9

Similar threads

Top Bottom