ISO 13485 - Separate Microbiology Audits
My company is a contract medical device manufacturer and currently searching for a new NB due to various reasons (we have had the same NB for 13 years). The current NB conducts separate microbiology audits and I questioned why that is. When the microbiology audit is performed they are auditing to the applicable 13485 standard and checking that we have our 14644 certification for all cleanrooms.
We manufacture in cleanrooms that are tested and certified by an outside company and do not perform any in-house sterilization. Our company's scope is "Manufacturing, inspection, packaging, and re-packaging of sterile and non-sterile medical devices." I am wondering if there is a requirement that we need to have the separate micro audit due to having "sterile" in the scope.
Thank you for any help!
Philip B
Yes, if you supply sterile medical devices, it is common to have a separate microbiology audit by your NB (we do). Although you are not sterilising the devices yourself, the audit will assess your controls over the sterilisation process, including selection and approval of sterilisation sub-contractors, sterilisation validation (eg from ISO 11137), bioburden monitoring, dose validation, sterilisation certification and release controls. As these are fairly specialist areas the NB likes to have an assessor trained in microbiology carry out the audit. I personally don't mind hosting microbiology audits, it keeps the scope nice and tight and you pretty much know what to expect.
levatorsuperioris
Not all NB require this to be separate, there are extensive requirements to audit Sterilization - these people are harder to come by. Some NB have enough of them on staff they perform the audit either offsite or through the regular audit.
sandra2014
Hello, I have 2 questions
1-can anyone tell me if there any medical device standard that requires supplier auditing?
2-If a company is already registered to ISO 13485:2016, will this suffice a supplier auditing requirement from a client?
I would appreciate any assistance from Elsmar Cove members.
Thank You!
chris1price
Take a look at ISO13485 7.4.1, it does not say you have to perform supplier audits; just "establish criteria for the evaluation and selection of suppliers" and among other criteria, that that it shall be risk based. That said, the more critical the supplier, the greater the likelihood that you should audit them. A supplier having ISO13485 is a good start, but shouldn't be the only criteria for making this decision.
Ralba
Hello Sandra,
We are ISO13485 certified, and we do not audit our suppliers. It has never come up for us in any of our audits.
You do, however, need to meet 7.4.1 like Chris stated above. This is a method you could use to evaluate that supplier, and it pretty common for extremely critical suppliers that need to be highly reliable. We evaluate using their ISO cert as well, but we also annually evaluate their metrics. That is, supplier Non-conformities and Supplier OTD. We require a thoroughly inspected sample before accepting them on the AVL if they are performing custom work of any kind.
Not saying you need to do all of this, just giving an idea of the controls that were necessary to evaluate them in our setting (CM in PCB assemblies).
sandra2014
Hello Sandra,
Thank You for your information, this helps! I appreciate it!
sandra2014
Yes , this is good advice, I believe some of our clients designated us as a critical supplier, hence annual audits are necessary from our clients. Thank You for clarifying this!
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Combined audit - ISO 9001 and ISO 13485 - Why Separate Audits? General Auditing Discussions 5
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 5
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M ISO 13485-2016 online certification ISO 13485:2016 - Medical Device Quality Management Systems 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 11
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 7
M Getting started in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 31
P ISO 13485:2016 MDSAP Certification Fee Survey ISO 13485:2016 - Medical Device Quality Management Systems 6
C SOP Template needed for ISO 13485 6.3 Infrastructure ISO 13485:2016 - Medical Device Quality Management Systems 9