Hi, I am a manufacturer of medical face mask and currently my face mask is a type IIR medical face mask under BS EN14683.

Just would like your kind advice whether the sterilization clause in ISO13485:2016 can be "non-applicable" ?

If the sterilization is non-applicable to us, then what is the cleanliness and contamination control procedure details required? Only the cleanroom verification and validation will be sufficient?

Please share your kind advice.

Thank you and have a nice day!
 
Last edited:
Elsmar Forum Sponsor

John C. Abnet

Teacher, sensei, kennari
Staff member
#2
Just would like your kind advice whether the sterilization clause in ISO13485:2016 can be "non-applicable"
Good day @amelol ;
There is no requirement in ISO 13485 for sterilization (unless sterilization is required).

Is your device... "sterile medical device" per term and definition 3.20 ?
Are there not statutory or regulatory requirements for your product to be sterile?

In regards to the "sterilization clause", sterilization is mentioned numerous times throughout the standard regarding documentation, and validation of ny sterilization that occurs. In particular, clause 7.5.2 requires your organization to identify and document "requirements for cleanliness". I would start there and as result, understand exactly what "requirements for cleanliness" do, or do not, exist.

Hope this helps.

Be well.
 
#3
Thanks John for sharing your thoughts. ;)

Good day @amelol ;
There is no requirement in ISO 13485 for sterilization (unless sterilization is required).

Is your device... "sterile medical device" per term and definition 3.20 ?
Are there not statutory or regulatory requirements for your product to be sterile?

Checked the face mask is not required to be sterile under the Malaysia medical device regulation. However, we do UV radiation oven sterilization on the face mask.

In regards to the "sterilization clause", sterilization is mentioned numerous times throughout the standard regarding documentation, and validation of ny sterilization that occurs. In particular, clause 7.5.2 requires your organization to identify and document "requirements for cleanliness". I would start there and as result, understand exactly what "requirements for cleanliness" do, or do not, exist.

Hope this helps.

Be well.
Checked the face mask is not required to be sterile under the Malaysia medical device regulation. However, we do UV radiation oven sterilization on the face mask. In this case, do I need to comply with the sterilization requirements?

Is cleanroom infrastructure and it’s cleanroom performance certified by NEBB considered fulfilling the “requirements for cleanliness”?
 

John C. Abnet

Teacher, sensei, kennari
Staff member
#4
In this case, do I need to comply with the sterilization requirements?
Good day @amelol ;
I want first issue the disclaimer that I do not know what regulatory or statutory requirement or ethics require your organization to sterilize or not to sterilize.

Having said that, you said in your prior post that "...the face mask is not required to be sterile under the Malaysia medical device regulation."

Remember, ISO 13485 is NOT a legal or regulatory requirement.

Do you have a copy of ISO 13485 and have you familiarized yourself with it? What is required by ISO 13485 is only what is stated within ISO 13485.

Hope this helps.
Be well.
 

Pjservan

Involved In Discussions
#5
Sterilization requirements only apply if one of the following exists:
  1. Regulatory Requirement
  2. Customer Requirement
  3. Internal Requirement
 
#6
Good day @amelol ;
I want first issue the disclaimer that I do not know what regulatory or statutory requirement or ethics require your organization to sterilize or not to sterilize.

Having said that, you said in your prior post that "...the face mask is not required to be sterile under the Malaysia medical device regulation."

Remember, ISO 13485 is NOT a legal or regulatory requirement.

Do you have a copy of ISO 13485 and have you familiarized yourself with it? What is required by ISO 13485 is only what is stated within ISO 13485.

Hope this helps.
Be well.
Yes, I have the copy of standard and attended awareness training for it. Just that certain requirements like sterilization is vague to be included or excluded with justification. So to certify with ISO13485, I am not sure if the sterilization requirements can be “non applicable” or not.
Thanks a lot for your inputs.
 
#7
Sterilization requirements only apply if one of the following exists:
  1. Regulatory Requirement
  2. Customer Requirement
  3. Internal Requirement
Thanks Pjservan for sharing your thoughts.
Internal requirement means it is depends on the organization’s decision to implement the sterilization process or not right? If implement sterilization process in the production of face mask, then the ISO13485 system will need to establish all the sterilization process control, verification and validation procedures n records. Correct?
 

planB

Super Moderator
#8
amelol,

you typically only talk about "sterilisation" when you render your product free of viable micro-organism with typically at least a sterility assurance level (SAL) of 10exp(-6). Is your UV irradiation process capable of providing this minimum SAL? Or is it more sort of a decontamination process (i.e. no real defined and validated microbial log reduction)

EN 14683:2019 specifies the microbial cleanliness in section 5.2.5: "[...]the bioburden of the medical mask shall be≤ 30 CFU/g [...]". There are no other defined cleanliness requirements in this standard other than stating "In the selection of the filter and layer materials, attention shall be paid to cleanliness." in section 5.1.1. Controlled environmental manufacturing conditions are surely favourable to meet this requirement. There

Hope this helps,
 
Thread starter Similar threads Forum Replies Date
D ISO 13485 for Hospital Sterilization Departments ISO 13485:2016 - Medical Device Quality Management Systems 2
C Validation of Sterilization Procedures - ISO 13485 Section 7.5.2.2 Document Control Systems, Procedures, Forms and Templates 4
xcanals_tecno-med.es ISO 13485 in Hospital Central Sterilization and Clinical Engineering Services ISO 13485:2016 - Medical Device Quality Management Systems 27
T Sterilization of Implantable bone screws - ISO 13485 and CE Mark EU Medical Device Regulations 14
C Cleaning and Sterilization - ISO 13485 sections 6.4 and 7.5.1.2.1 ISO 13485:2016 - Medical Device Quality Management Systems 3
W ISO 13485 Certification - No Sterilization, Packing nor Direct Shipping to End User. ISO 13485:2016 - Medical Device Quality Management Systems 6
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
P ISO 13485 (EN 46000) or ISO 9001 - Which Standard? Sterilization within a Hospital Hospitals, Clinics & other Health Care Providers 8
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 12
M ISO 13485 and document control for programs ISO 13485:2016 - Medical Device Quality Management Systems 6
M Customer Property - ISO 13485:2016 Clause 7.5.10 ISO 13485:2016 - Medical Device Quality Management Systems 9
pbojsen ISO 13485 Requirements versus FDA product classification and GMP exemptions - Audits ISO 13485:2016 - Medical Device Quality Management Systems 3
D Lead time to schedule an ISO 13485 audit General Auditing Discussions 2
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 1
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Operational Procedures for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 7
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
E ISO 13485 in Clinical Trial conduct: Applicable or No ISO 13485:2016 - Medical Device Quality Management Systems 2
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 3
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO 13485 Certification for Resin Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
T ISO 13485 - Assembly instructions written vs. online ISO 13485:2016 - Medical Device Quality Management Systems 5
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 8
N 93/42/EEC certification without ISO 13485 EU Medical Device Regulations 3
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
0 ISO 13485:2016 Chapter 8 Integration of the subsections ISO 13485:2016 - Medical Device Quality Management Systems 3
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
E ISO 13485 QMS certification as a Supplier ISO 13485:2016 - Medical Device Quality Management Systems 8
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
J Can signed agreements over-ride review of every "contract" under ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 9
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J Leveraging another company's ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 5
J New Job Position - Achieving ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 5