ISO 14971:2012 and the FDA

F

Frodeno

Hi Folks,

So the FDA lists ISO 14971:2007 in the recognised consensus standards database. Many EU companies are now 14971:2012 compliant. Is it the case that a 14971:2012 compliant risk file/risk analysis cannot be used with an FDA premarket notification? will FDA recognise the term AFAP over ALARP?

Thanks,

Frodo.
 

Ronen E

Problem Solver
Moderator
Hi Folks,

So the FDA lists ISO 14971:2007 in the recognised consensus standards database. Many EU companies are now 14971:2012 compliant. Is it the case that a 14971:2012 compliant risk file/risk analysis cannot be used with an FDA premarket notification? will FDA recognise the term AFAP over ALARP?

Thanks,

Frodo.

Hi Frodo,

What is the context in which you think you need the FDA to endorse your ISO 14971:2007 compliance?

Ronen.
 
F

Frodeno

Hi Ronen,

Thanks for replying. The context of this question is when the risk management documentation is being used within an FDA submission. I`m not expecting FDA to endorse 14971:2007 compliance - I`m wondering about their attitude to accepting documentation prepared using the 2012 version of the standard. Specifically, since many companies now are using the AFAP convention instead of ALARP when describing risk, will FDA accept this.

If FDA are not accepting documentation prepared in compliance with the 2012 version of the standard, does this mean that companies must prepare a 2007 compliant version of their risk documentation for submission purposes? i.e: should EU companies wishing to use thier risk documentation for FDA submission purposes maintain two versions one 2007 compliant and one 2012 compliant ? I would be really interested to hear from anyone who has experience with this.

Frodo
 

Ronen E

Problem Solver
Moderator
Hi Ronen,

Thanks for replying. The context of this question is when the risk management documentation is being used within an FDA submission. I`m not expecting FDA to endorse 14971:2007 compliance - I`m wondering about their attitude to accepting documentation prepared using the 2012 version of the standard. Specifically, since many companies now are using the AFAP convention instead of ALARP when describing risk, will FDA accept this.

If FDA are not accepting documentation prepared in compliance with the 2012 version of the standard, does this mean that companies must prepare a 2007 compliant version of their risk documentation for submission purposes? i.e: should EU companies wishing to use thier risk documentation for FDA submission purposes maintain two versions one 2007 compliant and one 2012 compliant ? I would be really interested to hear from anyone who has experience with this.

Frodo

Thanks for explaining.

In my opinion, in the context you describe the actual contents of the risk management documentation is far more important than formal compliance with any given version of the standard. On the other hand this might be something specific to your device - I don't know what it is, what specific controls / guidance apply etc.

I also think that the resulting documentation should not be very different, since the normative (obligatory) part of the 2 versions is identical. What you describe as a difference targeted at "meeting" EN ISO 14971:2012 is actually an enhancement directed at meeting the (underlying) intent of the MDD (as currently interpreted by the Commission), over and above what 14971 requires. Anyway, that excess is conservative relative to the original 14971 requirements and thus I can't imagine the FDA having real issues with a device / manufacturer that goes the extra mile in that direction.
 
F

Frodeno

Thanks again - I fully agree with your comments - my question was a general one rather than pertaining to specific device types but most of my work is in Programmable Electrical Medical Systems (PEMS). The stimulus for my question was an argument that I had with a RA/QA consultant a few weeks ago, who was adamant that FDA do not accept the 2012 version of the standard (specifically in the context of the use of AFAP). I found the reasoning a bit silly and I have a hard time beliving it but I do not have any direct experience of this. Over the coming months I will be preparing a 510(k) and I know that the risk analysis I will be using has been prepared using AFAP etc...so I guess I will see for myself.
 

mihzago

Trusted Information Resource
I agree with Ronen's assessment.
EN ISO 14971:2012 and ISO 14971:2007 are identical in terms of the normative portion of the standard. The differences are in the Z annexes that are included in the EN version of the standard. These annexes do not specify requirements but provide a guidance for complying with MDD, which are not at odds with complying with FDA requirements related to risk management.
 
Thread starter Similar threads Forum Replies Date
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 16
Q Information for safety EN ISO 14971:2012 - Customer Risk Reduction ISO 14971 - Medical Device Risk Management 6
A Our auditor requires us to attend a training on EN ISO 14971:2012 Other ISO and International Standards and European Regulations 3
W Risk Benefit Analysis - ISO 14971:2012 Requirements ISO 14971 - Medical Device Risk Management 27
B Interpreting Deviations 5 & 6 in Annex ZA in ISO 14971:2012 ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
M Labelling and the harmonized EN ISO 14971:2012 ISO 14971 - Medical Device Risk Management 9
F ISO 14971:2012 - 4.3 Initiating Events and Circumstances ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2012 - Verification of Implementation of Risk Control Measures ISO 14971 - Medical Device Risk Management 12
M MDD Europe: Use EN ISO 14971:2012 or ISO 14971:2007 Quality Manager and Management Related Issues 9
J EN ISO 14971:2012 Annex ZA Deviation #4 Clarification ISO 14971 - Medical Device Risk Management 3
J Annex ZA of ISO 14971:2012 - Deviation 7 - Incorrect Interpretation ISO 14971 - Medical Device Risk Management 10
Sam Lazzara Experiences with Notified Bodies addressing EN ISO 14971:2012 Annex Zs ISO 14971 - Medical Device Risk Management 5
K RISK ANALYSIS SAMPLE according to Annex ZA of EN ISO-14971-2012 Other Medical Device and Orthopedic Related Topics 1
R EN ISO 14971:2012 - Implementing Annex Z content deviations ISO 14971 - Medical Device Risk Management 1
renenatasha BS EN ISO 14971:2012, BS EN 62366:2008 and BS EN 15223-1:2012 Questions EU Medical Device Regulations 12
rob73 New version BS EN ISO 14971:2012 released - July 2012 ISO 14971 - Medical Device Risk Management 37
O ISO 14971 for Biologics? ISO 14971 - Medical Device Risk Management 6
O ISO 14971 on Hazards during service/installation ISO 14971 - Medical Device Risk Management 2
Q Risk Management ISO 14971 - Probability of Occurrence ISO 14971 - Medical Device Risk Management 8
Z Risk Management SOP ISO 14971 ISO 14971 - Medical Device Risk Management 1
K Help with ISO 14971: Benefit-Risk Analysis ISO 14971 - Medical Device Risk Management 3
B New to ISO 14971. Comparing to MIL-STD-882 ISO 14971 - Medical Device Risk Management 7
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
B Is 14971 Annex C checklist now in ISO/TR 24971 required to complete prior to 510k filing? ISO 14971 - Medical Device Risk Management 3
B ISO 14971:2019 amendment A11:2021 questions ISO 14971 - Medical Device Risk Management 5
K Questions about Table C.1 examples of hazards in Annex C of ISO 14971. EU Medical Device Regulations 1
M ISO 14971:2019 vs FMEA methodology ISO 14971 - Medical Device Risk Management 23
Y BS EN ISO 14971:2019+A11:2021 released ISO 14971 - Medical Device Risk Management 3
Q Harmonised Standards (EN ISO 13485 / EN ISO 14971) in MDR (2017/745/EU) ISO 13485:2016 - Medical Device Quality Management Systems 4
L PFMEA for test procedures (ISO 14971) ISO 14971 - Medical Device Risk Management 5
R ISO 14971 not harmonized ISO 14971 - Medical Device Risk Management 5
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 12
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
A ISO 14971 PFMEA Manufacturing Risk ISO 14971 - Medical Device Risk Management 2
K Overall residual risk according to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 12
Bill Hansen New ISO 14971:2019 Harm: unreasonable psychological stress, and cybersecurity ISO 14971 - Medical Device Risk Management 13
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B ISO 14971 Applied to Software ISO 14971 - Medical Device Risk Management 2
D Recent changes to ISO 14971 - SOP required for managing standard revisions ISO 13485:2016 - Medical Device Quality Management Systems 1
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
Ronen E Informational What's new in ISO 14971:2019 ISO 14971 - Medical Device Risk Management 2

Similar threads

Top Bottom