ISO 14971 Clause 9 Requirements - Post-Production Monitoring and Risk Management

C

Conquer

In ISO 14971:2009 clause 9 talks about post-production information and review and update of the risk management file.

I am looking for ideas/thoughts on how people meet this requirement. Do you yearly review each product and then create a report that gets added to the risk management file. My company has quite a few products and I am trying to determine an effective but least burdensome way to ensure we are meeting this clause of the standard.

Any feedback or input you have would be appreicated.
 

pkost

Trusted Information Resource
Re: Post-production monitoring

There are numerous ways to a carry out post market activities; they should all feed back into your risk management system.

1. Customer complaints
2. Other feedback from your customers, probably through sales reps, maybe questionnaires
3. Servicing history
4. Literature review etc.

If your products are CE marked then you should be carrying out clinical evaluations, this is what I would consider the key document that collates all of your post market activities. MEDDEV 2.7.1 offers guidance about how to carry out a clinical evaluation. Essentially it is an objective review of your all available literature regarding your product including that of competitive products and anything else the reviewer considers to be relevant

With regards to timescales to carry out the evaluation I risk assess the product, for instance our class 1 instruments have a review cycle of 3 years (there have never been any complaints regarding them and they have a lot of history). One of our class III products is being reviewed every year, perhaps less if we become aware of any relevent data.

throughout your PMS activities you may come across a reportable incident, this will of course require a more rapid response and cannot wait until your next evaluation!
 
S

SteveK

Re: Post-production monitoring

throughout your PMS activities you may come across a reportable incident, this will of course require a more rapid response and cannot wait until your next evaluation!

A case in point. I recently came across a recall of a class IIa device as part of my PMS (in Italy and Germany) due to a particular mode of failure which I was previously unaware of. Because we produce a similar device, this acted as a trigger for me to update my Risk Management file for this item.

pkost has given a spot on reply by the way!:applause:

Steve
 

Marcelo

Inactive Registered Visitor
You need an ACTIVE system to gather information regarding your product and similar products, meaning, do not rely only on reactive information such as complaint reports.

The examples mentioned are great, but please remember that you also need internal information (the "production" part).

As mentioned pkost, the new CE marking requirement for post-marketing clinical follow-up is a must, and even if not CE marked, i would suggest that something like that must be done (post market clinical follow-up will problably be required by other regulations in the future because this is exactly the main worry regarding product effectiveness).
 
Last edited:
R

Roland Cooke

My competitor makes an identical product to ours, I heard it killed someone last week. Luckily it wasn't a complaint against us, so there's no need to revisit our risk management documentation. :notme:

Joking aside, as Marcelo indicated, this has to be an active system, and you need to think fairly wide. Obviously the amount of effort required will be proportional to the risks and novelty of the device, how it used, and by whom.

For many established low-risk devices, an annual review of complaints etc may indeed be adequate. But for a novel public-use device, perhaps a daily review of social networking sites would be required?
 
M

Martin IT

To be practical, with the aim to monitor your products, you should involve your commercial department (product specialists, vendors, etc.). They are in close contact with the final user so they could see if the final user uses the product in a way that could be unsafe or dangerous. Very often, the product is enough safe, but become dangerous if used in a way that the designer has not thought yet or in conjunction with particular situations. In my opinion is more and more important know how the final user 'understands' and uses your product and is better if yuo have this kind of information before an accident has been occured!

Last point more people are involved in the safety risk monitor and better is it! Regular meetings is a way to keep all the people informed and more important involved in the process.
 

patmelad

Starting to get Involved
Can all of these items be reviewed and audited in one form and if so does anyone have a simple template to do so? I am looking to limit the amount of files that need to be maintained in our Risk Management file for our product. We are a very small company and I am the regulatory department...
Any Help would be appreciated.

Thank you
 

pkost

Trusted Information Resource
I won't say it is impossible to do on a single form, but I think it would be very difficult to capture all of the information.

I suspect it would only really work on a low risk device with an extensive history that is well understood and where the device hasn't changed in years/decades
 
Thread starter Similar threads Forum Replies Date
Sam Lazzara ISO 14971 Clause 7 - Evaluation of Overall Residual Risk Acceptability ISO 14971 - Medical Device Risk Management 3
T Defining Criteria for Risk Acceptability - ISO 14971 Clause 3.2 ISO 14971 - Medical Device Risk Management 4
O ISO 14971 for Biologics? ISO 14971 - Medical Device Risk Management 6
O ISO 14971 on Hazards during service/installation ISO 14971 - Medical Device Risk Management 2
Q Risk Management ISO 14971 - Probability of Occurrence ISO 14971 - Medical Device Risk Management 8
Z Risk Management SOP ISO 14971 ISO 14971 - Medical Device Risk Management 1
K Help with ISO 14971: Benefit-Risk Analysis ISO 14971 - Medical Device Risk Management 3
B New to ISO 14971. Comparing to MIL-STD-882 ISO 14971 - Medical Device Risk Management 7
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
I IEC 60812 or ISO 14971 for PFMEA? What should we use? ISO 14971 - Medical Device Risk Management 3
B Is 14971 Annex C checklist now in ISO/TR 24971 required to complete prior to 510k filing? ISO 14971 - Medical Device Risk Management 3
B ISO 14971:2019 amendment A11:2021 questions ISO 14971 - Medical Device Risk Management 5
K Questions about Table C.1 examples of hazards in Annex C of ISO 14971. EU Medical Device Regulations 1
M ISO 14971:2019 vs FMEA methodology ISO 14971 - Medical Device Risk Management 23
Y BS EN ISO 14971:2019+A11:2021 released ISO 14971 - Medical Device Risk Management 3
Q Harmonised Standards (EN ISO 13485 / EN ISO 14971) in MDR (2017/745/EU) ISO 13485:2016 - Medical Device Quality Management Systems 4
L PFMEA for test procedures (ISO 14971) ISO 14971 - Medical Device Risk Management 5
R ISO 14971 not harmonized ISO 14971 - Medical Device Risk Management 5
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 12
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 16
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
A ISO 14971 PFMEA Manufacturing Risk ISO 14971 - Medical Device Risk Management 2
K Overall residual risk according to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 12
Bill Hansen New ISO 14971:2019 Harm: unreasonable psychological stress, and cybersecurity ISO 14971 - Medical Device Risk Management 13
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B ISO 14971 Applied to Software ISO 14971 - Medical Device Risk Management 2
D Recent changes to ISO 14971 - SOP required for managing standard revisions ISO 13485:2016 - Medical Device Quality Management Systems 1
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
Ronen E Informational What's new in ISO 14971:2019 ISO 14971 - Medical Device Risk Management 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
Y When will Notified Bodies require MedDev manufacturers to fully implement ISO 14971:2019? ISO 14971 - Medical Device Risk Management 1
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
M Informational ISO 14971 / ISO TR 24971 revision update – atualizações sobre a revisão Medical Device and FDA Regulations and Standards News 1
R The difference b/w FMEA & Risk analysis as per iso 14971 ISO 14971 - Medical Device Risk Management 8
D Risk management according to ISO 14971 - When to document risk controls? ISO 14971 - Medical Device Risk Management 10
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
Q Information for safety EN ISO 14971:2012 - Customer Risk Reduction ISO 14971 - Medical Device Risk Management 6
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
A Our auditor requires us to attend a training on EN ISO 14971:2012 Other ISO and International Standards and European Regulations 3
S In a risk analysis, how can we tie mobile app security breach to ISO 14971? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
FoGia IMDRF opened a Consultation on Annex E & F and the link to ISO 14971 ISO 14971 - Medical Device Risk Management 4

Similar threads

Top Bottom