ISO 14971 - Product Risk Analysis and Reviews - New requirements?
W
Wicker
Hello all,
This is my first thread, I've looked to see if there are any similar ones regarding ISO14971 and haven't found anything.
I work for a large med device company and we were audited more than once last year for various reasons. One common area that came up was regarding compliance to ISO14971. The ISO auditors mentioned that we don't have a comprehensive review of risks for our products. They mentioned that we should have some type of regular review of product risks. The problem is that we have thousands of products that belong to fifty or so product families, a 'regular' review of risks would not be an easy task. We also have several ways of reviewing risks (e.g. FMEAs, Risk Mgmt Reports, mgmt review, complaints, health hazard analyses, to name a few) but most are performed independent of each other.
So my question is how do we get a reasonable handle on regular reviews of risks without backing ourselves into a corner? I don't want to go overboard and commit to something that will be very difficult to comply.
Has this been pushed by anyone else's ISO auditors? I hear that it may be asked more of from auditors.
madannc
Hello all,
Hi Wicker,
When you are in design I presume you perform risk analysis and where appropriate put in RCM's (Risk Control Measures)?
During the design phase here we have a template which just about mirrors 14971, the idea being that all areas of the standard are covered (presuming you have trained/competent risk assessors).
We revisit this review when upgrading and determine if the upgrade/enhancement affects the original analysis (and document).
So I guess I am asking with so many products was this noted during the review of the design for these? and are you looking for a fix to retrospectively perform Risk Analysis?
As for a regular review, this would be performed during design phase, and once product has been released...you would not revisit unless something changes (e.g. upgrade/enhancement) or you get a complaint to investigate.
Somehow I think I am stating the obvious here, maybe I should ask how do you perform risk analysis currently?
cheers Nigel
Roland Cooke
There is a ton of discussion about ISO14971 in the ISO13485 forum. From one perspective, ISO14971 is a subset of ISO13485, and in fact is referenced therein.
madannc
In the same way I would also say BS EN 60601-1 has a relationship with ISO 14971
Marcelo
I do not agree. If you get, for example, the MDD, you see that what the regulatory requirements want is that the risks related to the product be acceptable. Almost all countries with regulatory requirements use a risk management-based regulation system.
Using the process of ISO 14971, for example, you can guarantee that the risks are acceptable. Nor ISO 14971 nor the MDD nor almost any regulatory requirement has a REAL need for a quality system (because in fact a quality system is, in it´s root, worried about "doing always the same thing", not "doing safe things"). That´s way ISO 14971 cannot be a subset of ISO 13485.
Also, surely there´s a relatioshipo between ISo 14971 and IEC 60601, or any safety standard. Safety standards provide recognized ways to control risks. When your device meets a standard requirement, it means that the hazardous situation related to the requirement (and thus the risks of that particular hazardous situation) are deemed acceptable. So, in the risk control pohase of ISO 14971, you can use this do demonstrate the risk acceptability. Simple as that.
Roland Cooke
ISO13485:2003 7.1 makes direct reference to ISO14971 as a specific footnote. That is basically why I referred to ISO14971 as being a "subset" of ISO13485.
More importantly companies will build the requirements of ISO14971 into their overall quality management system, allowing them to build risk management concepts into all of their activities that are covered by the requirements of Clause 7.
Other safety standards tend to be very proscriptive, whereas ISO14971 is instead about establishing a system. Tools that can be used within that system (e.g. FMEA) are described in the Appendices.
babarian
Hello folks:
It's my first time in this forum.
Normally a risk management for each kind / family of product is needed, and this risk management need to be re-evaluated, e.g. customer complaints & state to the art of the product, periodically with a conclusion.
And I will make a conclusion for all the products comprehensively periodically, e.g. during the management review annually.
Q ISO 14971 Class II Medical Devices - Product Realization & Risk Management ISO 14971 - Medical Device Risk Management 5
V ISO 14971: 2007 Risk Management File - Separate file required for each product? ISO 14971 - Medical Device Risk Management 3
T Risk Management Plan for a Mature Product - ISO 14971 ISO 14971 - Medical Device Risk Management 2
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 6
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 10
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
A ISO 14971 PFMEA Manufacturing Risk ISO 14971 - Medical Device Risk Management 2
K Overall residual risk according to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 8
Bill Hansen New ISO 14971:2019 Harm: unreasonable psychological stress, and cybersecurity ISO 14971 - Medical Device Risk Management 13
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B ISO 14971 Applied to Software ISO 14971 - Medical Device Risk Management 2
D Recent changes to ISO 14971 - SOP required for managing standard revisions ISO 13485:2016 - Medical Device Quality Management Systems 1
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
Ronen E Informational What's new in ISO 14971:2019 ISO 14971 - Medical Device Risk Management 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
Y When will Notified Bodies require MedDev manufacturers to fully implement ISO 14971:2019? ISO 14971 - Medical Device Risk Management 1
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
M Informational ISO 14971 / ISO TR 24971 revision update – atualizações sobre a revisão Medical Device and FDA Regulations and Standards News 1
R The difference b/w FMEA & Risk analysis as per iso 14971 ISO 14971 - Medical Device Risk Management 8
D Risk management according to ISO 14971 - When to document risk controls? ISO 14971 - Medical Device Risk Management 10
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
Q Information for safety EN ISO 14971:2012 - Customer Risk Reduction ISO 14971 - Medical Device Risk Management 6
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
A Our auditor requires us to attend a training on EN ISO 14971:2012 Other ISO and International Standards and European Regulations 3
S In a risk analysis, how can we tie mobile app security breach to ISO 14971? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F IMDRF opened a Consultation on Annex E & F and the link to ISO 14971 ISO 14971 - Medical Device Risk Management 4
M Updates to EN 62366 & ISO 14971? Other Medical Device Related Standards 3
D IEC 60601-1 and ISO 14971 Assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
P ISO 13485 and ISO 14971 - one mandates the other? ISO 13485:2016 - Medical Device Quality Management Systems 8
D Rationale for Risk Acceptability Matrix - ISO 14971 ISO 14971 - Medical Device Risk Management 9
H Task analysis and ISO 14971 ISO 14971 - Medical Device Risk Management 9
M ISO 14971 and Stand-Alone Diagnostic Software ISO 14971 - Medical Device Risk Management 4
dgrainger Benefit - What is the definition of Benefit in ISO 14971? ISO 14971 - Medical Device Risk Management 7
Y Training as a risk control for ISO 14971 ISO 14971 - Medical Device Risk Management 13
W Risk Benefit Analysis - ISO 14971:2012 Requirements ISO 14971 - Medical Device Risk Management 27
thisby_ ISO 14971 - ALARP and P2 - New ISO 14971 does not allow the concept of ALARP? ISO 14971 - Medical Device Risk Management 3
C What is the difference between "Overall Risk" and "Risk"? (ISO 14971) ISO 14971 - Medical Device Risk Management 10
B New EU Medical Device Regulation & Reconciling with EN ISO 14971 EU Medical Device Regulations 41
B IFU and deviation 7 in ISO 14971 Annex ZA ISO 14971 - Medical Device Risk Management 1
B Interpreting Deviations 5 & 6 in Annex ZA in ISO 14971:2012 ISO 14971 - Medical Device Risk Management 1
B Our NB says that IEC 62304 is an ISO 14971 Requirement ISO 14971 - Medical Device Risk Management 1