ISO 14971 Risk Analysis - Effects of component failure on the patient
DuttonT
When doing an ISO 14971 risk analysis as a manufacturer - Do we take into account what might happen to a patient if a component fails?
Example: a light does not go on that should and causes an extended procedure.
Should we have documented that a light not going may cause a problem?
Thanks in advance.
Al Rosen
Holed-up in a Hotel in South Florida
When doing a 14971 risk analysis as a manufacturer - Do we take into account what might happen to a patient if a component fails?
Thanks in advance.
Yes, the risk analysis should take into consideration the risk of normal use as well as failure modes.
Marcelo
Sorry, i didn´t understand your question. The component - the light in the example - is a part of the equipment or device you manufacture??
DuttonT
If the equipment failed during an operation - and the failure caused the operation to take longer. That would be an extended procedure.
Marcelo
And this warning light would be a visual sign of an alarm which is a means of risk control for a particular hazard? Look, it´s really difficult to give a technical answer if you do not use the correct technical terminology.
Usualy "warning lights" are used in that way.
If the "warning light" is used in that way on the device, yes you have to document it because it´s a risk control measure, meaning, it warns the operator that a fault condition exists.
Anyway, you have to tak into account the failureof ANY component of the device which could lead into a hazard, not just a warning light.
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 3
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
R The difference b/w FMEA & Risk analysis as per iso 14971 ISO 14971 - Medical Device Risk Management 8
D Risk management according to ISO 14971 - When to document risk controls? ISO 14971 - Medical Device Risk Management 10
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
Q Information for safety EN ISO 14971:2012 - Customer Risk Reduction ISO 14971 - Medical Device Risk Management 6
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
S In a risk analysis, how can we tie mobile app security breach to ISO 14971? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
D Rationale for Risk Acceptability Matrix - ISO 14971 ISO 14971 - Medical Device Risk Management 9
Y Training as a risk control for ISO 14971 ISO 14971 - Medical Device Risk Management 13
W Risk Benefit Analysis - ISO 14971:2012 Requirements ISO 14971 - Medical Device Risk Management 27
C What is the difference between "Overall Risk" and "Risk"? (ISO 14971) ISO 14971 - Medical Device Risk Management 10
S Organizing Risk Analysis and Controls for a New Medical Device (ISO 14971) ISO 14971 - Medical Device Risk Management 4
A Is Risk Management Process compliant to ISO 14971 in absence of Hazardous Situations? ISO 14971 - Medical Device Risk Management 5
N Risk Severity Estimation for Medical Devices as per ISO 14971 ISO 14971 - Medical Device Risk Management 12
Marc Are you looking for ISO 14971 - Medical Device Risk Management? Risk Management Principles and Generic Guidelines 1
J Does anyone have an example of Risk-Benefit Analysis per ISO 14971? Other ISO and International Standards and European Regulations 2
M ISO 14971:2012 - Verification of Implementation of Risk Control Measures ISO 14971 - Medical Device Risk Management 12
N ISO 14971 Risk Analysis - Sections 4.2 and 4.3 ISO 14971 - Medical Device Risk Management 2
D ISO 14971 - Risk Analysis Best Practices ISO 14971 - Medical Device Risk Management 5
A ISO 14971 Figure E.1 that starts with Hazard and ends with Risk ISO 14971 - Medical Device Risk Management 3
A Risk Acceptance Criteria in ISO 14971 ISO 14971 - Medical Device Risk Management 19
M Risk Management (ISO 14971:2007) Internal Audit Checklist ISO 14971 - Medical Device Risk Management 7
K RISK ANALYSIS SAMPLE according to Annex ZA of EN ISO-14971-2012 Other Medical Device and Orthopedic Related Topics 1
K What is the policy for Risk Acceptability per ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 2
M The Sequence of ISO 14971 Risk Analysis Activities ISO 14971 - Medical Device Risk Management 14
Sam Lazzara ISO 14971 Clause 7 - Evaluation of Overall Residual Risk Acceptability ISO 14971 - Medical Device Risk Management 3
C Risk Analysis of a "Medical System" according to ISO 14971 ISO 14971 - Medical Device Risk Management 5
N ISO 14971:2007 vs. 2009 - Which Risk Management Standard is still accepted in the EU Other ISO and International Standards and European Regulations 2
E ISO 14971:2009 Risk Management Requirements CE Marking (Conformité Européene) / CB Scheme 2
M ISO 14971 Medical Device Risk Management FAQ ISO 14971 - Medical Device Risk Management 38
M Risk Management Plan Template - ISO 14971:2007 Compliant ISO 14971 - Medical Device Risk Management 13
M ISO 14971 Risk Analysis and use of a DFMEA (Design FMEA) ISO 14971 - Medical Device Risk Management 8
C Scope of Risk Management in ISO13485 vs. ISO 14971/EU MDD ISO 14971 - Medical Device Risk Management 2
C ISO 14971 Clause 9 Requirements - Post-Production Monitoring and Risk Management ISO 14971 - Medical Device Risk Management 7
A Where can I buy EN ISO 14971:2009 (Medical Device Risk Management)? ISO 14971 - Medical Device Risk Management 11
Q ISO 14971 Class II Medical Devices - Product Realization & Risk Management ISO 14971 - Medical Device Risk Management 5
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I Medical Device Risk Assessment to ISO 14971 ISO 14971 - Medical Device Risk Management 4
B Application of Risk Management - ISO 14971 for a Tooling Manufacturer ISO 14971 - Medical Device Risk Management 18
M Risk management, ISO 13485 and ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 10
G Risk Analysis - Templates of EN ISO 14971:2007 and others ISO 14971 - Medical Device Risk Management 8
M ISO 14971:2007 Risk Management - Class I A Sterile Medical Device ISO 14971 - Medical Device Risk Management 4
M ISO 14971 Risk Management Qualification of Personnel ISO 14971 - Medical Device Risk Management 5
Q ISO 14971 for Contract Manufacturers - Extensive risk management ISO 14971 - Medical Device Risk Management 16
V ISO 14971:2007 Application of Risk Management to Medical Devices ISO 14971 - Medical Device Risk Management 9
Q Risk Analysis of Software - ISO 14971:2007 ISO 14971 - Medical Device Risk Management 29
M ISO 14971 Risk Management questions and comments ISO 14971 - Medical Device Risk Management 38