Hi All,

I was hoping to get input on whether medical devices that are not reusable and have a Do Not Reuse symbol on them also need the Do Not Resterilize symbol. The thinking is that if it's not for reuse, then whether it is resterilizable or not is moot, because resterilizing it would only be associated with reuse. The devil's advocate response is, that conceivably if your sterile barrier is broken before use, if there is no Do Not Resterilize label, someone might subject it to a second sterilization before the single use. For our single-use disposables we are leaning toward just using the Do Not Reuse symbol, but we're not 100% confident. Thoughts? Thanks in advance.
 
Elsmar Forum Sponsor

indubioush

Quite Involved in Discussions
#2
If you think there is a risk that someone would accidentally resterilize and reuse, you can add it. If not, leave it off since the do-not-reuse label will suffice.
 

monoj mon

Involved In Discussions
#3
The devil's advocate response is, that conceivably if your sterile barrier is broken before use, if there is no Do Not Resterilize label, someone might subject it to a second sterilization before the single use.
Exactly for this reason I would use the "Do Not Re-sterilize" symbol on my label and IFU based on my risk assessment. Another reason for my decision is that uner MDR 2017/745,
  1. the GSPR 23.3 states "The following particulars shall appear on the sterile packaging: (j) an instruction to check the instructions for use for what to do if the sterile packaging is damaged or unintentionally opened before use"
  2. the GSPR 23.4 states "The instructions for use shall contain all of the following particulars: (l) if the device is supplied sterile, instructions in the event of the sterile packaging being damaged or unintentionally opened before use"
To fulfill the above requirements the first thing I would do is place a symbol and relevant description for "Do Not Re-sterilize".

I am not sure about your target market, but since this type of situation is already mentioned by regulatory authority, I would not take the risk of omitting that particular "Do Not Re-sterilize" symbol.

I hope this will be helpful for you to make a decision.
 

monoj mon

Involved In Discussions
#5
To fulfill these requirements, you can use this symbol:

View attachment 26877
Yes, true. This symbol is the primary one to go on the labeling but my reasoning says this one is not sufficient enough to fulfill the requirements as well as avert any situation like this one as Kat C has mentioned, when someone thinks it's okay to re-sterilize it.
The devil's advocate response is, that conceivably if your sterile barrier is broken before use, if there is no Do Not Resterilize label, someone might subject it to a second sterilization before the single use.
Perhaps I should have said apart from using the "Do not use if package is damaged" symbol as a continuation to this sentence.
To fulfill the above requirements the first thing I would do is place a symbol and relevant description for "Do Not Re-sterilize".
 
#6
Exactly for this reason I would use the "Do Not Re-sterilize" symbol on my label and IFU based on my risk assessment. Another reason for my decision is that uner MDR 2017/745,
  1. the GSPR 23.3 states "The following particulars shall appear on the sterile packaging: (j) an instruction to check the instructions for use for what to do if the sterile packaging is damaged or unintentionally opened before use"
  2. the GSPR 23.4 states "The instructions for use shall contain all of the following particulars: (l) if the device is supplied sterile, instructions in the event of the sterile packaging being damaged or unintentionally opened before use"
To fulfill the above requirements the first thing I would do is place a symbol and relevant description for "Do Not Re-sterilize".

I am not sure about your target market, but since this type of situation is already mentioned by regulatory authority, I would not take the risk of omitting that particular "Do Not Re-sterilize" symbol.

I hope this will be helpful for you to make a decision.
Yes, true. This symbol is the primary one to go on the labeling but my reasoning says this one is not sufficient enough to fulfill the requirements as well as avert any situation like this one as Kat C has mentioned, when someone thinks it's okay to re-sterilize it.


Perhaps I should have said apart from using the "Do not use if package is damaged" symbol as a continuation to this sentence.
Thank you, monoj mon and indubioush, for your discussion. Since you added the "Do not use if package is damaged" symbol to the discussion, I looked up its details in ISO 15223-1, and in the Informative Notes it says: "NOTE This symbol may also mean "Do not use if the product sterile barrier system or its packaging is compromised"." In that case, it seems like this symbol can satisfy GSPR 23.3 and GSPR 23.4 if the IFU defines the symbol as "Do not use if the product sterile barrier system or its packaging is compromised", which should specifically address the situation of someone thinking it's OK to use at all after package damage, resterilization or not.

In the end, my thought is that for a single-use, sterile product the symbols "Do not re-use" and "Do not use if the product sterile barrier system or its packaging is compromised" are sufficient, and cover the two scenarios where there would be forseeable risk of resterilization. The only other scenario where the device could be resterilized would be if the package is not damaged and someone decided to resterilize it for no apparent reason, and I believe that this is an extremely remote possibility and thus does not need to be labeled to address. I welcome comments.
 
Thread starter Similar threads Forum Replies Date
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
B Does EN ISO 15223-1:2016 include the graphic symbols to be added to software and IFU? Other Medical Device Related Standards 9
A MDRS Standards Harmonisation - copy of ISO 15223-1 EU Medical Device Regulations 3
J ISO 15223-1: Keep dry vs. humidity limitation label symbol Other Medical Device Related Standards 2
E Where can I buy ISO-15223-1 CD with graphic symbols? Other Medical Device Related Standards 13
E ISO 15223-1 (Medical Device Label Symbols) - Revisions, EU commission? Other ISO and International Standards and European Regulations 1
K EN ISO 15223-1:2012 Clarification or Examples on when to use Safety Symbols Other Medical Device Related Standards 3
R Is BS EN ISO 15223-1 a harmonised standard? EU Medical Device Regulations 2
rob73 When will BS EN ISO 15223-1:2012 replace EN 980 EU Medical Device Regulations 2
renenatasha BS EN ISO 14971:2012, BS EN 62366:2008 and BS EN 15223-1:2012 Questions EU Medical Device Regulations 12
M BS EN ISO 15223-1:2012 replaces EN 980 - What is the effect of the change? Quality Manager and Management Related Issues 44
R Use of EN 980 and ISO 15223 Symbols for Non Medical Mevices Other Medical Device Related Standards 6
bio_subbu ISO 15223-2 for safer use of symbols with Medical Devices - Released new standard Other Medical Device Related Standards 1
C ISO 15223 or EN 980 - What standard should be used in CE technical file? EU Medical Device Regulations 9
R Device Labeling Standards: Is ISO 15223:2000 Replacing EN980? ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 9001:2015 - Small Shop ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 0
M ISO 9001:2015 case study sample ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Sample of Nonconformity report for ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 0
Q ISO 9001 8.5.1 - Control of production and service performance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B Do IFU designs have to be document controlled under ISO 13485? Document Control Systems, Procedures, Forms and Templates 1
H ISO 13485 - Separate Microbiology Audits ISO 13485:2016 - Medical Device Quality Management Systems 1
M Case study help as per ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
M Case study solution help required as per ISO 9001 : 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 0
C Production and Post Production feedback - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 4
T ISO 13485 - 5.5.1 Responsibility and authority - Small Company Independence ISO 13485:2016 - Medical Device Quality Management Systems 13
A Refusal to discuss ISO 9001 obligations... what to do? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
B Can we be ISO 9001 certified without a physical office? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
J Informational ISO 24971:2020 Released ISO 14971 - Medical Device Risk Management 0
S ISO 45001 and outsourcing the transporation of products Occupational Health & Safety Management Standards 3
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
A Question on ISO 14001:2015 - Are annual audits required? ISO 14001:2015 Specific Discussions 8
N Which EN ISO 17664 version compliance to EU MDR? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 3
Sidney Vianna Informational New edition of ISO 29001 released (June 2020) Oil and Gas Industry Standards and Regulations 0
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
L Combining 3 ISO 9001 registrations into 1 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 7
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
N EN ISO 13640:2002 vs EN ISO 23640:2011 Other Medical Device Related Standards 1
O ISO 13485 vs. GMP - Comparison matrix wanted EU Medical Device Regulations 0
A ISO 10002:2018 Checklist Needed (Complaints Handling) Customer Complaints 5
N Sampling Plan for Internal Audits - ISO 2859 or 3951 - Or Neither? Internal Auditing 6
O ISO 13485 - Is management review required before stage 1? ISO 13485:2016 - Medical Device Quality Management Systems 6
BeaBea ISO 9001 Customer Feedback Methods - What has worked for your company? Service Industry Specific Topics 16
O In addition to the standard, what other ISO 13485 sources do people recommend? ISO 13485:2016 - Medical Device Quality Management Systems 5
Watchcat ISO 13485 for IVD (In-vitro Diagnostic Device) Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 9
G New to ISO 9001 - Where to begin? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Similar threads


















































Top Bottom