Clause 4.2.2 requires "When the laboratory is required by law or authorized by contractual arrangements to release confidential information, the customer or individual concerned shall, unless prohibited by law, be notified of the information provided".

Please give me a help to understand the difference between "be notified" and "be informed"? Does official informing the customer or individual concerned about releasing confidential information covers requirement of "be notified"?
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
So I brushed up on my legalese and came up with this explanation:

When used as verbs, inform means to instruct, train (usually in matters of knowledge), whereas notify means to give (someone) notice of (something).

So let's look at how the words are used here.

4.2.1 says that "The laboratory shall inform the customer in advance, of the information it intends to place in the public
domain." This would indicate that you are instructing the customer on what information shall be considered public domain and not confidential.

4.2.2 says "When the laboratory is required by law or authorized by contractual arrangements to release
confidential information, the customer or individual concerned shall, unless prohibited by law, be
notified of the information provided"

You originally defined what is confidential when you informed (trained) the customer of your terms. You are to notify (give notice to) the customer when you are required to release what has been previously accepted as confidential information.
 

Tallinec

Starting to get Involved
#3
So I brushed up on my legalese and came up with this explanation:

When used as verbs, inform means to instruct, train (usually in matters of knowledge), whereas notify means to give (someone) notice of (something).

So let's look at how the words are used here.

4.2.1 says that "The laboratory shall inform the customer in advance, of the information it intends to place in the public
domain." This would indicate that you are instructing the customer on what information shall be considered public domain and not confidential.

4.2.2 says "When the laboratory is required by law or authorized by contractual arrangements to release
confidential information, the customer or individual concerned shall, unless prohibited by law, be
notified of the information provided"

You originally defined what is confidential when you informed (trained) the customer of your terms. You are to notify (give notice to) the customer when you are required to release what has been previously accepted as confidential information.
But what means "be notified" if we do not consider clause 4.1.1. (because there is no need for us to place information in the public domain)?
 

dwperron

Trusted Information Resource
#4
But what means "be notified" if we do not consider clause 4.1.1. (because there is no need for us to place information in the public domain)?
17025 is written for both the calibration and test worlds. This is more applicable to testing labs, but could possibly involve calibration labs.

Imagine a medical test lab. As a customer, you would want one that maintains strict confidentiality of test results. So you would inform your customer of this policy, and the lab would agree that all test results are to be kept confidential.
Now consider that a police investigation gets a court order for the lab to release test results. Legally you must produce those results. Under 17025 you would be required to notify your customer that you have released the information.

This could possibly happen in the calibration world. Consider an airplane crash, and the investigation leads to questions of the accuracy of the equipment used to check the plane's avionics. A court order could compel you to release that calibration information, then you must notify your customer of the release.
 

Tallinec

Starting to get Involved
#5
17025 is written for both the calibration and test worlds. This is more applicable to testing labs, but could possibly involve calibration labs.

Imagine a medical test lab. As a customer, you would want one that maintains strict confidentiality of test results. So you would inform your customer of this policy, and the lab would agree that all test results are to be kept confidential.
Now consider that a police investigation gets a court order for the lab to release test results. Legally you must produce those results. Under 17025 you would be required to notify your customer that you have released the information.

This could possibly happen in the calibration world. Consider an airplane crash, and the investigation leads to questions of the accuracy of the equipment used to check the plane's avionics. A court order could compel you to release that calibration information, then you must notify your customer of the release.
As I understand your reply, «be notified» comes after „be informed“ or «be notified can not live without be informed».

I am in „lab business“ (non-medical and private) many years and nobody informed our lab about intention to place information about us in the public domain. And our lab have no practice and need to place information about customer (for example names of customer’s technicians) in the public domain.

May be the key difference between «be notified» and „be informed“ one shall find in juridical field. May be «be notified» shall be mandatory documented and approved by responsible person. May be “be notified” have juridical meaning in clause 4.2.2 of the standard „Be informed“ have more light-weighted, non mandatory meaning.

It is interesting to know you opinion.
 
Thread starter Similar threads Forum Replies Date
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
Douglas E. Purdy ISO/IEC 17025:2017 3rd Ed. Changes from 2nd Ed. ISO 17025 related Discussions 6
Douglas E. Purdy ISO/IEC 17025:2017 Clause 8 & Annex B ISO 17025 related Discussions 9
D Looking for PowerPoint presentation for ISO 17025:2017 ISO 17025 related Discussions 1
Ajit Basrur Informational ISO/IEC 17025:2017 Published - November 2017 ISO 17025 related Discussions 8
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 3
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
B ISO 17025 8.5 Actions to address risks and opportunities ISO 17025 related Discussions 7
G ISO 17025 - Requirement to Monitor Environment ISO 17025 related Discussions 10
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
B Supervision of personnel ISO 17025 Cl. 6.2.5(d)?? ISO 17025 related Discussions 1
D pH Validation for Water (ISO 17025) ISO 17025 related Discussions 0
S RSL (Remote Site Location) Laboratory - Does it need to be ISO 17025 IATF 16949 - Automotive Quality Systems Standard 5

Similar threads

Top Bottom