I recently went through a lab assessment and had a deficiency against clause 7.1.3 surrounding communication of a decision rule to customers. The clause states:
The lab is testing to methods that have decision rules "inherent" in them, and I make a statement about this in our lab manual. This was something that was confirmed during the assessment.
Separately, the lab also has a set of guard band limits to which it will typically pass/fail a test result. In all cases, the guard band limits are set within the limits identified by the method or standard. Because we are pass/failing results based on the guard band limits, the assessor gave us the deficiency under the reasoning that we are no longer using the limits (decision rule) that is inherent in the method, and therefore we must define those limits to, and gain agreement from the customer per clause 7.1.3
Wanted to get some different thoughts surrounding if this deficiency holds any weight? I'm having a difficult time with the interpretation that we need to disclose the use of guard band limits to customers, when all that we are claiming to comply with is the method.
When a customer requests a statement of conformity to a specification or standard for the test or calibration (e.g. pass/fail, in-tolerance/out-of-tolerance), the specification or standard and the decision rule shall be clearly defined. Unless inherent in the requested specification or standard, the decision rule selected shall be communicated to, and agreed with, the customer.
Separately, the lab also has a set of guard band limits to which it will typically pass/fail a test result. In all cases, the guard band limits are set within the limits identified by the method or standard. Because we are pass/failing results based on the guard band limits, the assessor gave us the deficiency under the reasoning that we are no longer using the limits (decision rule) that is inherent in the method, and therefore we must define those limits to, and gain agreement from the customer per clause 7.1.3
Wanted to get some different thoughts surrounding if this deficiency holds any weight? I'm having a difficult time with the interpretation that we need to disclose the use of guard band limits to customers, when all that we are claiming to comply with is the method.