ISO 17025 Calibration
We are a medical device company certified to ISO 13485 and are trying to understand the difference between a standard NIST calibration and a ISO 17025 calibration, specifically starting with being concerned for hi-pot and ground bond testers, and electrical safety testers. Would also be interested in knowing the difference between the two calibrations for simple measurement equipment like micrometers and calipers. Is there a different, does ISO 17025 only certify the lab performing the calibration? Is 17025 calibration required for medical device companies?
yodon
I think you're conflating 17025 with calibration. 17025 is a management system standard for test and calibration to which you can get certified for a particular scope. Indeed, many calibration service providers have 17025 certification. Just because a cal lab is certified to 17025 does not make them appropriate for ALL calibration activities - you need to check the scope of their certification.
13485 requires calibration to be done against "measurement standards traceable to international or national measurement standards..." Often, calibrated equipment is calibrated against NIST standards. If your team has such standards (and have the necessary competence to perform calibration) then you can certainly perform calibration internally.
17025 is not required for medical device companies. Even most large companies that do a lot of calibration internally don't pursue 17025.
AllTheThings
The difference between "NIST" (a calibration traceable to a National Metrology Institute) and ISO-17025 is as yodon said, a management system accreditation for the calibration lab itself.
When you use an accredited calibration lab, you are getting assurance of the following:
* The calibration was traceable, including the uncertainties. This is often neglected by non accredited labs.
* The lab performing the calibration has been evaluated as competent to perform that calibration
* The calibration is conducted according to recognized standards or methods, under controlled conditions.
* The results have been reported to you with all information required to evaluate the suitability of the calibration with respect to your processes AND the uncertainties, if you ask. This is important if you are intending to flow down measurement uncertainty into your design or manufacturing processes.
For your purposes, you do not have to be ISO-17025 accredited. But selecting calibration labs that are accredited gives you significantly better assurance that a calibration is valid than going to someone claiming "NIST traceable". The FDA encourages ISO-17025 calibrations on a voluntary basis, considering it a best practice. It is referenced in numerous places, including here: Accreditation Scheme for Conformity Assessment (ASCA)
Now, the risk to you of using a non-accredited calibration lab (Note: 'non-accredited' cals from an accredited lab are typically OK) comes in section 7.5.9 and 7.6. If a cal lab is NOT accredited, and you use them, your 13485 auditor would be well justified in asking you to prove traceability and calibration. Are you prepared to audit a non-accredited cal supplier with respect to traceability? Do you have the staff with the skillsets to do that?
Calibration is roughly defined as a relation between two values with measurement uncertainty declared. Are you prepared to audit a non-accredited calibration supplier with respect to their measurement uncertainties?
Using a 17025 accredited cal lab obviates the need for auditing your cal supplier with respect to the 13485 requirement in section 7.6a, and relieves you of some recordskeeping that would otherwise be required under 7.5.9.1.
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
S Is it mandatory for a Calibration Lab to be ISO 17025 accredited? General Measurement Device and Calibration Topics 7
G ISO 17025 Calibration Laboratory Assessor's Common Questions and Procedures ISO 17025 related Discussions 11
J ISO 17025 Guide and Ideas for a New Calibration Business ISO 17025 related Discussions 18
P Calibration in the UK to ISO 17025..cant find anyone ISO 17025 related Discussions 8
N Implementing ISO 17025 in a Calibration Laboratory ISO 17025 related Discussions 8
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
Crusader Calibration Source not ISO 17025 accredited ISO 17025 related Discussions 6
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
S What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
A Calibration system must meet ISO 10012, ISO 17025 or ANSI/NCSL Z540? ISO 17025 related Discussions 38
L Calibration by a non-ISO 17025 Laboratory ISO 17025 related Discussions 4
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
Z Force & Strain Calibration Uncertainty for ISO/IEC 17025 ISO 17025 related Discussions 2
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
D Calibration of Load Cells - ISO 17025 & MSA (Measurement Systems Analysis) ISO 17025 related Discussions 5
D ISO 17025 and Calibration Laboratory Risk Management Requirements ISO 17025 related Discussions 1
1 Pharma Industry stance on ISO 17025 Accreditation - In-House Calibration Laboratories ISO 17025 related Discussions 3
M ISO 17025 - Recommended or Required and other Calibration questions ISO 17025 related Discussions 5
N ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ISO 17025 related Discussions 30
M Specific Requirement for Calibration - ISO 17025 Clause 5.6.2 Interpretations ISO 17025 related Discussions 15
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
S ISO 17025 Validation Requirements - "calibration using reference standards or..." ISO 17025 related Discussions 5
H Finding an ISO 17025 accredited laboratory to perform calibration ISO 17025 related Discussions 14
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
suildur Assuring the Quality of Test and Calibration Results - ISO IEC 17025 Clause 5.9 ISO 17025 related Discussions 13
K Subcontracting of Tests and Calibration - ISO 17025 Clause 4.5 ISO 17025 related Discussions 1
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
D Calibration Frequency of CMM as per ISO/IEC 17025 ISO 17025 related Discussions 11
U I need All ISO 17025 Documents for a calibration laboratory ISO 17025 related Discussions 7
H Can you tell me a suitable "tooling" for Caliper Calibration - ISO 17025 ISO 17025 related Discussions 16
M Old calibration standard methods - Question for ISO 17025 auditors..... ISO 17025 related Discussions 1
R ISO 17025 section 4.9 - Control of Nonconforming Testing and Calibration Work ISO 17025 related Discussions 6
T Calibration Procedures - ISO/IEC 17025, Clause 5.5.9 - Outsourced Calibration ISO 17025 related Discussions 10
B Pin Gage Calibration Procedure - ISO 17025:2005 accreditation requirements? ISO 17025 related Discussions 8
B ISO 17025 logo & ISO 9001 logo on calibration certificate/report ISO 17025 related Discussions 7
apestate How do you find nearby ISO 17025 registered calibration laboratories? ISO 17025 related Discussions 1