ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements
tacom
ISO 17025 Clause 5.4.2 says
“The laboratory shall confirm that it can properly operate standard methods before introducing the tests or calibrations. If the standard method changes, the confirmation shall be repeated.”
According to this statament The Standard test method must be confirmed before use in the first time by laboratory.
My question:
İf the Standard test method does not changes at all
a) we do not have to confirm method any more
or
b) we have to confirm that we properly operate Standard methods for predetermined period (for instance once a year)
which one is correct.
Thanks in advance
TACO
Marcelo
If it dos not change, there's no need to reconfirm it. Think of the confirmation as a qualification, meaning, you confirm that you can operate the method and the only need to reconfirm it if there's any change in the process.
tacom
Thanks Marcelo
What do you mean saying “…the only need to reconfirm it if there's any change in the process”
I think if any of the paremeters (new equipment, new laboratory staff, noticeable change in equipment suh as usage of new column in chromotography etc. .) contributed to test result changed significantly after the inital confirmation , reconfirmationstudies should be done.
As it can be easily known that many testing equipment especially used instrumental analysis will began to lost their performance with time .
For instance , Dedection limit (LOD) may get worse in the course of time
What is your opinion?
TACO
Marcelo
I think if any of the paremeters (new equipment, new laboratory staff, noticeable change in equipment suh as usage of new column in chromotography etc. .) contributed to test result changed significantly after the inital confirmation , reconfirmation studies should be done.
We meant almost the same thing. However, I do not think you need to reconfirm the process if new staff id added - this is handled by the training process. As you mentioned, the standard only require reconfirmattion (requalification) "If the standard method changes".
This is generally related to process limits, not to changes in standard method. If your equipment cannot fulfill it's objective inside your test method (process) you need to replace the equipment. There's no change to the test method here.
Hershal
Metrologist-Auditor
Generally, if the method is a validated published method/procedure, the lab does not need to verify the method/procedure, but will have to prove to the assessor that they can conduct the test or calibration in accordance with the method/procedure. The verification then basically means that the method/procedure is up to date and that the lab can complete it with any changes it may have.
tacom
Dear Marcelo and Hershal
AOAC INTERNATIONAL describe confirmation (verification) as given below in its guide called “ H o w t o M e e t I S O 1 7 0 2 5 R e q u i r e m e n t s f o r M e t h o d V e r i f i c a t i on)”
In this guide, to confirm is the same as to verify.
Verification that a laboratory can adequately operate a standard method requires that the laboratory provideobjective evidence the performance parameters specified in the test method have been met with the matrices to which the method is being applied. Most often, the critical requirements are the accuracy and the precision(generally accepted as repeatability and reproducibility) which are reflected in the measurement uncertainty. The objective evidence is the accuracy and precision obtained from actual lab data.
Therefore confirmation studies generally include precision (repetability, intermediate precision) , trueness (bias,/recovery) and LOD&LOQ studies. Some publication I have read (published by European Communities) claim “verification should be performed every year if the method is used regularly”.
Normally accreditation process renews once evey four year according to accreditation body’s request. I believe, under normal condition this studies must be performed again once evey four year.
What I need to know is whether it is really necessary to perform these studies every year like the way some publications says or not.
Thanks your interest
TACO
Hershal
Metrologist-Auditor
I note the quote includes the statement that verification is the same as confirmation. Confirm annually.
If your national AB has their own rules then of course you need to abide by them.
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
Douglas E. Purdy ISO/IEC 17025:2017 Clause 8 & Annex B ISO 17025 related Discussions 9
N ISO 17025 Clause 4.10 regarding Improvement ISO 17025 related Discussions 3
N ISO 17025 Clause 4.6 Purchasing Services & Supplies (Compliance) ISO 17025 related Discussions 3
N ISO 17025 clause 5.8.3 Records of Non-Conforming Test Items ISO 17025 related Discussions 3
A Addressing every clause in ISO 17025 ISO 17025 related Discussions 1
N Measurement Uncertainty Explanation per ISO/IEC 17025:2005 clause 5.10.4.2 ISO 17025 related Discussions 3
M Specific Requirement for Calibration - ISO 17025 Clause 5.6.2 Interpretations ISO 17025 related Discussions 15
W Issuing Authority - ISO 17025 Clause 4.3.2.3 ISO 17025 related Discussions 3
J Small Laboratories & ISO 17025 Clause 4.1.4 - Small company 4 employees ISO 17025 related Discussions 13
C Control of Equipment Manuals - ISO 17025 Clause 4.3.1 ISO 17025 related Discussions 6
suildur Assuring the Quality of Test and Calibration Results - ISO IEC 17025 Clause 5.9 ISO 17025 related Discussions 13
K Subcontracting of Tests and Calibration - ISO 17025 Clause 4.5 ISO 17025 related Discussions 1
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
howste Condition of Item Calibrated - ISO 17025 Clause 5.10.2f - Calibration Certificate ISO 17025 related Discussions 13
V CMM Inspection Layouts vs ISO 17025 Clause 7.6.3.2 Requirements ISO 17025 related Discussions 5
V ISO/IEC 17025 Clause 4.4 - Review of requests, tenders and contracts ISO 17025 related Discussions 4
T Calibration Procedures - ISO/IEC 17025, Clause 5.5.9 - Outsourced Calibration ISO 17025 related Discussions 10
A How to Audit ISO 17025 Clause 4.2.7 - Upper Management ISO 17025 related Discussions 1
Hershal ISO/IEC 17025:2005 Clause 4.1.5 and 4.1.6 ISO 17025 related Discussions 5
D Appropriate Equipment for ISO/IEC 17025:2005 Clause 5.6.3.3 - Intermediate Check ISO 17025 related Discussions 5
J Issuing Authority on Forms - Required? ISO 17025 Clause 4.3.2.3 ISO 17025 related Discussions 2
K Is formal Internal Auditor training required for ISO 17025:2004 - Clause 4.14.1 ISO 17025 related Discussions 5
K AS9100 6.2.2 - Internal ISO 17025 Laboratory - Clause 5.2 Personnel Best Practices ISO 17025 related Discussions 1
A How do I address ISO 17025:2005 Clause 4.10 Improvement ISO 17025 related Discussions 3
V What is a US National Equivalent to ISO 17025? TS 16949 Clause 7.6.3.2 ISO 17025 related Discussions 6
S ISO 17025 Clause 5.9 - Calibration results - How to achieve what ISO wants ISO 17025 related Discussions 7
T ISO 17025 and ISO/TS-16949 Clause 7.6.3.1 Requirements - In-House Laboratory ISO 17025 related Discussions 2
J What are National Equivalences for Guide 25 & ISO/IEC 17025? Clause 4.11.2.b.1 ISO 17025 related Discussions 5
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 3
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6