ISO 17025 Compliance within an ISO 9001 Organization
Sarge
I have been tasked with bringing our testing lab into alignment with (as opposed to accredited to) ISO 17025. The lab is part of a larger ISO 9001 registered organization.
Does anyone have any experience with this? Are there any particular pitfalls I should watch out for?
Should there be a separate Quality Manual and such for the lab?
Any input would be appreciated
P
Phiobi
Re: ISO 17025 within an ISO 9001 organization
I'm not sure if this is the "best" way to do this, but recently I came accross the same situation within a company and they were ISO9001:2008 but had established procedures based on 17925 clause 5. They get the best of both worlds... business system to 9001 and technical controls to 17025.
qusys
Yes. ISO 17025 requests to have a dedicated manual.
My suggestion is to well undestand the reason why you would like to pursue the certifcation for your Lab.
Do your Lab provide extenal services that other customers buy?
If not, Iso 9001 or ISO TS are sufficient for your purpose.
You can also consider to follow the standards ISO 17025 without being certified from a Registrar.
Hope this helps
Jen Kirley
Quality and Auditing Expert
Good morning,
There may be a squeamishness about making a quality manual within a quality manual so to speak, but in the case of 17025 I would support such an idea because the standard does have specific requirements.
An alternative is an appendix in the existing quality manual, and/or attachments dealing with the technical processes/unique elements.
I generally find that my organization's ISO/TS requirements are in general more strict than 17025. So there are a great many elements that can, and do say llittle more than "See ___ (site procedure for the subject)." I would certainly not advise duplicating element descriptionsin a new manual, lest you have to change two manuals when one subject gets updated. That could be a mess, IMO.
I hope this helps!
suildur
First of all, you have not stated if there will be an accreditation application; that also have effects on the system which will be established.
Anyway, we, I mean the company in which I am also a shareholder, is going to be accredited according to ISO IEC 17025 as a calibration laboratory. Also, there will be validation and testing of clean rooms in our job according to ISO 14644 series, which also means we will have to apply for ISO 17020 accreditation as an inspection body.
Novadays, I am on the final step of finishing the implemantation of an integrated system, 17025 plus 17020. I have stated that the system is based on ISO 9001 in various procedures. As you may probably have seen, ISO IEC 17025 has a cross table with ISO 9001. In fact, all management systems are parallel to ISO 9001. So, I do not think that it would be wrong to go on as you have proposed.
Yes, there are specific features for documentation, infrastructure, work conditions, data control, monitoring etc. according to ISO IEC 17025, but that doesnot mean you cannot build an integrated system. In Turkey, there are several accredited CBs which are also an NB, inspection body and testing body which use only one manual for all those different systems.
Here you can see my documentation preview for my system too:
UNFORTUNATELY THE LINK DID NOT SURVIVE THE OCTOBER 2018 SOFTWARE MIGRATION
Also, if the lab will not be accredited, but still continue with ISO 9001 certification plus application of the ISO IEC 17025, I can propose something else. First of all, you can state the position of the lab and use of the ISO IEC 17025 in the ISO 9001 manual, then you can establish a separate manual for the lab, which will include all the ISO IEC 17025 specific procedures and the related cross references to ISO 9001 QM and ISO 9001 procedures for common processes (documentation, control of records, purchasing etc.). You can bound the lab's instructions, SOPs, forms etc. to the lab QM. That way, the ISO IEC 17025 specific procedures will not be separate procedures in the ISO 9001 system, and, you will not be expanding your whole system with lots of techically specific procedures, as you will only have a separate QM for the lab and your existing ISO 9001 system at the end.
Sarge
Thanks for the quick responses!
Hope this helps
We are the product test lab for a manufacturer of a government regulated (international) product line. We design, manufacture and sell safety devices, so to do our testing right is not optional.
As I stated in my original post, we plan to align with 17025 but not seek accredidation at this time. Just use 17025 as a "best practice" and at some later date actually seek accreditation if and when it makes good business sense. To that end, we want to set everything up according to the spec the first time.
Hershal
Metrologist-Auditor
Practical basis: First, compare the standards, and by that I mean specific requirements.
An example is Management Review, which has a prescriptive agenda in ISO/IEC 17025:2005 Clause 4.15.1; and also requires the top management in charge of the lab participate in the review. Such a review can be included in the ISO9001 review, but the laboratory must be specifically addressed and the minutes (however kept) must reflect that.
Another is purchasing. When you eventually seek accreditation, any vendors you are using for the lab must be able to support your accreditation. That is, meet appropriate requirements. As an example, a manufacturer that is ISO9001 would be adequate and appropriate, but a calibration lab that is only ISO9001 would not be, unless you had also assessed them, and then make sure you have the auditors qualifications (including technical) handy.
As to the separate manual, a smaller manual can then be written for the lab that addresses requirements of ISO/IEC 17025:2005 that are not addressed by ISO9001.
Hope this helps.
JodiB
Resurrecting this discussion thread to keep all info in one place, as my question is somewhat similar to the original poster's....
My org has multiple certifications for ISO 9001 (different functional groups had their own), but the certifications have never specifically included the operational group that operates the field analytical labs that I am now trying to bring into 17025 accreditation. The quality group intends to merge the multiple ISO 9001 certifications into only 2 multi-site certs and has republished the quality manual as the Level 1 for both multi-site certs. The ISO 9001 quality manual states that it is intended to be applicable to the entire organization.
I am writing a quality manual for the ISO 17025 accreditation of the labs. I'm a bit hung up on where our quality manual falls in the documentation hierarchy. Are we supposed to consider the ISO 9001 manual as a higher level governing document, which would put the 17025 manual as a Level II doc? Or do we disregard the quality group's 9001 manual and place our own 17025 manual as our Level I? ( I don't mean disregard, as in ignoring some of their basic requirements, but disregard in terms of how our 17025 quality system is set up..)
Does an accreditation for 1702 somehow fall "under" the company's ISO 9001 certification? Or are they two separate things? Would the labs need to be specifically listed on the scope of the company's ISO 9001 certification for the 9001 to apply?
Jen Kirley
Since 17025 includes all the essentials such as control of documents, supplier control, training etc., it is meant to accomplish QMS development and maintenance independent of 9001. So, the quality manual doesn't need to include activities covered under 17025, and your 17025 facility doesn't need to refer to the parent organization's 9001 quality manual; it would have its own.
However: if you are like my people, some of your processes (such as document control and training) are shared. If there is any particular or special treatment in those processes for this lab, such should be included in their documentation on a process-by-process basis.
JodiB
So the quality manual for the labs 17025 is independent and doesn't have to reference any other quality system and therefore is its own Level 1.... Correct? That's what I was hoping for. The quality manager in charge of the 9001 program was trying to tell me that my 17025 manual would be a Level II since it was specific to the laboratories.
But the laboratories have never been included in the 9001 certification so I didn't see how that could be true.
Seems to me that the labs will have our own stand-alone accreditation and essentially be just a supplier to the rest of the organization. (Our testing is done to support our own internal operations.) We may closely mirror some of the activities used in the 9001 certification (such as doc control), but there is no requirement that we do things exactly the same way since it is a separate accreditation.
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
N ISO 17025 Clause 4.6 Purchasing Services & Supplies (Compliance) ISO 17025 related Discussions 3
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
T Laboratories Claiming ISO 17025 Compliance - Sick of it..... ISO 17025 related Discussions 19
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4