Help resolve a dispute on interpretation of ISO17025 requirements within TS16949:2009.
During a recent CB audit, we got a minor NC when the auditor checked the 17025 accredited lab scope of our calibration service provider on line and did not find the specific test documented in the CoC.
When we checked with the calibration lab, they said:
1. The lab is ISO 17025 certified but only those tests specifically requiring an "accredited calibration" are listed on their 17025 lab scope. All other tests, including the cal we required, are listed in thier lab capability.
2. The lab's 17025 CB agrees that all tests performed by the lab comply with ISO17025, section 4.
We agree with our CB that there was a non-conformity - the calibration service provider did not have a lab scope meeting the definition in TS16949:2009 (listed below) for the test we required.
Where the dispute is:
Our CB interprets 7.6.3.2 requirements (listed below) as requiring all tests to be listed on the 17025 accredited lab scope. The calibration service provider interprets the requirements to mean that ISO 17025 accreditation is required and having a lab scope is required but that having all tests on an ISO 17025 accredited lab scope is not. Their position is their internal procedures for each test meet the requirements for a lab scope as defined in TS 16949.
We're in the middle. Our CB says all test must be on the ISO17025 lab scope. Our cal service provider says they can add them but it will increase our calibration costs.
What should we do?
[FONT=Arial,Bold][FONT=Arial,Bold]laboratory scope - [/FONT][/FONT]controlled document containing:
* specific tests, evaluations and calibrations that a laboratory is qualified to perform,
* list of the equipment which it uses to perform the above, and
* list of methods and standards to which it performs the above
[FONT=Arial,Bold][FONT=Arial,Bold]7.6.3.2 External laboratory[/FONT][/FONT]
External/commercial/independent laboratory facilities used for inspection, test or calibration services by the organization shall have a defined laboratory scope that includes the capability to perform the required inspection, test or calibration, and either:
* there shall be evidence that the external laboratory is acceptable to the customer, or
* the laboratory shall be accredited to ISO/IEC 17025 or national equivalent.
qusys
ISO/TS 16949 requirements for ISO 17025
What should we do?
In my opinion, this is not a non conformity.
Your supplier is ISO 17025 certified and they report the capability to do the test inside their manual and procedure .
There is no requirement that the same items have to be reported ( one by one) in the certificate.
Stijloor
Friends,
For correct interpretation: A laboratory can be ISO 17025 accredited not certified. ;)
Stijloor.
qusys
Re: ISO/TS 16949 requirements for ISO 17025
Friends,
Stijloor.
Agree with you , Stijloor as per requirement.
But then the organization should demonstrate thi evidence, for example, with assessment of the Customer or second part assessment approved by the Customer ( see note 1).
I have understood that the Lab that the organization uses for their scope is "certified" vs 17025 given that they showed to the CB the certificate on the web. So, it already over meet the requirement according to my opinion, always remembering the control of this service supplier by the organization certified ISO TS.
Pls, correct me if I am wrong:bigwave:
JAltmann
From my interupation of the events that took place your CB is correct, and their was a NC. The story you relay of the lab's opinion and their CB's opinion is incorrect, the lab can only preform work accredited to ISO-17025 for which they have been granted such authority. They may however preform other services that are not accredited as they deem fit to do so.
You since you are conforming to ISO-16949 have the responsibility that any calibration work being preformed by a 3rd that work must be preformed by an accredited soucre, unless the customer has given you a waiver for this calibration services or if no such accredited soucre exsists.
One thing i am not sure of was this work for some form of testing or specifically calibration?
Helmut Jilling
Auditor / Consultant
Help resolve a dispute on interpretation of ISO17025 requirements within TS16949:2009....
...Where the dispute is:
What should we do?...
The NC is correct. The lab's position is illogical. If they have a certificate, and they have a scope, if these documents don't synchronize, it would make the whole process pointless. A lab is accredited for a certain scope of activities. That is why a 17025 certificate is linked to the scope of what was verified. If a lab is accredited to calibrate micrometers, it is does mean they are also accredited to calibrate electronic instruments like a CMM. They are only accredited to perform the tests that are listed and signed off on the scope, because those are the tests that are audited and verified.
The problem is the more tests that are verified and accredited, the more expensive the audit gets for the lab. So, there are a LOT of games being played in the lab field. And this sounds like it falls into that category.
If the accreditating body also agrees with the lab, then I would check whether that lab is among the approved major ones. Something does not sound right here. Their certificate may not even be issued by an approved accreditation body.
Perhaps Hershal will weigh in here. He is very knowledgeable in this area.
N
ntrprizxile
Clarification:
The lab in question has ISO 17025 accreditation (I incorrectly stated they were certified). The scope of their accreditation includes many services, including calibrations, that we have them perform. The particular calibration of an instrument in question was not on their lab scope.
The lab is an nationally known lab and is accredited by a very reputable ISO 17025 CB. I would venture to guess that many on this thread has heard of this company and everyone on this thread is familiar with their CB.
The question in dispute is whether 7.6.3.2 requires that all calibrations be ISO 17025 accredited calibrations. The text says, the laboratory must have a lab scope and that it must be ISO 17025 accredited. Both items the lab in question says they have.
The lab claims and their ISO 17025 CB affirms that the lab is accredited and that all activities in the lab fall under the same quality management system required of ISO 17025 section 4.
We found fault in how they presented their "lab capability" for services not included on their accredited lab scope as it did not meet the definition found in the TS 16949 standard and requested they address that as a proper lab scope - accredited or not.
We are faced with a decision: Require the lab to include all activity we give them on the accredited lab scope and face increased costs, or contest the interpretation of our CB.
qusys
Clarification:
Ok, understood now.
Thks for explanation.
Having betetr specified , I must correct my previous post.
This is a non conformity for you.
Remain that fact related the audit vs tihs supplier:bigwave:
sitapaty
What should we do?
Your CB is correct.Go to a dfferant calibration supplier.
-Sitapaty
justdavew32
I have a question regarding this. We received a Minor NC in relation to weighing scales.
The Calibration House we used has a defined scope and is 17025 accredited to carry out the test. They are also iso9001.
We used this company to carry out calibration on our weighing scales and they supplied us with ISO certs. Does every calibration have to be carried out under the 17025 banner or do they just have to have the test on their scope?
Thanks in advance for any advice here.
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
S What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
T ISO 17025 - 4.1.5b Internal/External commercial, financial, and other pressures ISO 17025 related Discussions 3
J ISO/IEC 17025 - Which External Documents require control? ISO 17025 related Discussions 6
L Acceptable use of a non accredited external ISO/IEC 17025 laboratory ISO 17025 related Discussions 5
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 3
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2