ISO 17025 Facilities and environmental conditions SOP
#1
Would anyone be kind enough to share a template for facilities and environmental conditions SOP for 17025 accreditation?
Elsmar Forum Sponsor
dwperron
Trusted Information Resource
#2
There are no 17025 SOP's for environmental conditions.
17025 states:
The facilities and environmental conditions shall be suitable for the laboratory activities and
shall not adversely affect the validity of results.
The requirements for facilities and environmental conditions necessary for the performance of
the laboratory activities shall be documented.
You will need to determine what environmental conditions need to be met for whatever testing or calibration that your organization is performing, document those requirements, show that you have controls in place to maintain those conditions, and collect data proving that you continually meet those requirements.
#3
Hi Klyoyo. I just posted for you the Procedure that we are using and it was accepted during many Audit assessments. Please feel free and ask any question. Kindest regards, Constantin
#4
The SD Log that we are suing includes the LINE VOLTAGE, AMBIENT TEMPERATURE, RELATIVE HUMIDITY, SIGNATURE of the TESTING PERSONNEL who is recording the data and OBSERVATIONS. IF any of the MEASURED VALUES is OUTSIDE the REQUIRED LIMITS. (in our case established by the Standards that we have under our Scope of accreditation) it will trigger a Report to the Technical Manager and the Testing will be put on hold.
#5
Hello,
I was given the job to implement the iso17025 but have no idea where to start. I just saw that you shared some SD log and procedure but cannot find the place to download it. I'm sorry i'm writing in here but because i'm new i cannot send PM until i write 10 of my own posts. Unfortunately i'm newbie so spamming 10 useless posts only to get to somebody doesn't make sense to me but i understand the restriction. If you have some tips where i should look first at EC regarding iso17025 i would very appreciate it. Thank you!
Jim Wynne
Leader
Admin
#6
Hello,
You can get the procedure here: ISO 17025 PROCEDURE FOR FACILITIES AND ENVIRONMENT (elsmar.com)
Thread starter Similar threads Forum Replies Date
gard2372 ISO 17025 Certified Calibration Laboratory Has Moved Facilities ISO 17025 related Discussions 6
E ISO 17025 Key Performance Indicators (KPIs) ISO 17025 related Discussions 5
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
D Training for ISO 17025 ISO 17025 related Discussions 2
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
Q Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 2
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 14
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
Similar threads
Top Bottom