ISO 17025 Method Verification - One sided “F” test or two sided “F” test

tacom

Involved In Discussions
#1
The Standard deviation obtained in house method verification study must be compared to precision found in the colloborative study. The aim of this study is to demostrated that the result produced by laboratory is acceptable (method verification). That mean my standart deviation must be smaller or equal than found in the colloborative study
So, I decided to use “F test . What should i do? One sided “F” test or two sided “F” test
Thaks in advance
 
Elsmar Forum Sponsor

itsbiodiversity

Involved In Discussions
#2
Are you performing testing such as chemical analysis or are you performing calibrations? Are you trying to conform to 17025 (2005) Section 5.9? Are you familiar with En score and z scores? The most important is the En score.

The En score is calculated according to this formula: (Xi - Xref) / (SQROOT(ucsquared + ucrefsquared)) where Xi is your value, Xref is the "assigned" value, ucsquared is your uncertainty squared, and ucrefsquared is the "assigned" uncertainty.

Z score is calculated according to this formula: (Xi - Xref) / stdev for proficiency assessment.

The z score is commonly used, but it does not take measurement uncertainty in to account which can lead to "false" failures. For instance - some instruments like calipers have a resolution of 0.0005". If you have a value one digit of resolution different than the "assigned" value your z score may be magnitudes greater than is allowed. I would analyze any z score results that are out of tolerance.

Here is a great link that helped me along the way:
https://www.eurachem.org/images/stories/Guides/pdf/Eurachem_PT_Guide_2011.pdf
 
Thread starter Similar threads Forum Replies Date
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
D ISO 17025 5.10.2(e) - Test Method ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
Q Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
K ISO 17025 Facilities and environmental conditions SOP ISO 17025 related Discussions 6
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 5
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 2
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 14
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4

Similar threads

Top Bottom