ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements
L
LClaeys
My company asked me to put together a quality manual according to ISO17025.
I'm trying to understand the validation of methods and what a method is.
Example: Customer send us a voltmeter. I will calibrate this voltmeter using manufacturer's specifications. I write a procedure in Met/Cal that connect a calibrator to the voltmeter and test a certain amount of test points.
At this point I just use my procedure and calibrate the unit, any errors of the use of standards or the procedure I fix along the way since I need to use the unit to see if my standards and procedure is correct.
According to 5.4.4 you need to validate a method before use?
So is my method: calibration according to manufacturer's specifications?
And how do I validate my method?
dwperron
This is one of those 17025 paragraphs that you can conduct a seminar on.
First, a "method" is the procedure you will use to calibrate the instrument. The earlier sections of 5.4.X talk about choosing an appropriate method, or procedure, that is adequate for your needs. There is an emphasis on using "accepted" procedures, but the lab can develop their own procedures. The notes after section 5.4.4 give guidance on what should be included in the procedure. And yes, you must have a documented and controlled procedure so that the calibration can be duplicated. Your Met/Cal procedure will satisfy this requirement, but you must document that this is the procedure to be used, exercise revision control, etc.
Concerning validation, 5.4.5.2 says you shall (must) validate "non-standard methods, laboratory-designed / developed methods, standard methods used outside their intended scope, and amplifications and modifications of standard methods". That means that if you use one of the procedures that is taken from a reputable organization (manufacturer procedure, GIDEP, ASTM, AMSE, etc.) and use it as written you do not need to validate it. If you deviate from the procedure (delete test points, add test points, substitute equipment, etc.) you must validate that your changes still will provide valid results for your intended uses. For instance, if you delete tests from the original you must show that this will still be valid for your uses. If you substitute equipment you must show the equipment you are using is adequate for the job. If you add or modify tests you must show how they are adequate to meet your purposes. In your case, since you are creating your own Met/Cal procedure, it must be validated.
The validation of the procedure is "the provision of objective evidence" that the method works, and as such must be documented and retained as a record. The Note 2 of 5.4.5.2 spells out the validation techniques to be used. In your case here the simple way would be to perform the same calibration steps manually and compare your results to the Met/Cal results. In other cases you can use different standards to see if you get the same results, show how this same method works for similar instruments, or some other method that works for you to show that you get consistent and adequate results using your method. Be sure to record your validation results as evidence that you performed the validation.
This isn't really as bad as it looks: the last section of 5.4.5.2 Note 2 says that your validation technique can include "scientific understanding" and "practical experience", so if you can demonstrate that you know what you are doing, that you understand the process and the risks to the measurements involved, know your measurement requirements, and that you have documented your validation you should be all set.
Hershal
Metrologist-Auditor
Very simple. Manufacturer states certain things. Now, in your internal procedure, refer to the manufacturer procedure. That takes care of the validation part.
There is an exception, where you may need different data points. Those specific points you must validate. Submit a plan to management listing points or methods and how results may be verified (e.g., another lab). Provide results to management, they sign off, validated
J Recommendations on calibration interval (frequency) ISO/IEC 17025:2005 para 5.10.4.4 ISO 17025 related Discussions 5
K ISO 17025 Query Relating to Monitoring of Equipment ISO 17025 related Discussions 2
A Help required in establishing Laboratory Information Management System as per ISO 17025, Cl.7.11.2 ISO 17025 related Discussions 0
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
V ISO 17025 Accreditation Consulting Needed ISO 17025 related Discussions 1
D Is Good Laboratory Practice (GLP) Required in ISO 17025? ISO 17025 related Discussions 4
W Looking for IATF 16949 (and ISO 17025) QMS software Suggestions Quality Tools, Improvement and Analysis 8
R Process Maps for ISO 17025 ISO 17025 related Discussions 2
K ISO 17025 Method Validation and Verification for Test Lab ISO 17025 related Discussions 5
R ISO 17025 Lead Assessor Training ISO 17025 related Discussions 0
J ISO 17025 Calibration Other Medical Device Related Standards 2
E ISO 17025 Key Performance Indicators (KPIs) ISO 17025 related Discussions 5
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
D Training for ISO 17025 ISO 17025 related Discussions 2
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
Quality Specialist Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
K ISO 17025 Facilities and environmental conditions SOP ISO 17025 related Discussions 6
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 4
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 14
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3