ISO 17025 - Please share a sample procedure for monitoring the validity of results
dwperron
This is now a two part process.
Part one is the internal process. How can you be sure you are getting valid results? In 7.7.1 they give a long list of possible solutions, choose whatever ones work best for your lab.
Most labs already have a cross check program, where you bump one one of your standards against another (5520A against a 3458A) and record (and chart the results, assessors really like this!) the results. You can see when the results drift, and can act to determine and fix problems while they are still small. Record your results, document your methods.
You can perform the same calibration with different instruments to see if you get the same results. - Document the records.
You send something out for calibration? See if you can duplicate those results when it gets back.
Part two is proficiency testing. That pretty much is the same as it used to be.
You need a program where you do proficiency testing on each discipline in your scope of accreditation, or you get involved in interlaboratory comparisons (round robins). The idea is to compare your results with somebody else's.
All this needs to be documented, and your accreditation body will have their own requirements as to how this gets done, so check with them.
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 0
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
B ISO 17025 8.5 Actions to address risks and opportunities ISO 17025 related Discussions 7