ISO 17025 Query Relating to Monitoring of Equipment

klyoyo

Registered
Hi All,

We are in the early stages of becoming accredited and I have a query relating to monitoring our equipment. For example, our incubators are calibrated and mapped yearly, and the temperature is displayed on the outside panel of the incubator. We are recording the temperature on the display twice daily, am and pm. Is this sufficient, or do we need another thermometer or temperature logger?

Thanks!
 

dwperron

Trusted Information Resource
Is this sufficient, or do we need another thermometer or temperature logger?

Your organization is the judge as to what is "sufficient" based on the risk to your products and processes.

Personally, I would have trouble with your incubator monitoring. The annual calibration is a token snapshot of performance - it does nothing to tell you how the incubator actually performed over that year. Twice daily logging is a big improvement, but what if someone leaves the door open for a period of time in between those readings? How could that affect your product?

I am a big fan of dataloggers with alarms. They provide an actual record of the incubator temperature, and the alarms alert you when you have a problem.
 

Seanmhc

Registered
Is this sufficient, or do we need another thermometer or temperature logger?

Your organization is the judge as to what is "sufficient" based on the risk to your products and processes.

Personally, I would have trouble with your incubator monitoring. The annual calibration is a token snapshot of performance - it does nothing to tell you how the incubator actually performed over that year. Twice daily logging is a big improvement, but what if someone leaves the door open for a period of time in between those readings? How could that affect your product?

I am a big fan of dataloggers with alarms. They provide an actual record of the incubator temperature, and the alarms alert you when you have a problem.

I agree with you, dwperron.

I would recommend justifying any risk to product or process due to any failure modes identified that might cause a drop/raise in temperature outside of specification. This should be covered when installing/mapping the equipment for instance, how quickly does the temperature drop once the door has been opened, or if there is a power outage? and how long does it take to ramp back up? Is this accounted for in your incubation timings or any validated process.

A form of temperature monitoring device in your hottest/coolest locations indicating during your mapping process is a good way forward, with the ability to indicate min/max/current temps.

One thing that is also worth noting is the medium in which the items you are incubating are stored in. A temperature logger placed in the incubator will only be insulated by the air around it and itself. If the product you are incubating is in a liquid form, it is good practice to have a logger stored in liquid to monitor changes to this temperature.
 

VKF1923

Registered
Hi All,

We are in the early stages of becoming accredited and I have a query relating to monitoring our equipment. For example, our incubators are calibrated and mapped yearly, and the temperature is displayed on the outside panel of the incubator. We are recording the temperature on the display twice daily, am and pm. Is this sufficient, or do we need another thermometer or temperature logger?

Thanks!


We were doing this in the lab I work in and it wasn't enough.
What we ended up doing, was buying an Easycloud thermometer. It allows us to monitor incubator temps 24 hours a day, and we print the report once monthly. That seemed to pacify the accreditation board.
 
Thread starter Similar threads Forum Replies Date
G ISO 17025.2017 Clause 7.8.4.3 Certificate of Calibration Label ISO 17025 related Discussions 5
GStough ISO 17025 Lead Auditor Training - Punyam Academy ISO 17025 related Discussions 6
A Help required in establishing Laboratory Information Management System as per ISO 17025, Cl.7.11.2 ISO 17025 related Discussions 0
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
V ISO 17025 Accreditation Consulting Needed ISO 17025 related Discussions 1
D Is Good Laboratory Practice (GLP) Required in ISO 17025? ISO 17025 related Discussions 4
W Looking for IATF 16949 (and ISO 17025) QMS software Suggestions Quality Tools, Improvement and Analysis 8
R Process Maps for ISO 17025 ISO 17025 related Discussions 2
K ISO 17025 Method Validation and Verification for Test Lab ISO 17025 related Discussions 5
R ISO 17025 Lead Assessor Training ISO 17025 related Discussions 0
J ISO 17025 Calibration Other Medical Device Related Standards 2
E ISO 17025 Key Performance Indicators (KPIs) ISO 17025 related Discussions 5
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
D Training for ISO 17025 ISO 17025 related Discussions 2
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
Quality Specialist Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
K ISO 17025 Facilities and environmental conditions SOP ISO 17025 related Discussions 6
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 4
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 14
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3

Similar threads

Top Bottom