ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications

#1
Dear all,

could you give me advice how to satisfy requirements for personnel competences if it is applied for testing laboratory - for Non destructive testing.
My question is: Can department Manager (person responsible for the laboratory activities) be person who is not certified as NDT Level 3?
How does it comply with EA-4/15 G:2015 requirements?
My company has certified NDT level 3 personnel for every accredited method, but once again, does department manager who will lead department shall be also level 3?

Thank you in advance for help.
 
Elsmar Forum Sponsor

Gus

Involved In Discussions
#2
From what i can tell (from a quick google search on what is NDT and EA-4/15 G:2015) you should'nt have a problem as long as:

your department manager is not responsible for reviewing (technical) test reports or technical employee qualification reviews, i think it matters most that he/she is qualified in 17025 if he is approving test reports.

as long as you can evidence that your department manager is competent to perform whatever activities he/she is directly involved in performing regarding accredited requirements then you should be fine, he/she certainly doesn't have to be a technical expert if he doesn't perform technical work (reviewing technical work is also technical work IMO), but your management system has to reflect this wherever you have stated authorized roles and responsibilities.

that's my 2 cents on that
 
Thread starter Similar threads Forum Replies Date
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
G ISO 17025 - Requirement to Monitor Environment ISO 17025 related Discussions 10
rob73 Requirement for independent certification of device using an ISO 17025 laboratory EU Medical Device Regulations 2
Sean Kelley What is the ISO 17025 Certified Laboratory Requirement for External Calibration? ISO 17025 related Discussions 8
R Implementation of ISO/IEC 17025 Requirement for our Test Laboratory ISO 17025 related Discussions 2
M Specific Requirement for Calibration - ISO 17025 Clause 5.6.2 Interpretations ISO 17025 related Discussions 15
D Scales calibrated by an ISO 17025 accredited organization? TS 16949 requirement? ISO 17025 related Discussions 44
R Is passing 'Measurement Audit' a requirement for ISO 17025 Acrreditation? ISO 17025 related Discussions 3
Sidney Vianna How to audit this ISO 17025 requirement? ISO 17025 related Discussions 7
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 3
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2

Similar threads