We are a calibration lab. accredited since 2013 based on 17025 (2005), and now we intend to expand the scope to the on-site calibrations. I wonder if the mobile laboratory can be presented to the assessors for the demonstration for the on-site scope? Also, Could someone explain the standard requirements of on-site calibration.
dwperron
First, yes you can, and you will have to, demonstrate your competence at performing on site calibration service to the assessor.
The requirements for on site calibrations are the same as for in lab services, with a few special cases:
6.3.5 When the laboratory performs laboratory activities at sites or facilities outside its permanent
control, it shall ensure that the requirements related to facilities and environmental conditions of this document are met.
For this most labs will define the environmental conditions that are acceptable for on site work, and often have a disclaimer that names responsible parties who can authorize on site work if the conditions deviate. This also means recording the pertinent environmental data.
6.4.3 The laboratory shall have a procedure for handling, transport, storage, use and planned
maintenance of equipment in order to ensure proper functioning and to prevent contamination or deterioration.
You will need a procedure for how you handle transporting equipment from the lab to the on site locations.
6.4.4 The laboratory shall verify that equipment conforms to specified requirements before being
placed or returned into service.
You will need to verify that your equipment survived the trip, and is operational. Keep records of what you checked and what equipment you used for the checks. You will need to do this again when you return to the lab.
Most accrediting bodies also have internal policies for on site calibrations. You will definitely check with your accreditation body. I am familiar with A2LA's policies, they are in their R104 Field Calibration Requirements document. Those labs will also have a disclaimer statement on their scope of accreditation, such as:
Field calibration service is available for this calibration and this laboratory meets A2LA R104 – General Requirements: Accreditation of Field Testing and Field Calibration Laboratories for these calibrations. Please note the actual measurement uncertainties achievable on a customer's site can normally be expected to be larger than the CMC found on the A2LA Scope. Allowance must be made for aspects such as the environment at the place of calibration and for other possible adverse effects such as those caused by transportation of the calibration equipment. The usual allowance for the actual uncertainty introduced by the item being calibrated, (e.g. resolution) must also be considered and this, on its own, could result in the actual measurement uncertainty achievable on a customer’s site being larger than the CMC.
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
T ISO 17025 Section 5.9.1 - Assuring Quality of Tests - Requirements ISO 17025 related Discussions 4
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
B ISO 17025 Accredited Certificate Requirements ISO 17025 related Discussions 2
T Is the use of Control Charts Mandatory or Optional? ISO 17025 Requirements ISO 17025 related Discussions 6
N ISO 17025 Requirements for Environmental Conditions for In-House Calibration ISO 17025 related Discussions 5
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
T ISO 17025 Clause 5.4.2 - Calibration and Test Method Requirements ISO 17025 related Discussions 6
A Training Requirements for ISO 17025 Internal Auditor ISO 17025 related Discussions 19
D ISO 17025 and Calibration Laboratory Risk Management Requirements ISO 17025 related Discussions 1
M ISO 17025 and Design/Process Control Requirements ISO 17025 related Discussions 2
S ISO 17025 requirements not enough to prove NIST Traceability? ISO 17025 related Discussions 10
P NABL ISO/IEC 17025 Manual Requirements ISO 17025 related Discussions 3
N ISO 17025 External Calibration Laboratory Requirements within ISO/TS 16949 ISO 17025 related Discussions 28
S ISO 17025 Validation Requirements - "calibration using reference standards or..." ISO 17025 related Discussions 5
A How to comply with ISO/IEC 17025 Laboratory Requirements ISO 17025 related Discussions 7
V CMM Inspection Layouts vs ISO 17025 Clause 7.6.3.2 Requirements ISO 17025 related Discussions 5
G ISO 17025 definitions - Requirements for document management system ISO 17025 related Discussions 5
Hershal Accreditation to ISO 17025:2005 meets ISO 9001:2008 QMS requirements ISO 17025 related Discussions 5
B Pin Gage Calibration Procedure - ISO 17025:2005 accreditation requirements? ISO 17025 related Discussions 8
Hershal What about...? Meeting the requirements of ISO 17025, ISO 17020, and ISO Guide 65 ISO 17025 related Discussions 12
M 21 CFR Part 11 Control of Data Requirements in light of ISO 17025 ISO 17025 related Discussions 8
B ISO 17025 MU Requirements - Construction Material Testing Lab Uncertainty Budgets ISO 17025 related Discussions 10
A Training requirements for ISO 17025? ISO 17025 related Discussions 5
S Trying to find the actual ISO 17025:2005 standard requirements ISO 17025 related Discussions 6
E To conduct ISO 17025 Internal Audit - Auditor Requirements ISO 17025 related Discussions 10
gard2372 A2LA & ISO/IEC 17025:1999 vs. European Requirements ISO 17025 related Discussions 7
T ISO 17025 and ISO/TS-16949 Clause 7.6.3.1 Requirements - In-House Laboratory ISO 17025 related Discussions 2
R Guide 25 vs. ISO17025 - Substantially more requirements in ISO 17025 ISO 17025 related Discussions 12
Marc ISO 17025 & Guide 25 are being rewritten to include ISO 9001 requirements ISO 17025 related Discussions 8
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 2
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 6
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 1
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1