Has anyone ran into this? How did you answer it? I did some calibration for a new customer, and they sent it back saying they need NIST test numbers. I asked where they were getting that from, since I thought we stopped using them in the late 90's, and this is what they sent me:
Our nuclear requirement comes from ASME NQA-1-2008, Requirement 12, Paragraph 301, “Calibration shall be against and traceable to certified equipment or reference standards having known valid relationships to nationally recognized standards, or to international standards known to be equivalent to and verified against corresponding nationally recognized standards.”
The Nuclear Regulatory Commission (NRC) compared required nuclear requirements with 17025. The reviewers found that 17025 addresses all but two of the nuclear administrative requirements for calibration. The following two requirements must be included in purchase orders to commercial-grade calibration suppliers.
1. Calibration report/certificate shall include identification of the laboratory equipment/standards used, including NIST traceability
2. Calibration report/certificate shall include as-found and as-left data. (17025 requires that as-found data be reported only if the item requires adjustment or repair.)
To meet the above requirement NIST numbers shall be documented or the calibration is found deficient and data taken with the equipment is suspect until NIST traceability can be positively confirmed or data is re-taken with equipment having correctly documented traceability.
We are an ISO17025 lab and this calibration was within our scope...
Our nuclear requirement comes from ASME NQA-1-2008, Requirement 12, Paragraph 301, “Calibration shall be against and traceable to certified equipment or reference standards having known valid relationships to nationally recognized standards, or to international standards known to be equivalent to and verified against corresponding nationally recognized standards.”
The Nuclear Regulatory Commission (NRC) compared required nuclear requirements with 17025. The reviewers found that 17025 addresses all but two of the nuclear administrative requirements for calibration. The following two requirements must be included in purchase orders to commercial-grade calibration suppliers.
1. Calibration report/certificate shall include identification of the laboratory equipment/standards used, including NIST traceability
2. Calibration report/certificate shall include as-found and as-left data. (17025 requires that as-found data be reported only if the item requires adjustment or repair.)
To meet the above requirement NIST numbers shall be documented or the calibration is found deficient and data taken with the equipment is suspect until NIST traceability can be positively confirmed or data is re-taken with equipment having correctly documented traceability.
We are an ISO17025 lab and this calibration was within our scope...