ISO 17025 section 4.9 - Control of Nonconforming Testing and Calibration Work
remedina
I have experience with ISO 9001 but new to ISO 17025. I am used to concept of nonconforming product in ISO 9001, but when it comes to nonconforming work I am a little confused. Does this section of the ISO 17025 calls for the control of nonconformamces in performing the test or calibration work, like if the work is done improperly? I get confused when I think of an item subject to test. If the test is done properly and the item does not pass the test, this is not a nonconformace as defined in this 4.9 section, right? Any help will be appreciated.
Hershal
Metrologist-Auditor
You are exactly correct.
In particular this applies if the lab discovers the non-conforming work after the work is complete, and hopefully still before it reaches the customer. If you experience this then employ the corrective action protocols.
If the test or calibration is completed correctly and the item does not pass, then it is NOT non-conforming work, it is a failure of the unit under test.
One other point, this is one of the MOST difficult Clauses in the Standard.
Z
zamclachia
4.9 states 'when any aspect of its testing and/or calibration work, or the results of this work, do not conform to its own procedures or the agreed
requirements of the customer'
If I look at the typical work flow in the lab a non conformance can be raised at any stage of the process depending on what has been written in your procedures ( Or missing from your procedures). It may be agreed that the customer is notified once the sample is logged in. If this does not happen, a non conformance (NC) needs to be raised. If the sample integrity is compromised due to poor storage conditions - NC. If your procedures state that you shall run a reference material with the analytical run and the results shall be in specified limits. If either condition is not respected -NC. Anything that could negatively affect the management or technical system results in a NC. The purpose of a NC is to improve the quality system regardless of where the issue may be. :2cents:
L
lucasso
How is this requirement to be interpreted:
"4.9.1
a) <...> and actions (including halting of work and withholding <...>) are defined and taken when nonconforming work is identified."
Does this mean that
a) actions should be defined AFTER nonconformity is identified and, say, written up in the Nonconforming work protocol,
b) or that actions should be defined and writen up in Quality Manual in advance in case the nonconformity happens?
If i read the English version of the standard it kind of implies the a) version, but if i read the translation of the standard to my native language it kinda suggests the b) version.
Which is it?
Hershal
Metrologist-Auditor
It means that the policy and procedure have to define at least general steps to be taken in advance and people trained on the procedure and steps (keep records of the training).
It also means that if non-conforming work is found then the steps must be implemented right away, and if additional specific steps are necessary which is possible, those can be identified at the time and documented.
Also, the policy/procedure MUST SPECIFICALLY identify the person(s) having authority to authorize resumption of work.
Hope this helps.
T ISO 17025 Section 5.9.1 - Assuring Quality of Tests - Requirements ISO 17025 related Discussions 4
S Reference to original Calibration Certificate - ISO 17025 Section 5.10.9 ISO 17025 related Discussions 4
C Accredited Calibration Certificate Content - ISO 17025 Section 5.10.4.2 Question ISO 17025 related Discussions 7
R ISO 17025 section 4.2.2 and Quality Policy Content ISO 17025 related Discussions 4
Ajit Basrur How to accomplish section 5.9.2 of ISO 17025 ? ISO 17025 related Discussions 10
K What is Section 5.4.5.2 of ISO / IEC 17025 about? ISO 17025 related Discussions 1
Hershal PT/ILC - ISO 17025 section 5.9 - Proficiency Test/Inter-Laboratory Comparisons ISO 17025 related Discussions 27
Marc ISO 17025 Section 5.4.7.2.a - Software Validation Concern ISO 17025 related Discussions 0
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1