ISO 17025 sections state "Shall have a Policy and Procedure"
A
ambrosedierdre
#1
Throughout 17025, sections state 'shall have a policy and procedure'. Does this mean that you need both a policy, AND a procedure. eg, complaints 4.8.
Elsmar Forum Sponsor
Jen Kirley
Quality and Auditing Expert
Staff member
Admin
#3
Good day ambrosediere,
The short answer is yes.
But that doesn't mean you must have a tall stack of documents. A2LA's Explanations for the ISO/IEC 10725 Requirements says:
Must the quality manual include all of the laboratory?s procedures? 04/30/2005, 4.2.2
In many clauses, the standard will require a policy and procedure. The policies related to quality must be defined in the quality manual (see 2011 explanation of Section 4.2.2 regarding this). However, the procedure may be found in other documents as long as it is referenced from the quality manual (per Section 4.2.5). If the procedure is included in the quality manual, it is considered sufficient to satisfy both the policy and procedure elements.
The point is this: a policy statement is a higher level directive than a procedure, which tells people how the process works. In 17025 we need both, but they don't need to be separated.
I hope this helps!
Thread starter Similar threads Forum Replies Date
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
Similar threads
Top Bottom