ISO 17025 Validation Requirements - "calibration using reference standards or..."
sokolasty
Hello
I'm reading chapter "Validation" of ISO 17025 now. One of the proposed method of validation is:
"calibration using reference standards or reference materials".
I don't understand it - well, I get the words, but I don't see the meaning. Can you help?
BradM
Re: Validation way acc. to 17025: "calibration using reference standards or..."
I assume you mean the words in quote, in your quote.:) Is that correct?
Basically it means using standards traceable to N.I.S.T. or an equivalent nationally recognized body. They would be higher order standards than the devices you are calibrating within your lab.
The reference materials refers to chemical compositions and such that are not necessarily traceable to a national standard, but when made using a certain process, are "considered" standards.
Did that answer your question?:)
sokolasty
Thanks Brad, I get your point, but it seems my question was not clear enough.
This is a quote from ISO 17025 (determination of performance is a validation here):
The techniques used for the determination of the performance of a method should be one of, or a combination of, the following:
— calibration using reference standards or reference materials;
— comparison of results achieved with other methods;
— inter-laboratory comparisons;
— systematic assessment of the factors influencing the result;
— assessment of the uncertainty of the results based on scientific
understanding of the theoretical principles of the method and practical
experience.
I don't understand, how can I validate a calibration method by a calibration.
BradM
Actually, you pose a very good question. The statement in the standard does appear a bit circular in logic.
This is just kind of my take on this. I would look at why this section is here.
Say you have a request to calibrate a new instrument. You write up a procedure, and start calibrating them. Strangely, you begin to have a high # of failures in the as-found category. Too, the customers begin complaining their instrument does not seem to operate correctly. Where is the problem?
You question the techs, you at the standard, etc. to try to find the issue. Ironically, after spending a lot of time and effort, you realize the procedure was wrong. There were a couple of steps left out that really made a difference.
The intent of the validation requirement quoted is to validate the process.
  • Does it yield the results it should?
  • Have you identified all the variables involved?
  • Does workstation A yield similar results as workstation B?
  • Do days of the week affect the results?
Perform the calibration (with all the data, of course) then send it to another lab (maybe a competitor). Are the results fairly similar? How do their uncertainty compare to yours?
I feel I am rambling:D, so I will stop.:D
Just remember you need to verify the entire process (and document the verification) prior to considering it acceptable for performing calibrations. Once it is verified, all involved will have more confidence in it.:)
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
D pH Validation for Water (ISO 17025) ISO 17025 related Discussions 0
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
T ISO 17025 Validation Procedure - Need Example ISO 17025 related Discussions 11
Marc ISO 17025 Section 5.4.7.2.a - Software Validation Concern ISO 17025 related Discussions 0
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6