ISO 9001:2015 - Record requirements for out of calibration tool

Ninja

Looking for Reality
Staff member
Super Moderator
#11
Howdy, ... , Totally on board with Sidney (a rare enough thing and nice to see).
it was a torque screwdriver. ... and doesn't impact the quality of the product.
We had a tool that the calibration house reported as out of calibration, so we took it out of service
Why would you send it out for calibration then? Do you send your coffeemaker out for temperature calibration?
On the same line...why did you take it out of service?
If you don't care about the torque, at least it's a screwdriver, right? ...and you don't just need the screws in...

why send it to an outside lab for calibration?
Yup...that...thanks Sidney.

I would mark the thing as "calibration not required"...and then (just as a CYA) would put a statement in the calibration record as to WHY calibration was not required. Soup to nuts...15 minutes.
 
Elsmar Forum Sponsor

Big Jim

Super Moderator
#12
We are preparing for our annual surveillance audit, and in reviewing prior CARs(so I can ensure we are prepared to show effectiveness) I came across a CAR I'm not sure why we were written up for. Kind of late as it's said and done, but just for my understanding... We had a tool that the calibration house reported as out of calibration, so we took it out of service but we had no records of who decided what and took what actions so we were written up. Specifically - ISO 7.1.5 requires that "The organization shall determine if the validity of previous measurement results has been adversely affected when measuring equipment is found to be unfit for its intended purpose, and shall take appropriate action as necessary.". To make sure we're in compliance today, I read ISO section 7.1.5 as well as our internal procedures and I don't see where it specifically says records must be maintained. Am I misunderstanding or missing something?

Interesting question. Especially since it was a requirement in ISO 9001:2008 to keep a record of the investigation. I usually recommended writing up the investigation using the corrective action system so you didn't need to create another recording method.

Since it is clear that ISO 9001:2015 doesn't mandate that you keep a record of the investigation, keep in mind that an auditor can still pursue if you are doing it or not by interview. Also keep in mind that if you do determine that that you shipped nonconforming product you will need to write a nonconformance.

So it may still be a good idea to document the investigation.
 

somashekar

Staff member
Super Moderator
#13
Thanks guys. I like how pragmatic the people on this forum are, or said another way how you keep the perspective of quality as well as what makes good business sense.

In our specific business, we sell relatively low cost products and if we were to go overboard we wouldn't be making money. In regards to the finding we were written up for, it was a torque screwdriver. It is used for consistency and doesn't impact the quality of the product. To clarify a bit more if screws were falling out of products customers wouldn't be happy, but it's not a situation where if the screw isn't tightened to x inch / lbs the product won't work. I don't see anywhere where it says a record is required, so was this a valid CAR - where is the requirement that we violated?
"Its used for consistency and doesn't impact the quality of the product"
You have taken all the steps to have a calibrated torque screwdriver for consistency purpose which is a very valid quality characteristic, there by giving assurance for your low cost product.
Having done this much, have you mentioned the reference of this torque screwdriver in your final inspection / test records., even if you have only one such torque screwdriver ?
This habit of having the reference of the unique instrument that is used in your setup, appearing in your inspection records., will help in the valid traceability of such products down the line as well as in the field depending upon your identification / traceability procedure in place.
When the torque screwdriver was taken out of service being reported as out of calibration, you could respond with confidence that the previous batches were traced based on your inspection records, and verified to have been good, A record of such verification would not be required based upon your risk based thinking.
I see it all about how your calibration process is made to interact with the identification and traceability process and the records process you have established.
 

Jim Wynne

Staff member
Admin
#14
In regards to the finding we were written up for, it was a torque screwdriver. It is used for consistency and doesn't impact the quality of the product.
What does "...used for consistency..." mean? Why bother if it doesn't mean anything? Is there a torque specification? If not, then how do you know if "consistency" has been achieved?
 

embedded

Involved In Discussions
#15
This is a really good discussion. I'll follow up with the design team to see if we can move away from these torque screwdrivers for this application, we did this for a number of other applications but not entirely sure why they didn't want to move away for this particular application(note these are tiny screws, picture the screws that hold your phone together, pretty much the same application here).

Here is where I'm really confused though. Doesn't the ISO specification or our procedures explicitly have to say a record is required/retained for this to be a finding? I'm also a new internal auditor and my understanding is a finding is a non-conformance that is situation where a requirement was not met. I don't see the requirement stating a record is retained so I don't understand how this could be a non-conformance / finding. Can you help me to better understand?
 

Jim Wynne

Staff member
Admin
#16
This is a really good discussion. I'll follow up with the design team to see if we can move away from these torque screwdrivers for this application, we did this for a number of other applications but not entirely sure why they didn't want to move away for this particular application(note these are tiny screws, picture the screws that hold your phone together, pretty much the same application here).

Here is where I'm really confused though. Doesn't the ISO specification or our procedures explicitly have to say a record is required/retained for this to be a finding? I'm also a new internal auditor and my understanding is a finding is a non-conformance that is situation where a requirement was not met. I don't see the requirement stating a record is retained so I don't understand how this could be a non-conformance / finding. Can you help me to better understand?
In order to address the question directly, please give us the exact text of the whole NC statement.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#17
I don't see the requirement stating a record is retained so I don't understand how this could be a non-conformance / finding. Can you help me to better understand?
As you insist on this aspect, let me offer the following takes:

1. As already mentioned previously in this discussion, IRRESPECTIVE, of a record being explicitly required or not, audits are based on OBJECTIVE EVIDENCE. So, as an auditor, internal or external, your system has to establish evidence that the out of tolerance condition has been identified, evaluated, analyzed and dispositioned by a competent and authorized individual, as the standard requires.

2. I can EASILY make a case that such scenario falls under 8.7.2 - nonconforming output, which mandates records to be retained. The situation where parts were produced with a tool that is later found to be out of specification is CLEARLY a nonconforming output.

Voilá. :naughty:
 

Big Jim

Super Moderator
#18
As you insist on this aspect, let me offer the following takes:

1. As already mentioned previously in this discussion, IRRESPECTIVE, of a record being explicitly required or not, audits are based on OBJECTIVE EVIDENCE. So, as an auditor, internal or external, your system has to establish evidence that the out of tolerance condition has been identified, evaluated, analyzed and dispositioned by a competent and authorized individual, as the standard requires.

2. I can EASILY make a case that such scenario falls under 8.7.2 - nonconforming output, which mandates records to be retained. The situation where parts were produced with a tool that is later found to be out of specification is CLEARLY a nonconforming output.

Voilá. :naughty:

1. Objective evidence isn't limited to documents and records. It also includes what is learned from interview and observation. If the standard does not require a record be kept and auditor CANNOT insist on seeing one.

2. At this point it is NOT clear the the out of calibration tool resulted in nonconforming product. An engineering study probably needs to be undertaken to determine if the screws are too tight or too loose. A study that might be little more than taking one of the assemblies in question and taking a screwdriver to the screws in question.

Your point that 8.7.2 could be applied in a situation like this is well taken, but it is too soon to make a conclusion from the evidence submitted so far.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#19
. At this point it is NOT clear the the out of calibration tool resulted in nonconforming product. An engineering study probably needs to be undertaken to determine if the screws are too tight or too loose. A study that might be little more than taking one of the assemblies in question and taking a screwdriver to the screws in question.
Irrespective of the conformity status of the product, the situation is WITHOUT A QUESTION, a nonconforming output. The output being the assembly process. The torque requirements were off spec. So, nonconforming output has to be addressed and record is mandatory.
 

Big Jim

Super Moderator
#20
Irrespective of the conformity status of the product, the situation is WITHOUT A QUESTION, a nonconforming output. The output being the assembly process. The torque requirements were off spec. So, nonconforming output has to be addressed and record is mandatory.
How did you establish that there was a torque spec? The OP did not say so in any of his posts. Finding a torque tool is use doesn't establish that a torque spec is in play.
 
Thread starter Similar threads Forum Replies Date
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J Opportunity in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Sequence of ISO 9001:2015 Implementation Steps ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S ISO 9001:2015 Internal Auditing Internal Auditing 8
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
A What must be recorded? (ISO 9001:2015, subclause 10.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S ISO 9001:2015 vs 21 CFR Part 211 matrix Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
M ISO 9001:2015 and AS6081:2012 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Implementation ISO 9001: 2015 ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
eule del ayre Documented Information - Periodic Review of Documents? IATF 16949:2016 / ISO 9001:2015 IATF 16949 - Automotive Quality Systems Standard 34
J Audit Checklist for Integrated Management System for ISO 9001:2015, ISO 14001 & OHSAS18001 (IMS) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
lanley liao How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
T ISO 9001:2015 - Small Shop ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M ISO 9001:2015 case study sample ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Sample of Nonconformity report for ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Case study help as per ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
M Case study solution help required as per ISO 9001 : 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
F Logistics and IT clauses in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
V Quality Objectives - ISO 9001 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
Gun46 ISO 9001 : 2015 Lead Auditor Exam General Auditing Discussions 16
Q ISO 9001:2015 man days for surveillance audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
I ISO 9001:2015 Section 9.3.2 C1 "customer satisfaction and feedback from relevant interested parties" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
I ISO-9001:2015 and "Relevant Parties" Management Review Meetings and related Processes 2
I Master Document Access - ISO 9001:2015 clause 7.5.3 Document Control Systems, Procedures, Forms and Templates 5
qualprod Why new clause (7.1.6) in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
I Where can I beg, steal (just kidding of course) or borrow good training material on the ISO 9001:2015 standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M ISO 9001:2015 8.2.1 Contingency Plan required for small Business? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Help with Drawing Review - ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
V ISO 9001:2015 7.2 Competence ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 35
B ISO 9001:2015 Transition: Much Easier Than You Think! Misc. Quality Assurance and Business Systems Related Topics 0
A Process Documentation for ISO 9001:2015 Internal Auditing 2
C Migrate ISO 9001:2015 Certificate Registrars and Notified Bodies 4
J ISO 9001:2015 Small Operation Management Review General Auditing Discussions 6
E Three basic questions about ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C ISO 9001:2015 Monitoring and measuring resources. Application a service industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
W ISO 9001:2015 Management Review Input Template wanted ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
G Is ISO 9001:2015 certification worth it for a company that does only contract manufacturing? Quality Management System (QMS) Manuals 14
N ISO 9001:2015 Customer Complaints Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
O Performance Measurement ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Does ISO 9001-2015 have a requirement for manufacturing equipment to be numbered? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
V ISO 9001: 2015 Production drawing Control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
KCalaba21 ISO 9001:2015 6.2.2 E - "e) how the results will be evaluated." ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3

Similar threads

Top Bottom