ISO 9001 Section 7.6 - Customer has supplied torque specifications
Big Jim
#11
Oops!
After posting I realized two others have already provided nearly the same results.
according to clause no 7.6 ISO 9001:2008 the measuring equipments should be calibrated so, is it required to calibrate avometers and multimeters.
and if yes, what is meant by (where necessary ) which mentioned in that clause.
is it means it is not mandatory, waiting for your esteemed feedback
In context, what is stated in the standard is "Where necessary to ensure valid results . . .
If the instrument is needed for valid results, it needs to be calibrated. If it is not needed for valid results, it doesn't need to be calibrated. There are not very many circumstances that one would be using a measuring tool where valid results are not needed.
#12
Oops!
This is another misinterpretation of the wording of the clause. It doesn't say that calibration is necessary when the device is needed for valid results. It says that calibration is necessary when calibration is necessary to ensure valid results. It's an important distinction. All measurement devices (and measurement systems) are expected to provide valid results, but calibration isn't always necessary to ensure valid results.
#13
and, of course, (good, beam type) torque wrenches aren't actually 'measuring' torque, they are 'set' to a value, derived from an accurate device...:notme:
Big Jim
#14
I fail to see the distinction, at least on this application.
somashekar
#15
Assume the customer has supplied torque specs. I believe it is, others disagree. What do you think?
Customer provides torque spec so apply it in your product realization process, by a varification at your specified intervals, or prior to the use.
If customer has not provided, then you determine it and then apply it in your product realization process, by a varification at your specified intervals, or prior to the use.
cheap uncalibrated torque wrenches
Does not matter. Can it be verified against your national /international standards for the torque determined ?
Can it be safeguarded from adjustment to prevent any unintentional adjustment ?
Then I do not care if its cheap or has no any calibration.
If you cannot prove that it is varified, and show records of same, I will stop its use.
#16
What you wrote was "If the instrument is needed for valid results, it needs to be calibrated." That's not what the standard requires, however. It says "Where necessary to ensure valid results [calibration is required]."
In other words, the standard is saying that if lack of calibration can lead to invalid results, you must calibrate. This is fundamentally different from saying that if the device must produce valid results, it must be calibrated.
Big Jim
#17
I still fail to see the distinction, but if it makes you feel better you're welcome to believe so. Particularly in this application, the results is the same.
Can we move onto something more important?
I ISO 9001:2015 Section 9.3.2 C1 "customer satisfaction and feedback from relevant interested parties" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
D Monitoring and measuring resources - Example of how section ISO 9001 7.1.5 would apply to a manufacture of software ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Relevant internal and external communication - ISO 9001:2015 section 7.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Applicability of ISO 9001:2015 Section 8.5 - Engineering company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Q ISO 9001 Section 7.6 - Computer Software ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
K ISO 9001:2015 Section 8.5.3 - Property belonging to customers or external providers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
K ISO 9001 Section 7.3 Design and Development Question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
K Exemption from Section 7 clauses of ISO 9001 - Scientific Research Company Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
L ISO 9001 Section 6 - Human Resources ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R ISO 9001:1987 or 1994 - Wording of Purchasing in section 6.0 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
H ISO 9001 Section 7 when you don't manufacture anything? Service Industry Specific Topics 6
T Updating to ISO 9001:2008 - I need to add section 7.3 Design and Development ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
I Clarification of ISO 9001: 2008 Section 7.6 - Monitoring and measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
K No Formal Design Process - Audit Finding - Help with Section 7.3 of ISO 9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
G Can a testing laboratory exclude section 7.3 of ISO 9001? Quality Management System (QMS) Manuals 2
D ISO 9001:2000 Checklist for Auditing Personnel/HR section Internal Auditing 12
M Confused about measurement of software - Section 7.6 of ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Plug Gage Calibration Report does not include actual data - ISO 9001:2000 section 7.6 General Measurement Device and Calibration Topics 3
T Addressing ISO 9001 Section 7 in the Quality Manual when Design is Excluded Design and Development of Products and Processes 6
Q Training Records - Proof of "training" per ISO 9001 section 6.2.2 (b) Records and Data - Quality, Legal and Other Evidence 33
U ISO 9001:2000 Section 1.0 Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
S ISO 9001 Section 7 Requirements - Contract assembly ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
A Need help with section 7.5 of ISO 9001:2000 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D Differences in ISO 9001:2000 Section 8.3 vs. ISO13485:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
G Nuclear - ASME Section III N-Stamp - Requirements over and above AS9100 and ISO 9001 Various Other Specifications, Standards, and related Requirements 12
T ISO 9001:2000 Section 8 - Measurment Analysis and IMPROVEMENT requirement? Preventive Action and Continuous Improvement 1
Q Audit report template ISO 9001/14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
Q Process matrix examples of ISO 9001 & 14001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Q ISO 9001 - Reseller Exclusions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N ISO 9001 - Training business with fewer than 5 employees ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
J Opportunity in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
R Remote Audits for ISO 9001 (or any other standard) General Auditing Discussions 31
T Relationship between ISO 9001 and ISO – IEC BS EN 870079- 34 2020 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Sequence of ISO 9001:2015 Implementation Steps ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
qualprod Business Continuity Planning in ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P ISO 9001 certification with zero customers? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11