ONE OF MY CLIENT IS PROVIDING LABORATORY SERVICES IN THE AREAS OF AIR / WATER POLLUTION. THIS LAB RECEIVES SAMPLES FOR ANALYSIS AND IT PROVIDES BACK ANALYTICAL REPORTS.
WHERE REQUIRED, THIS LAB PROVIDES SUITABLE POLLUTION CONTROL MEASURES.
ADDITIONALLY, THIS LAB HAS ENGINEERING EXPERTISE ALSO.
THIS LAB IS INTERESTED IN GETTING A RECOGNITION THROUGH ISO CERTIFICATION. HOW DO THEY GO ABOUT : SHOULD THEY GO FOR ISO9002 OR ISO 9001 ?
WILL THIS CERTIFICATION HELP THEM GET A SPECIAL RECOGNITION FOR THIER LABORATORY.
THANKS
Dr P R MADHAVAN
 
Elsmar Forum Sponsor
D

Don Winton

#2
If by engineering, you mean design and development, they could go for either ISO 9001 or ISO 9002. If they are willing to have the design and development a part of the audit scope, do 9001. If not, do 9002.

Will the certification help them get special recognition? ISO 900x is a Quality Management System that is registered against a standard. The registration certifies that a Quality Management System exists and is in compliance with the standard. Only their customers or potential customers can determine if this is worthy of “special recognition.”

Regards,
Don
 
D

Don Winton

#3
Subject: iso 9001 - laboratory
Date: Sun, 14 Mar 1999 06:51:30 +0530
From: "Dr.MADHAVAN" <[email protected]>
To: [email protected]

thanks for the response

the client will go for iso9001 as he has full scope for design.

1. since there is no process involved , how do i give treatment to clause 4.9?

2. again, the laboratory which is serving the environment sector (air & water pollution), will it be expected that the procedures for the laboratory activity comply with provisions of IEC25 even though the lab may not get certified for IEC25 separately.

dr p r madhavan

since there is no process involved , how do i give treatment to clause 4.9
I would suggest that the procedures for 4.9 be prepared regardless. Remember, “the range and detail of the procedures that form part of the quality system depend on the complexity of the work, the methods used, and the skills and training needed by personnel involved in carrying out the activity.” When the question is asked, there is a procedure in place, whether it is in actual use or not.

...that the procedures for the laboratory activity comply with provisions of IEC25...
I would humbly suggest that this be considered separately from ISO 9001.

Regards,
Don
 
S

SteelWoman

#4
This is an old thread, but I found it through the search feature because I have a strange issue I need some help with. I have a company I deal with, a supplier, who has recently acquired a cold mill facility which includes a lab. The lab performs tests like tensiles, chemistry analysis, etc. The newly acquired facility is not yet certified to any standard. I know that with customer approval to do so (which we have) we can accept their testing results on their own material while they pursue certification. However, if I read 7.6.3.2 correctly if we use them to process someone else's material (another vendor's substrate) for us, and we want them to test the material, we ALSO have to have either them certified to a lab standard (not practical at this time) or a copy of their lab scope.

The PROBLEM is I'm running into a novice quality person at their facility who is inexplicably refusing to give me a lab scope document. I've explained that it's really just a list of the testing their lab is ALREADY doing - a document that would take her about 5 minutes to prepare from scratch, which they don't even have to do because they inherited the scope document from the previous owners. But she keeps coming back to me telling me why she is not "required" to have a lab scope! For whatever reason, she's turning this into a big pissing contest and I'm not going to be able to get a scope from them. Is customer approval to use them for testing, despite the lack of certificaton and lack of a lab scope document sufficient?

I also got the same response from her when I asked her to complete a Vendor Self Assessment - she told me they aren't certified so they shouldn't have to fill one out??? I'm starting to lean toward scheduling a full day audit over there of their new facility and their lab - if I can't get the simple documents completed I don't see where I have much choice but to perform an audit myself? Before you ask, this is not a vendor we can ditch - they provide a critical product.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#5
A lot of your question is dependent on YOUR certification/accreditation.

If you are registered to ISO 9K2K or to a sector-specific, then someone else may give a better answer, but I suspect you will have to do a good evaluation of them.

If your organization, or the part of your organization that uses this other lab, is accredited under ISO/IEC 17025, then you are in a higher level playing field and the rules are different - and much more involved. The comments you mention suggest that you need an accreditation of that lab ultimately, as you need a scope.

Feel free to contact me directly if 17025 is the driving issue and I may be able to help answer the questions a bit better.

Hershal
 

Cari Spears

Super Moderator
Staff member
Super Moderator
#6
SteelWoman said:
I also got the same response from her when I asked her to complete a Vendor Self Assessment - she told me they aren't certified so they shouldn't have to fill one out???
Sheesh! If she knew anything, she'd know that it's BECAUSE they aren't certified that people want these things.
 
D

D.Scott

#7
Did you try explaining that it isn't a requirement of their certification (which they don't have), it is a requirement of doing business with you. If their quality group won't provide what you need, get on their salesmen and have them put some pressure on. There doesn't seem to be much choice about scheduling an audit though. In my opinion you will need to do one anyway.

Dave
 
Thread starter Similar threads Forum Replies Date
A What is "a router" in ISO/TS 9002 Quality management systems ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Sidney Vianna Informational ISO/TS 9002 - Guidelines for the application of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
A Timeframe to transition to ISO 9001:2000 - ISO 9002:1994 certificate expires in 2007 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
S Tracking Down ISO 9002:1994 version ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Sidney Vianna ISO 9002:2000 - I did not know it existed - Not to mention QS-9000:2000 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
V Need to update to ISO 9001:2000, current registered to ISO 9002 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
J ISO 9002 - Small distributorship is beginning to pursue certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Audited against ISO 9002/3/1:1994 until December 2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
T New ISO 9002 Document Control Systems, Procedures, Forms and Templates 1
W ISO 9002 in small business that provides telecommunications service Service Industry Specific Topics 8
N Combine ISO 9002 and ISO 14001 - Could you please give me some suggestions? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M ISO 9002 Section 4.7 - Heat Treatment - Supplies and related services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
L ISO 9002 Implementation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Z Auditor Findings ISO 14001:2015 vs. 45001:2015 ISO 14001:2015 Specific Discussions 1
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Supplier Management ISO 13485: 2016- Which supplier needs to fill in a self assessment form? ISO 13485:2016 - Medical Device Quality Management Systems 6
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
G ISO 17023 2017-11 - Suggestions for good books ISO 17025 related Discussions 0
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M ISO 9001:2015 and AS6081:2012 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
C Implementation ISO 9001: 2015 ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
J Possible to get ISO 13485 certified with only OEM Product? ISO 13485:2016 - Medical Device Quality Management Systems 4
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 2
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 0
L Biological Assessment (ISO 10993-1) Other Medical Device Related Standards 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
eule del ayre Documented Information - Periodic Review of Documents? IATF 16949:2016 / ISO 9001:2015 IATF 16949 - Automotive Quality Systems Standard 34
J ISO 9001:2015, ISO 14001 & OHSAS18000 (IMS) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
G ISO 14001 - 6.1.3 Compliance Obligations ISO 14001:2015 Specific Discussions 1
D Rules for Paper Forms outside of an eQMS - 3 Questions (ISO 13485) Document Control Systems, Procedures, Forms and Templates 9
S Qualification question - ISO 13485 Reliability Analysis - Predictions, Testing and Standards 0
K ISO 13485 clause 8.5.2 'Any necessary CA shall be taken without undue delay' ISO 13485:2016 - Medical Device Quality Management Systems 10
Aymaneh ISO 11607-1: 2019 main changes Other Medical Device Related Standards 2
G National Structural Steel Specification 7th Edition - Do I now have to be audited against ISO 3843-3 as well as ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J How much to charge for helping a startup company with initial ISO 13485 certification? Consultants and Consulting 3
J ISO 13485 System 'soft start' - How to best reflect this in initial audits, management review minutes and other records? ISO 13485:2016 - Medical Device Quality Management Systems 3
L How to understand the clause 6 Planning of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
B ISO 8536-4 Contamination Index ISO 13485:2016 - Medical Device Quality Management Systems 0
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1
J Recommendations for online ISO 19011 training? Training - Internal, External, Online and Distance Learning 6
D ISO 13485 8.2.1 and 8.2.2 - Customer Feedback and Customer Complaints ISO 13485:2016 - Medical Device Quality Management Systems 5
S Requirements to obtain ISO 50001 Certification ISO 14001:2015 Specific Discussions 2
A ISO 11135:2014, B.1.4, BI resistance x product bioburden ISO 13485:2016 - Medical Device Quality Management Systems 6
Sravan Manchikanti How to interpret '8.3 Control of nonconforming product' for SaMD device while implementing ISO 13485 & MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 4

Similar threads

Top Bottom