ISO/IEC 17025 Sampling Plans - Sampling clauses not applicable for test labs?

E

el_quixiote

#1
Hi. We are starting a project to get our laboratories accredited under iso/iec 17025.

We already have QS-9000 certification since > 5 years.

Our labs are not responsible for sampling nor for designing sampling plans. These activities are deployed by operation people and by product engineering people respectively.

Besides, our labs are not responsible for the GO/NO GO decision. This decision is taken by QA people using a system that has the sampling results and the customer requierements for each product.

So, our labs are focused on determination of properties of each sample that is delivered to them.

My question is: If we declare in our Quality Manual that the sampling clauses are not applicable for our labs, would it be a nonconformancy ?

OMC

Thanks in advance....
 
Elsmar Forum Sponsor
T

tomvehoski

#2
You did not mention what type of sampling you are doing. Is it selecting x parts out of lot size y, or sampling chemicals, solutions, etc.

I don't think you can claim an exemption. If other departments write the sampling procedures, just adopt those as your laboratory procedures. Make sure they meet all of the requirements of 5.7. It may not be enough to say sample 10% or 10 ml from container. Do you need to make sure you get samples from throughout a production run? Does a liquid have to be mixed prior to sampling? There could me many other factors to address.

Tom
 
K

Ken K

#3
Section 5.7 of our manual simply states:

Sampling requirements of ISO17025 does not apply to the lab"

It has been audited and approved with no questions asked.
 
E

el_quixiote

#4
thanks you for your responses.

We produce steel flat products and our sampling procedures refers to chemical analysis and mechanical properties.

We follow ASTM standar sampling procedures or meet customer requierement if it is more specific.

We can write down some references about the sampling standars we are following. Question: Is an ISO/IEC 17025 auditor entitled to make judgements about the sampling plans we are using?
 
T

tomvehoski

#5
If you are using ASTM standard procedures, I would simply note what those are and use them as your procedure. There is no need to rewrite them, unless you feel they need clarification.

The auditor judgement will depend on the type of tests you are doing. If you customer says follow ASTM Standard # 12345, and it references a sampling plan, you must. If there is no requirement you should be able to use what is appropriate and generally accepted for the application.

Tom
 
Thread starter Similar threads Forum Replies Date
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
Douglas E. Purdy ISO/IEC 17025:2017 3rd Ed. Changes from 2nd Ed. ISO 17025 related Discussions 6
Douglas E. Purdy ISO/IEC 17025:2017 Clause 8 & Annex B ISO 17025 related Discussions 11
Ajit Basrur Informational ISO/IEC 17025:2017 Published - November 2017 ISO 17025 related Discussions 8
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
D ISO/IEC 17025 Implementation Workshop Ideas ISO 17025 related Discussions 2
Hershal ILAC is preparing to start the possible revision to ISO/IEC 17025 ISO 17025 related Discussions 2
J ISO/IEC 17025 - Which External Documents require control? ISO 17025 related Discussions 6
Z Force & Strain Calibration Uncertainty for ISO/IEC 17025 ISO 17025 related Discussions 2
R Implementation of ISO/IEC 17025 Requirement for our Test Laboratory ISO 17025 related Discussions 2
K Where to buy ISO/IEC 17025:2005 ISO 17025 related Discussions 4
G A2LA's ISO/IEC 17025 and Accreditation Training - Opinions from previous attendees ISO 17025 related Discussions 2
M Does ISO/IEC 17025:2005 require the Laboratory to have separate ISO 9001 Procedures ISO 17025 related Discussions 4
P NABL ISO/IEC 17025 Manual Requirements ISO 17025 related Discussions 3
QMMike TS 16949 - A2LA or ISO/IEC 17025 required for material testing? ISO 17025 related Discussions 6
N Measurement Uncertainty Explanation per ISO/IEC 17025:2005 clause 5.10.4.2 ISO 17025 related Discussions 3
G Is Agilent?s automated network analyzer verification ISO/IEC-17025 compliant? ISO 17025 related Discussions 4
R Example of ISO/IEC 17025 Implementation Project Plan ISO 17025 related Discussions 8
I Unapproved Accreditation Bodies offer ISO/IEC 17025 Accreditations ISO 17025 related Discussions 3
suildur Assuring the Quality of Test and Calibration Results - ISO IEC 17025 Clause 5.9 ISO 17025 related Discussions 13
Hershal Training in ISO/IEC 17025:2005 - May 2010 Washington, DC ISO 17025 related Discussions 3
A Compliance with ISO/ IEC 17025:2005 Clause 5.9 for Calibration Laboratories ISO 17025 related Discussions 6
A How to comply with ISO/IEC 17025 Laboratory Requirements ISO 17025 related Discussions 7
D Calibration Frequency of CMM as per ISO/IEC 17025 ISO 17025 related Discussions 11
Hershal Training in ISO/IEC 17025 and 17020 ISO 17025 related Discussions 0
V ISO/IEC 17025 Clause 4.4 - Review of requests, tenders and contracts ISO 17025 related Discussions 4
T Calibration Procedures - ISO/IEC 17025, Clause 5.5.9 - Outsourced Calibration ISO 17025 related Discussions 10
Hershal Update from MSC........ ISO/IEC 17025 - Challenges and Solutions ISO 17025 related Discussions 1
Hershal Reviving the discussion on ISO/IEC 17025 ISO 17025 related Discussions 4
Hershal Introduction to ANS/ISO/IEC 17025 and Z540 ISO 17025 related Discussions 2
Hershal ISO/IEC 17025:2005 Clause 4.1.5 and 4.1.6 ISO 17025 related Discussions 5
S Project Kick-Off - First visit agenda to kick off a ISO/IEC 17025:2005 ISO 17025 related Discussions 1
D Do Supplier's Laboratories Need to be Accredited to ISO/IEC 17025 ISO 17025 related Discussions 15
S Laboratory is accredited to ISO/IEC 17025 but that is not stated on the Cal Cert ISO 17025 related Discussions 7
B What Laboratories are the 'National Equivalent' of Accredited to ISO/IEC 17025? ISO 17025 related Discussions 3
Hershal ISO/IEC 17025 and ISO/IEC 17020 training ISO 17025 related Discussions 2
D Appropriate Equipment for ISO/IEC 17025:2005 Clause 5.6.3.3 - Intermediate Check ISO 17025 related Discussions 5
A Linearity and Eccentricity Point Definitions - Ref: ISO/IEC 17025:2005 ISO 17025 related Discussions 1
J ANSI/NCSL Z540-1-1994 vs. ANSI/ISO/IEC 17025:2000 ISO 17025 related Discussions 4
Hershal For accredited labs that expect more - Certification of ISO/IEC 17025 assessors ISO 17025 related Discussions 4
WCHorn ISO/IEC 17025 Registrars - I'm confused about "accredited" registrars for ISO 17025 ISO 17025 related Discussions 18
A Sampling - What does it mean by Sampling in ISO:IEC 17025 ISO 17025 related Discussions 1
Hershal Measurement Uncertainty - Need help with ANS/ISO/IEC 17025 and MU ISO 17025 related Discussions 0
Hershal Measurement Uncertainty Training - ANS/ISO/IEC 17025, ISO/IEC 17020 ISO 17025 related Discussions 0

Similar threads