ISO TS 24971 Updates and formal objection of the European Commission to ISO 14971
M
MIREGMGR
#21
Re: Updates on ISO TS 24971 and formal objection of the European Commission to ISO 14
If your NB won't be practical about the user being a willing and able participant in use-risk management, you could discuss with the NB the possibility that you would cease declaration of conformance to 14971, and instead conform to the Essential Requirements via your own risk procedure that is similar to 14971 except in the areas of present conflict, where your procedure would be based on common real world practice.
Elsmar Forum Sponsor
Salisburysteve
Starting to get Involved
#22
Hi,
Our NB are making a major NC out of not reviewing 14971 compliance with the BS EN 2009/2012 update. Their case is not that we haven't reduced/assessed/managed risk appropriately but that we cannot meet ER requirement to reference 'state of the art'.. Since BS EN ISO 14971:2012 is the most recent risk standard it must be state of the art..
Glad to hear someone has ditched 14971 and gone straight to ER justification.. Tempting...
Regards,
Steve
Peter Selvey
Staff member
Super Moderator
#23
I think in practice manufacturers have excessively relied on warnings in the instructions for use as a get out of jail free card in risk management. A lot of standards seem to encourage this as well.
It seems as if Deviation 7 is a knee jerk reaction to the other extreme by prohibiting the use of any information as a risk control. This is likely to be modified in the future back to something reasonable. There are simply too many cases where information to the user does form a valid risk control, with contraindications being a good example.
The real problem is a failure to properly verify the effectiveness of a risk control. Most warnings really don't reduce the risk, and therefore they should not be listed as a risk control. Risk management tables often show a risk as being significant prior to the warning, and then magically reduced once the warning is in place.
I find that a lot of warnings and cautions are a bit of a smokescreen, and if you look into the subject carefully, the you find there was no real risk in the first place, or the true risk control was somewhere else.
So while Deviation 7 is clearly wrong, it puts a spotlight on a serious issue.
For those with the stomach that would like to challenge Deviation 7, note that Europe has an effective constitution in the form of a treaty (OJ 2010/C 83/01). Article 4.4 states:
Under the principle of proportionality, the content and form of Union action shall not exceed what is necessary to achieve the objectives of the Treaties
Deviation 7, and in particular statement 7c is obviously one which could be challenged legally under the principle of proportionality. Disallowing all information to the user as a risk control obviously makes no sense.
Marcelo
Inactive Registered Visitor
#24
Also, if you use information for safety as a risk control measure by ISO 14971, you need to route them to the usability engineering process aa part of the way to verify it?s implementation and effectiveness. Only putting the info in youor risk management file is not enough, but that?s what people usually do.
G
GPjeri
#25
Re: ISO 24971 July 2013 doc
Has anyone purchased this and would you recommend getting it; did you find it helpful?
Thanks in advance...:)
Thread starter Similar threads Forum Replies Date
J Informational ISO 24971:2020 Released ISO 14971 - Medical Device Risk Management 0
M Informational ISO 14971 / ISO TR 24971 revision update – atualizações sobre a revisão Medical Device and FDA Regulations and Standards News 1
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
M ISO 14971 and ISO TR 24971 revision ISO 14971 - Medical Device Risk Management 32
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
W First time being audited (ISO 9001), asking for advice ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet ISO 26262 ISO 26262 - Road vehicles – Functional safety 3
Marc ISO 26262- Road vehicles – Functional safety ISO 26262 - Road vehicles – Functional safety 0
John C. Abnet ISO 26262 IATF 16949 - Automotive Quality Systems Standard 0
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
K PDCA cycle and ISO processes alternative model Quality Management System (QMS) Manuals 14
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S ISO 11137- Simulated product vs SIP Other Medical Device Related Standards 2
D Which ISO Standard to purchase? ISO 13485:2016 - Medical Device Quality Management Systems 7
V ISO 10360-5: 2020 Gap analysis and Action plan Excel .xls Spreadsheet Templates and Tools 1
Q ISO 9001 - Reseller Exclusions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S Inventory Listing and ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO 45001 6.1.2.1 Hazard Identification Occupational Health & Safety Management Standards 1
T The difference between ISO 14644-3:2005 and ISO 14644:2019 Other Medical Device Related Standards 2
S Any ISO standards around Artificial Intelligence and Machine Learning? Medical Information Technology, Medical Software and Health Informatics 4
R AS9100D internal audit checklist or ISO 9001 2015 to AS9100 D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
N ISO 9001 - Training business with fewer than 5 employees ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
J Opportunity in ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
P ISO 8 classified medical manufacturing room Qualification and Validation (including 21 CFR Part 11) 1
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
Sidney Vianna Release of ISO 10013:2021, Quality management systems – Guidance for documented information Other ISO and International Standards and European Regulations 0
Similar threads
Top Bottom