Hello everyone
Info:
- We have a 15m^3 EtO sterilization chamber, with a usable volume of 12m^3.
- Our current process spec validation (and subsequent requalification reviews and reduced requalifications) has been audited with no findings for 5+ years.
- We currently use 18x BIs for routine monitoring, as per the recommendation of ISO 11135:2014 Table C.3.
- (According to the table, we can actually use 16, corresponding to the usable load volume)
- BIs are placed in validated IPCDs for routine monitoring
- In 2000+ cycles over 10 years, we have never had a postive BI or device testing non-sterile.
- We have a 3h exposure time. With fractional cycles we have all-kill of 36 distributed BIs and inoculated devices at 30min exposure.
My question:
- Annex C is marked as "informative", and not "normative"
- Table C.3. is titled "Recommended number of BIs", and not "required"
- It seems that the number is not an explicit requirement of the standard, provided that a reduction in the number is robustly justified
- Can we use historic data, data from fractional cycles, temperature/RH profiles etc to justify a reduction in BI's, and limiting them to the 3x most challenging locations?
Reason:
- We do 30+ cycles per month, and the cost of BIs per cycle is substantial
- Our devices are low cost, but large - thus the production cost is heavily influenced by cost per cycle.
- 18x seems excessive considering we have never had a failed BI using this process spec in this chamber.
- As mentioned, the standard "recommends" the number in an "informative" Annex
Thanks in advance!
Info:
- We have a 15m^3 EtO sterilization chamber, with a usable volume of 12m^3.
- Our current process spec validation (and subsequent requalification reviews and reduced requalifications) has been audited with no findings for 5+ years.
- We currently use 18x BIs for routine monitoring, as per the recommendation of ISO 11135:2014 Table C.3.
- (According to the table, we can actually use 16, corresponding to the usable load volume)
- BIs are placed in validated IPCDs for routine monitoring
- In 2000+ cycles over 10 years, we have never had a postive BI or device testing non-sterile.
- We have a 3h exposure time. With fractional cycles we have all-kill of 36 distributed BIs and inoculated devices at 30min exposure.
My question:
- Annex C is marked as "informative", and not "normative"
- Table C.3. is titled "Recommended number of BIs", and not "required"
- It seems that the number is not an explicit requirement of the standard, provided that a reduction in the number is robustly justified
- Can we use historic data, data from fractional cycles, temperature/RH profiles etc to justify a reduction in BI's, and limiting them to the 3x most challenging locations?
Reason:
- We do 30+ cycles per month, and the cost of BIs per cycle is substantial
- Our devices are low cost, but large - thus the production cost is heavily influenced by cost per cycle.
- 18x seems excessive considering we have never had a failed BI using this process spec in this chamber.
- As mentioned, the standard "recommends" the number in an "informative" Annex
Thanks in advance!