ISO13485:2016 Clause 4.2.3 - Changes in Technical File Requirements
RA Lille
Dear all ,
As you know ISO 13485:2016 now contains a clause about technical file of the medical device (clause 4.2.3).
As part as this requirement has been newly written down in the ISO text, what's really new ?
we received an e-mail from our NB saying that "requirements for technical file are becoming more strict and you're asked to take them in consideration".
Does anyone know what 'more strict' means ?
Has anyone already experienced technical file audit according to new ISO 13485:2016 ?
Any idea of what will be deciphered by the auditors ?
Any idea or tool is welcome !
thanks !
yodon
Can't answer your questions but would certainly like to see this discussion advanced.
Certainly, the overall expectations are increasing. With the latest version of 60601 (3rd edition, amendment 1, I think), there's a detailed review of the Risk File (assessment of compliance to 14971), a review of software design artifacts (compliance to 62304), and a review of the Usability Engineering File (to assess compliance to IEC 62366).
My understanding is if these aren't assessed by a certified test lab, the NB will have to do the detailed review (at extra cost to the manufacturer).
While not exactly imminent, the transition from the MDD to the MDR will probably also have some impact.
Have you asked you NB for any particulars? Again, I'd certainly be interested in hearing where all this is headed.
Ronen E
Problem Solver
Moderator
Dear all ,
thanks !
I've briefly reviewed the topic and here is what I found.
There really isn't much of a material change. It's more a matter of elaborating the requirement. I think that in practice there shouldn't be grave repercussions.
ISO 13485:2003 had the following paragraph under 4.2.1 (Documentation Requirements / General):
For each type or model of medical device, the organization shall establish and maintain a file either containing or identifying documents defining product specifications and quality management system requirements (see 4.2.3). These documents shall define the complete manufacturing process and, if applicable, installation and servicing.
(The emphasis is mine, to highlight the bits that are relevant to ISO 13485:2016 4.2.3).
In essence this is almost identical to the new requirement.
Annex A to ISO 13485:2016 (comments on change between the 2016 and 2003 versions) only states (under 4.2) "Lists the documents that would be included in the medical device file". In that regard, yes, maybe the 2016 version is a little more detailed, but I don't see anything new in it except for explicitly calling out labelling, intended use and IFU (in my view these are part of the device specification, though) and the various aspects of device manufacturing (in my opinion measuring and monitoring are definitely integral parts of device manufacture). Please note that manufacturers following the MDD Annex VII had to have all that in their Tech Files anyway.
Perhaps the only significant difference is the requirement that the File will support demonstrating compliance with ISO 13485 and with applicable regulatory requirements. In the case of NB scrutiny, the latter is mostly complied with via the ER checklist (or similar) that was included in the File anyway. The former, though, as well as full comliance with the latter, might call for an extended checklist that addresses each and every requirement in ISO 13485:2016 and the applicable requirements in the relevant MDD conformity assessment Annex.
Again, I think that in practice you shouldn't see any heightened requirements, it's more a case of formalising the existing practice.
Cheers,
Ronen.
Marcelo
The final revised requirement in fact is not right (this was a problem we had by dividing the revision work in several groups).
The expectation when we revised the requirement was to have a more clear requirement for a technical file/DMR (we also created a separate file for device design in 7.3.10) because the original requirement rom 2003 mixed both.
The requirement as published requires that you have a file "containing or referencing documents generated to demonstrate conformity to the requirement of this International Standard and compliance with applicable regulatory requirements".
One interpretation of this, for example I that this file has have documents that show how to you comply with ISO 13485 (a checklist with file links, maybe?) and all applicable regulatory requirements (this would be crazy).
Please note that are several more topics in the final text of ISO 13485 that are somewhat weird if you really read the standard with scrutiny (I counted some 30 + last time I revised the standard for the last time when publishing our Brazilian version).
Hopefully some of those can be solved in the guidance document that's going to be published in 2017.
Ronen E
Problem Solver
Moderator
1. Why do you think that "would be crazy"?
2. What other interpretations do you have in mind for that clause?
Thanks,
Ronen.
Marcelo
Well, the checklist in the case of Iso 13485 would have to separate all the standard requirements with a line for each requirement (how much would that be, 200 requirements?) and show compliance with everyone of them.
Now imagine doing the same for each specific requirement in each specific regulations that you need to comply.
Besides, I don't think this would fit well in a QMS standard, why would a QMS standard require that? I tried to include a requirement for manufacturers to have at least a procedure to define how they register a product (which I think is extremely reasonable) and even this was not accepted because it would be too much of a burden (what we ended up was with the requirement that "The organization shall communicate with regulatory authorities in accordance with applicable regulatory requirements".).
Well, another one would be a mix of both, a technical file + monstrous spreadsheet with all applicable requirement of ISO 13485 and applicable regulations and the real documents that show compliance with each specific requirement.
Ronen E
Problem Solver
Moderator
Thanks for your explanations, Marcelo.
I don't see that as much different from what manufacturers are currently being expected to present on audit. For example see the ER checklist that has become the standard although it is not explicitly required in the MDD. It's not like that checklist only has a few lines / requirements...
Bottom line, regulatory authorities and certification auditors don't expect manufacturers to start looking for supporting evidence on audit day and they don't expect a lot of verbal debating over meeting the requirements. You could say that they expect to be "spoon fed" with compliance evidence, but I think it makes some sense in terms of audit efficiency, audit duration and total cost (which is eventually rolled over to manufacturers).
In my opinion that's not really another interpretation, it's just another way of implementing the same approach.
I think that s. 4.2.3 is written quite clearly and indeed there is some added burden in it, however that added burden is not directly related to the device's Tech File. Rather, it is in the need for documenting how ISO 13485:2016 is met. Seems like this was included in 4.2.3 by mistake because it's not directly related to the product; it's more about the organisation and its QMS.
Documenting regulatory compliance in detail is not a huge addition in most cases I know because the regulation often requires such documentation anyway, at least to a large extent.
Cheers,
Ronen.
RA Lille
Dear all,
thanks for this interesting debate.
As Ronen said, the supporting evidences to demonstrate that QMS is compliant with ISO13485 is not part of the MD tech file, but should be displayed pieces after pieces to feed the certification auditor.
Moreover, compliance with all applicable regulatory requirements should be supported by the general procedure for compliance with regulations and communication with competent authorities. It's RA stuff...
Cheers,
Philip B
Apologies I am a little late to this interesting debate but I have just started to amend our QMS to 13485:2016. It strikes me that 4.2.3 (a) and (b) are technical file requirements and (c) - (f) are QMS requirements. Therefore I am not sure how best to demonstrate this within our system - or perhaps it doesn't matter as long as they are addressed somewhere? Has anyone had a transition audit and can therefore shed some light on this?
b-quality
I'm confused by this discussion. In Clause 4.2.3 of ISO 13485:2016 (Medical Device File) I was told this was an expanded version of the FDA DMR. No mention was made to a Tech File. Can anyone elaborate on this?
T ISO13485:2016 Clause 6.3 - Infrastructure ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO13485:2016, MDSAP and Internal Audits ISO 13485:2016 - Medical Device Quality Management Systems 8
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 8
A ISO13485:2016 name and ownership changes questions ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO13485:2016 7.5.6 Process Validation Responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 1
D Customer Audit Finding ISO13485:2016 7.6 ISO 13485:2016 - Medical Device Quality Management Systems 7
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 6
S PPT for ISO13485:2016 Employee Training (Request) ISO 13485:2016 - Medical Device Quality Management Systems 0
qualityegghead ISO13485:2016 4.2.4 Control Of Documents - Lifetime of the Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO13485 2016 - What the upshot is of failing to upgrade on a medical diagnostic device ISO 13485:2016 - Medical Device Quality Management Systems 1
H ISO13485:2016 Measurement System - Methods General Measurement Device and Calibration Topics 4
G ISO13485:2016 Cl. 7.5.1 - Production and Service Provision - Medical Device Repairs ISO 13485:2016 - Medical Device Quality Management Systems 3
M Looking for Cross reference of AS9100:2016 to ISO13485 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M ISO13485:2016 - Exemption vs. Not Applicable ISO 13485:2016 - Medical Device Quality Management Systems 16
K Implementing ISO13485:2016 in the Middle of a New Device Project ISO 13485:2016 - Medical Device Quality Management Systems 1
A When will ISO13485 Update be Released? EU Medical Device Regulations 1
J Do Software Subcontractors need to be ISO13485 compliant in the EU? EU Medical Device Regulations 3
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
M Address change for a company with CE/ISO13485 EU Medical Device Regulations 2
A Document "Correspondence IATF 16949 vs ISO13485" available? IATF 16949 - Automotive Quality Systems Standard 0
O ISO13485 implementation - Are internal audits expected before stage 1 audit? Design and Development of Products and Processes 3
N Can a (class I) medical device be manufactured in a GMP certified site with no ISO13485? EU Medical Device Regulations 18
V Iso13485 certification vs CE technical audit ISO 13485:2016 - Medical Device Quality Management Systems 3
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Nonconformance to a Note in ISO13485 clause 6.2 Human Resources Effectiveness ISO 13485:2016 - Medical Device Quality Management Systems 13
C ISO13485 exclusions for design and automation company ISO 13485:2016 - Medical Device Quality Management Systems 2
R ISO13485/ISO9001 Convenient Document Management System Document Control Systems, Procedures, Forms and Templates 3
R ISO9001 & ISO13485 QMS for 2 types of products ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M ISO13485 Documentation packs (manual, procedures, forms etc) ISO 13485:2016 - Medical Device Quality Management Systems 3
C ISO13485/9001 clause 6.3 (Infrastructure) ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO13485 cert expired - Product still CE Marked CE Marking (Conformité Européene) / CB Scheme 1
V Revised ISO std vs ISO13485:2003 - Help Please!!! ISO 13485:2016 - Medical Device Quality Management Systems 4
C Is there a strong market out there for ISO13485 certification? ISO 13485:2016 - Medical Device Quality Management Systems 11
B Must a Design Outsourcing Company have ISO13485? ISO 13485:2016 - Medical Device Quality Management Systems 3
L ISO13485 past practice labeling vs customer requirement Other ISO and International Standards and European Regulations 2
S CMDCAS ISO13485 - Bilateral agreement with SAI global to conduct Certification Audits ISO 13485:2016 - Medical Device Quality Management Systems 7
Q Multiple Quality Standards - AS9100 and ISO13485/9001 Cross Reference Matrix Quality Manager and Management Related Issues 3
C ISO13485 Incoming Receiving Parts Verification ISO 13485:2016 - Medical Device Quality Management Systems 13
Q ISO13485 Section 7.5.1.1 and 7.5.2.1 for a Distributor ISO 13485:2016 - Medical Device Quality Management Systems 2
W Setting up from scratch - Looking for help (ISO9001 / ISO13485) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Outsource Production Line need ISO13485 Certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
somashekar Clause 7.5.5. as in the text and in the Annex B (ISO13485:2003) ISO 13485:2016 - Medical Device Quality Management Systems 1
U Do we have to have foreign legal documents? (ISO13485 4.2.3 document control) ISO 13485:2016 - Medical Device Quality Management Systems 3
0 Medical Device Distribution Centre and Implementing ISO13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
C Writing a Sterilization Procedure when the Process is Subcontracted - ISO13485 Quality Management System (QMS) Manuals 1
J ISO13485 Certification Scope - Registering a Subset of our Products ISO 13485:2016 - Medical Device Quality Management Systems 9
J ISO13485 Clause exclusions for a Medical Software Company ISO 13485:2016 - Medical Device Quality Management Systems 9
P How to streamline different documents in a multi plant ISO13485 QMS ISO 13485:2016 - Medical Device Quality Management Systems 8
D FDA Guidance on Interim Use of ISO13485 for Medical Device Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 3
S Implementing ISO13485 in a new startup company - Neurological devices ISO 13485:2016 - Medical Device Quality Management Systems 12