My company's ISO13485 certificate is expiring shortly due to issues with the NB and we may be in a situation where we will not be certified to ISO13485, but the products are CE Marked until 2018.

Can the products still be sold in Europe, while we wait for teh cert to be issued

How does this affect distribution to the US? We are compliant with 21 CFR 820 and have undergone 2 recent FDA audits with no observations.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Hi,

The MDD does not mandate ISO 13485 certification, so the fact that you won't be ISO 13485 certified does not negate your EU compliance (and the right to CE Mark).

Technically, you may be EC certified until 2018, but practically that certification needs to be ascertained annually, via a NB surveillance audit. In your next such audit you will have to convince the NB that your organization meets the quality management system requirements in the MDD, that apply. Even then, ISO 13485 certification (or compliance) is a preferred path, but it's certainly not the only path. If you comply with the applicable MDD requirements (in the relevant annex), that should suffice. So, don't despair.

From a regulatory standpoint, your ISO 13485 certification (or lack thereof) should not affect your situation in the USA. Compliance with 21 CFR part 820, especially one concurred by the FDA, is enough. There may, however, be some marketing significance because some organizations require their suppliers to be ISO 13485 certified.

Cheers,
Ronen.
 
Thread starter Similar threads Forum Replies Date
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
M Address change for a company with CE/ISO13485 EU Medical Device Regulations 2
A Document "Correspondence IATF 16949 vs ISO13485" available? IATF 16949 - Automotive Quality Systems Standard 0
O ISO13485 implementation - Are internal audits expected before stage 1 audit? Design and Development of Products and Processes 3
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 8
N Can a (class I) medical device be manufactured in a GMP certified site with no ISO13485? EU Medical Device Regulations 18
V Iso13485 certification vs CE technical audit ISO 13485:2016 - Medical Device Quality Management Systems 3
A ISO13485:2016 name and ownership changes questions ISO 13485:2016 - Medical Device Quality Management Systems 2
D ISO13485:2016 7.5.6 Process Validation Responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 1
D Customer Audit Finding ISO13485:2016 7.6 ISO 13485:2016 - Medical Device Quality Management Systems 7
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 5
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S PPT for ISO13485:2016 Employee Training (Request) ISO 13485:2016 - Medical Device Quality Management Systems 0
qualityegghead ISO13485:2016 4.2.4 Control Of Documents - Lifetime of the Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
S Nonconformance to a Note in ISO13485 clause 6.2 Human Resources Effectiveness ISO 13485:2016 - Medical Device Quality Management Systems 13
C ISO13485 exclusions for design and automation company ISO 13485:2016 - Medical Device Quality Management Systems 2
R ISO13485/ISO9001 Convenient Document Management System Document Control Systems, Procedures, Forms and Templates 3
S ISO13485 2016 - What the upshot is of failing to upgrade on a medical diagnostic device ISO 13485:2016 - Medical Device Quality Management Systems 1
R ISO9001 & ISO13485 QMS for 2 types of products ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M ISO13485 Documentation packs (manual, procedures, forms etc) ISO 13485:2016 - Medical Device Quality Management Systems 3
H ISO13485:2016 Measurement System - Methods General Measurement Device and Calibration Topics 4
C ISO13485/9001 clause 6.3 (Infrastructure) ISO 13485:2016 - Medical Device Quality Management Systems 3
G ISO13485:2016 Cl. 7.5.1 - Production and Service Provision - Medical Device Repairs ISO 13485:2016 - Medical Device Quality Management Systems 3
M Looking for Cross reference of AS9100:2016 to ISO13485 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M ISO13485:2016 - Exemption vs. Not Applicable ISO 13485:2016 - Medical Device Quality Management Systems 16
K Implementing ISO13485:2016 in the Middle of a New Device Project ISO 13485:2016 - Medical Device Quality Management Systems 1
T ISO13485:2016 Clause 6.3 - Infrastructure ISO 13485:2016 - Medical Device Quality Management Systems 1
R ISO13485:2016 Clause 4.2.3 - Changes in Technical File Requirements ISO 13485:2016 - Medical Device Quality Management Systems 18
V Revised ISO std vs ISO13485:2003 - Help Please!!! ISO 13485:2016 - Medical Device Quality Management Systems 4
C Is there a strong market out there for ISO13485 certification? ISO 13485:2016 - Medical Device Quality Management Systems 11
B Must a Design Outsourcing Company have ISO13485? ISO 13485:2016 - Medical Device Quality Management Systems 3
L ISO13485 past practice labeling vs customer requirement Other ISO and International Standards and European Regulations 2
S CMDCAS ISO13485 - Bilateral agreement with SAI global to conduct Certification Audits ISO 13485:2016 - Medical Device Quality Management Systems 7
quality1 Multiple Quality Standards - AS9100 and ISO13485/9001 Cross Reference Matrix Quality Manager and Management Related Issues 3
C ISO13485 Incoming Receiving Parts Verification ISO 13485:2016 - Medical Device Quality Management Systems 13
Q ISO13485 Section 7.5.1.1 and 7.5.2.1 for a Distributor ISO 13485:2016 - Medical Device Quality Management Systems 2
W Setting up from scratch - Looking for help (ISO9001 / ISO13485) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Outsource Production Line need ISO13485 Certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
somashekar Clause 7.5.5. as in the text and in the Annex B (ISO13485:2003) ISO 13485:2016 - Medical Device Quality Management Systems 1
U Do we have to have foreign legal documents? (ISO13485 4.2.3 document control) ISO 13485:2016 - Medical Device Quality Management Systems 3
0 Medical Device Distribution Centre and Implementing ISO13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
C Writing a Sterilization Procedure when the Process is Subcontracted - ISO13485 Quality Management System (QMS) Manuals 1
J ISO13485 Certification Scope - Registering a Subset of our Products ISO 13485:2016 - Medical Device Quality Management Systems 9
J ISO13485 Clause exclusions for a Medical Software Company ISO 13485:2016 - Medical Device Quality Management Systems 9
P How to streamline different documents in a multi plant ISO13485 QMS ISO 13485:2016 - Medical Device Quality Management Systems 8
D FDA Guidance on Interim Use of ISO13485 for Medical Device Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 3
S Implementing ISO13485 in a new startup company - Neurological devices ISO 13485:2016 - Medical Device Quality Management Systems 12
E Career Plan - ISO13485 Lead Auditor Course/IRCA certified ISO9001 Lead Auditor Course Professional Certifications and Degrees 4
S Scope of ISO13485 and the term 'Related Services' ISO 13485:2016 - Medical Device Quality Management Systems 1
P What's the difference between ISO13485 and FDA/QSR? ISO 13485:2016 - Medical Device Quality Management Systems 16

Similar threads

Top Bottom