ISO14971:2019 - Verification of implementation and effectiveness of risk control

mithilesh

Registered
Hi All

I am trying to implement ISO14971:2019 in our companies risk management procedure and have a question on implementation of risk control measure:

ISO 14971:2019 requires implementation of risk control measures, verification of implementation and verification of the effectiveness of those risk control measures.

Does anyone have any specific examples on how to verify implementation of risk control vs. how to verify effectiveness of the risk controls.

One example i can think of is: for a medical device such as an insulin pen injector; one of the risk control was the OD of pen system to control the dosing of the pen. The implementation of the risk control would then be the drawing and the design verification test protocol. The verification of the risk control would then be the design verification test report .

Can you guys provide other specific examples. For example..if i identify a risk control as Instruction for Use..what would be the verification of implementation and verification of effectiveness for this?

Thank you
 

yodon

Leader
Super Moderator
@monoj mon is right - verification confirms the content is included in the IFU and validation (probably via a summative study per 62366) would validate that the instructions were understandable. Bear in mind that if you consider the content deviation from EN ISO 14971:2012 that says labeling alone cannot reduce risk, "effectiveness" is not really a consideration, you just want to be sure that you have communicated in an understandable manner.
 

ThatSinc

Quite Involved in Discussions
look at what the hazardous situation and harm is, and then look at what your control measure is - how can you prove that the control measure has reduced the probability of the hazardous situation or harm from occurring.

some abbreviated potentials below from previous experience;

Risk/Issue - Control Measure - VoE
Device damaged in transit - Robust shipping boxes - transit validation
Device degrades over time (fluid in a bottle) - device formulation for long shelf life / packaging size to ensure it's used in a timely manner - stability studies for life / validation on time taken for users to exhaust a bottle following opening
Skin sensitivity to device - product formulation - ISO 10993 mandated studies on sentisation etc.


@monoj mon Bear in mind that if you consider the content deviation from EN ISO 14971:2012 that says labeling alone cannot reduce risk, "effectiveness" is not really a consideration, you just want to be sure that you have communicated in an understandable manner.

I thought that this myth was debunked a long time ago?

Instructions for safety and disclosure of residual risk are two completely separate entities.

Putting a warning on a UV lamp that says "Exposure to UV may cause skin cancer" does not reduce risk. Providing an instruction that users must wear appropriate safety equipment, etc. If shown to be understood, and verified as effective through any usability studies, certainly can reduce the risk.
 

zoneofindifference

Starting to get Involved
look at what the hazardous situation and harm is, and then look at what your control measure is - how can you prove that the control measure has reduced the probability of the hazardous situation or harm from occurring.

some abbreviated potentials below from previous experience;

Risk/Issue - Control Measure - VoE
Device damaged in transit - Robust shipping boxes - transit validation
Device degrades over time (fluid in a bottle) - device formulation for long shelf life / packaging size to ensure it's used in a timely manner - stability studies for life / validation on time taken for users to exhaust a bottle following opening
Skin sensitivity to device - product formulation - ISO 10993 mandated studies on sentisation etc.




I thought that this myth was debunked a long time ago?

Instructions for safety and disclosure of residual risk are two completely separate entities.

Putting a warning on a UV lamp that says "Exposure to UV may cause skin cancer" does not reduce risk. Providing an instruction that users must wear appropriate safety equipment, etc. If shown to be understood, and verified as effective through any usability studies, certainly can reduce the risk.

Any thoughts on whether you could point to drug product characterizations studies as either verification of implementation of a risk control measure or verification of effectiveness for a combination product?
 

yodon

Leader
Super Moderator
I suppose there's a case if you are, say, mitigating overdose and the (combo) product is limited to an amount that cannot result in overdose. The design can be verified to limit the dosage to a safe amount and the characterization study might support that the limit is effective for preventing overdose.
 

Fritz

Registered
Hi everybody

I'm also unsure what is the "implementation of risk control" and the "verification of effectiveness of the risk controls" for an specific example.
  1. Events leading to the harm: Insufficent undersized design of the prothesis to resist loadings during activities of daily living -> Fatigue fracture of the prothesis -> Fall of the patient
  2. Harm: Bone fracture
  3. Risk control: Pre-clinical endurance testing.
  4. Implementation of the risk control: a) Standard operating procedure demanding pre-clinical endurance testing in accordance with ISO 10328. b) Work instructions how to perform testing in accordance with ISO 10328. c) A specific test description including an acceptenace criterion based on ISO 10328.
  5. Verification of effectivness of the risk control: Mechanical test report that the device passed the test in accordance with ISO 10328.

Is this correct or is the mechanical test report the "Implementation of the risk control"? But what is the "Verification of effectivness of the risk control" then?

Thank you
 

ThatSinc

Quite Involved in Discussions
Everything you have listed there is a part of your verification of effectiveness, and ongoing production controls

The actual control measure is what you have done to ensure that the design doesn't fracture.

The clue is in the events you have listed about the part being undersized.

Your risk control measure is the dimensions of the part, your VoI is the technical drawing of the part showing the dimensions.

Your VoE is as you have everything there, the evidence that shows the part doesn't fail under the loads defined by the user and use specifications, or the standard in this case.
 
Last edited:

Fritz

Registered
Hi ThatSinc

Thanks for the explanation.

I'm asking because others might say, that the verification of effectivness of the risk control to reduce the harm (the injury of the prothesis user) can only be shown by a clinical study or post market surveillance.

Kind regards
 

ThatSinc

Quite Involved in Discussions
If it can only be shown by a clinical study or post market surveillance and the risk without it is unacceptable - I guess you'd better do a clinical study rather than find out it's not effective through post market surveillance.

This example seems to have it covered through the standard, if the standard has tensile strength requirements and test methods?
 
Thread starter Similar threads Forum Replies Date
O ISO14971 for EUA ISO 14971 - Medical Device Risk Management 3
H QMS ISO 13485:2016 - ISO14971 IEC60304 etc ISO 13485:2016 - Medical Device Quality Management Systems 6
J Should a company create different SOPs for ISO14971? ISO 14971 - Medical Device Risk Management 1
E ISO14971 - Determination of possible hazards (SU, UU, PC) ISO 14971 - Medical Device Risk Management 3
S KPIs (Key Performance Indicators) and Risk Management Process (ISO14971:2007) ISO 14971 - Medical Device Risk Management 5
N post production Information as per ISO14971 ISO 14971 - Medical Device Risk Management 1
S ISO 14971 and 13485 - Is ISO14971 just a supplement to ISO13485? ISO 14971 - Medical Device Risk Management 2
N Ansi/aami/iso13485:2003 + Iso14971:2000 + Iec60601-1-4:2000 + Mdd93/42/eec IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A ISO14971:2001 - Risk Management for Medical Devices - Help need copy ISO 14971 - Medical Device Risk Management 10
D EN 61010-1:2010 vs 61010-1:2010+A1:2019 CE Marking (Conformité Européene) / CB Scheme 2
G Brazil Regulations RDC 270/2019 Other Medical Device Related Standards 0
B Risk Management Procedure updates needed for 14971:2019 ISO 14971 - Medical Device Risk Management 11
M AIAG-VDA DFMEA 2019 - it is compulsory? FMEA and Control Plans 7
B ISO 14971:2019 amendment A11:2021 questions ISO 14971 - Medical Device Risk Management 5
T SQL Server 2019 - Master Data Services - Validation needed? ISO 13485:2016 - Medical Device Quality Management Systems 4
M ISO 14971:2019 vs FMEA methodology ISO 14971 - Medical Device Risk Management 23
Y BS EN ISO 14971:2019+A11:2021 released ISO 14971 - Medical Device Risk Management 3
R What are the new changes in EN ISO 11137-1:2015+A2:2019? Other Medical Device Related Standards 2
T The difference between ISO 14644-3:2005 and ISO 14644:2019 Other Medical Device Related Standards 2
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 12
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 16
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
J FMEA Handbook 2019 IATF 16949 - Automotive Quality Systems Standard 3
classical_quality Reflecting: Improvements Made 2019 & 2020? Coffee Break and Water Cooler Discussions 2
Aymaneh ISO 11607-1: 2019 main changes Other Medical Device Related Standards 2
K Overall residual risk according to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 12
Bill Hansen New ISO 14971:2019 Harm: unreasonable psychological stress, and cybersecurity ISO 14971 - Medical Device Risk Management 13
dgrainger Informational UK - Medicines and Medical Devices Bill 2019-21 UK Medical Device Regulations 0
Ajit Basrur Why is ISO 14644-3:2019 not approved by US? Other Medical Device Related Standards 2
T Implant Card - Article 18.1(a) and MDCG 2019-8 clarification EU Medical Device Regulations 3
Richard Regalado Top 10 operational risks of 2019 for business continuity planning Business Continuity & Resiliency Planning (BCRP) 6
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
J UK Medicines and Medical Devices Bill - 2019 ISO 13485:2016 - Medical Device Quality Management Systems 0
A ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations Other Medical Device Related Standards 12
Ronen E Informational What's new in ISO 14971:2019 ISO 14971 - Medical Device Risk Management 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
Sidney Vianna IAQG News IAQG has a new website - December 2019 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
Y When will Notified Bodies require MedDev manufacturers to fully implement ISO 14971:2019? ISO 14971 - Medical Device Risk Management 1
Sidney Vianna Continual decline of ASQ - can it be saved? Nov 2019 ASQ - American Society for Quality 29
B Four Key Performance Indicators for Continual Improvement (Dec. 3 2019) [Deleted] Training - Internal, External, Online and Distance Learning 2
Richard Regalado ISO 22301:2019 has been published - Nov 2019 Business Continuity & Resiliency Planning (BCRP) 0
Marc Fiat Chrysler merges with PSA Group, owner of Peugeot and Citroën - Oct 2019 World News 0
M Informational ANVISA – CONSULTA PÚBLICA Nº 734, DE 22 DE OUTUBRO DE 2019 – Registro de dispositivos médicos para diagnóstico in vitro (in vitro diagnostic medical m Medical Device and FDA Regulations and Standards News 0
M Informational ANVISA – CONSULTA PÚBLICA No 730, DE 14 DE OUTUBRO DE 2019 – Registro de dispositivos médicos (Public consultation – registration of medical devices) Medical Device and FDA Regulations and Standards News 0
Marc From the "I'm glad I'm retired" Department (Oct 2019) Career and Occupation Discussions 7
M Informational EU – Minutes of the 24 July 2019 SCHEER Working Group on safety of breast implants in relation to anaplastic large cell lymphoma (BIA-ALCL) meeting Medical Device and FDA Regulations and Standards News 0
M Informational EU – Flyer – Call for expression of interest for expert panels on medical devices and in vitro diagnostic medical devices (2019/C 323/05) Medical Device and FDA Regulations and Standards News 0
M Informational EU – MDCG 2019-12 Designating authority’s final assessment form: Key Information Medical Device and FDA Regulations and Standards News 0
M Informational EU – MDCG 2019-6 v2 Questions and answers: Requirements relating to notified bodies Medical Device and FDA Regulations and Standards News 0

Similar threads

Top Bottom